Wiki Extreme Makeover – Derivatizations in Chromatography – Part 2 LC
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • New
  • Consumables
  • A Beginner’s Guide to Hydrophobic Interaction Chromatography
  • A simple tip to protect your columns
  • A Tip for Preparing Robust and Consistent Mobile Phases
  • Achieve accurate and repeatable gas flow meter measurements
  • AdvanceBio Columns Blog Series
  • Agilent 123 Meter Sodium probe maintenance
  • Agilent Collection of Columns, Supplies, and Standards Resources
  • Analysis of microplastics in the environment
  • Analyzing ADCs by HIC
  • Avoiding downtime in the lab: top tips for GC/MS success
  • Best Practices for Aqueous Mobile Phases
  • Best Practices for Making Good Connections
  • Bursting Tubing and Columns (GC and HPLC)
  • Calculating Column Volume
  • Cannabis Potency Testing: a Reliable, Cost-Effective Method
  • Carbohydrate Analyses in LC
  • Checking Your Intuition - Sub 2 µm vs Superficially Porous
  • Choosing the right GC Injection Technique
  • Choosing the right pore size for size exclusion chromatography
  • Columns, Supplies, and Standards Knowledgebase
  • Consumables Applications and Workflows
  • Consumables Recommended Supplies Lists for Agilent.com
  • Custom Product Request
  • Extreme Makeover – Derivatizations in Chromatography – Part 1 GC
  • Extreme Makeover – Derivatizations in Chromatography – Part 2 LC
  • Fake It Until You Make It: When BioInert Isn’t an Option
  • Flipping Amino Acid Analysis on Its Head
  • Glycans at a glance:  Analyzing therapeutic glycoproteins
  • Handle and Care of Syringes
  • Help! My Peaks Look Strange - Fronting and Tailing in GC
  • Help! My Peaks Look Strange - Saddle Points - LC/GC Troubleshooting
  • Hydrophobic Interaction Chromatography of Proteins and mAbs
  • Importance of Silica Particle Strength for Sub-2 µm SEC Columns
  • KB: Ferrules recommended for GC self-tightening column nut
  • LC Method Translation - the Dwell Volume
  • Minimize spectroscopy workflow disruptions
  • Minimizing Metals for Best HILIC Results
  • More than just a drink: Analyzing the elemental composition of beer
  • Multi-Attribute Methods – Peptide Mapping Part IV
  • Must See Webinars
  • Nomenclature of CFC's/Freons/Halons/Coolants
  • Oligonucleotide Analysis - Unexpected Details Matter
  • Optimizing Bonding Chemistry for Sub-2 µm SEC Particles
  • Pass the Salt, Please – Mobile Phase Preparation for HIC
  • Pesticides and their stability during GC analyses
  • Pre-Columns - the forgotten art of using retention gaps
  • Problematic polar analytes? Hello HILIC…
  • Protecting your laboratory productivity
  • Recommended Reading
  • Sample Prep Pointers - Peptide Mapping Part I
  • Save your results with sample filtration
  • Simplified cone inspection with the new Agilent LED measuring magnifier
  • Software tool for the ADM Flow Meter (G6691A)
  • Software-Supported Method Development - the Scanview Program
  • Stay Safe: A Win-Win for Solvent Storage
  • Streamline your sample processing
  • The importance of chemical composition for vial performance
  • Tips & Tricks for Amino Acid Analysis – Part I
  • Tips & Tricks for Amino Acid Analysis – Part II
  • Tips & Tricks for Amino Acid Analysis – Part III
  • Tips & Tricks for Amino Acid Analysis – Part IV
  • Tips for Smooth Sailing with HIC
  • Troubleshooting HPLC autosamplers
  • Troubleshooting HPLC degassers
  • Troubleshooting Sequence Coverage – Peptide Mapping Part III
  • UV, MS, TFA, and Formic Acid – What to use? Peptide Mapping Part II
  • You Need Lamps or Chemical Standards for Atomic Absorption Single-Element Analyses?
Still Need Help?

Post your question in our User Forum or Contact Support.

Extreme Makeover – Derivatizations in Chromatography – Part 2 LC

Created by Dr. Know Dr. Know over 1 year ago | Last modified by Dr. Know Dr. Know over 1 year ago

Part 2 – Liquid Chromatography

N. Reuter*, ACG Global Technical Support for CSD, Middelburg, The Netherlands

Introduction

Wikipedia defines “derivatization“ as follows:

 

Derivatization is a technique used in chemistry, which transforms a chemical compound into a product of similar chemical structure, called a derivative.

Generally, a specific functional group of the compound participates in the derivatization reaction and transforms the educt to a derivative of deviating reactivity, solubility, boiling point, melting point, aggregate state, or chemical composition. Resulting new chemical properties can be used for quantification or separation of the educt. Derivatization techniques are frequently employed in chemical analysis of mixtures

 

That means derivatization is “wet“ chemistry.  Though chromatographers may not prefer this type of chemistry, it is quite often used for specific purposes, as mentioned in the first paragraph of the word definition. Something similar, but different.

 

 

In HPLC, derivatizations can be more complex than their GC counterparts, because of the condensed state in which everything takes place. Increasing volatility is only a possible issue for LC/MS applications. Where-as all derivatizations in GC are pre-column, in HPLC you also can have post-column processes.


Classic Pre-Column Derivatizations

Pre-column derivatizations are carried out as off-line chemical reactions between reagent and sample prior to the injections on the HPLC column.

 

Figure 1: Flow path of a standard HPLC analysis with pre-column derivatization.


Chromotags

If volatility is not the issue, then detectability is. The classic detector in HPLC is the UV detector, so com-ponents need to be UV-active to be detected. If the molecule is not, you need to insert so-called chromophores into the molecules. Reagents that do this are called chromotags.  This table lists common chromotag reagents:

                                                                                                                                                                                               

Reagent Abbreviation

Full Name

Dbs-Cl (Dabsyl chloride)

4-Dimethylamino-azobenzene-4'-sulfonic    acid chloride

DNBA

3,5-Dinitrobenzyloxyamine hydrochloride

DNBC

3,5-Dinitrobenzoylchloride

DNBDI

O-(3,5-Dinitrophenylmethyl)-N,N-diisopropylthiourea

DNPBA

3,5-Dinitrophenylmethyl n-propylamine hydrochloride

FDNB

Fluoro-2,4-dinitrobenzene

NIC

1-Naphthylisocyanate

Ninhydrine

4,5-Benzo-2,2-dihydroxy-cyclopentan-1,3-dione

PBPB

p-Bromophenyl bromomethyl keton

PITC

Phenylisothiocyanate

PNBA

O-(p-Nitrophenylmethyl)hydroxylamine    hydrochloride

PNBA

O-(p-Nitrophenylmethyl)hydroxylamine    hydrochloride

PNBDI

O-(p-Nitrophenylmethyl)-N,N-diisopropylthiourea

PNBPA

p-Nitrophenylmethyl    n-propyl amin hydrochloride

SDPNA

N-Succinimidyl-3,5-dinitrophenylacetate

SNPA

N-Succinimidyl-p-nitrophenylacetate

Table 1: Common chromotag reagents.

 

All of these reagents carry at least one aromatic ring or even larger pi-electron systems, which are responsi-ble for the UV detection. Dabsylchloride can, for example, be used to derivatize amines, thiols, imidazoles, phenols and alcohols.

 

Figure 2: Dabsylation of alcohols.

 

A typical procedure is as follows:

General procedure: Derivatives are formed within 30 minutes at room temperature or at 70 °C in 10 minutes in acetone, buffer or acetonitrile.

Example for alcohols: Mix 1 mL of alcohol with 10 mL 100 mM sodium bicarbonate and 20 mL 10 mM dabsyl chloride, reflux for 10 minutes, cool, extract with diethylether, wash with pH 3 citrate buffer, evaporate the organic layer, reconstitute and inject an aliquot.

 

 


Fluorotags

Besides the standard UV detection, fluorescence detection is also often used in HPLC. Chemical reagents called fluorotags are used to introduce groups with fluorescent properties into the molecules. Table 2 shows a list of common fluorotag reagents:

                                                                                                                                                                                     

Reagent Abbreviation

Full Name

AECF

2-(9-Anthranyl)ethyl chloroformate

BDMQ

3-Bromomethyl-6,7-dimethoxy-1-methyl-2(1H)-quinoxalinone

BMC    (BrMmC)

4-Bromomethyl-7-methoxycoumarine

BrMaC

4-Bromomethyl-7-acetoxycoumarine

DANS    hydrazide

1-Dimethylaminonaphthalene-5-sulfonic    acid hydrazide, Dansyl hydrazide

Dithiobis(amino-dimethoxybenzene

2,2'-Dithiobis(1-amino-4,5-dimethoxybenzene)

Dns-Cl (DANS-Cl)

1-Dimethylaminonaphthalene-5-sulfonic acid chloride, Dansyl chloride

EDTN

1-Ethoxy-4-(dichloro-s-triazinyl) naphthalene

Fluorescamine    (FLURAM)

4-Phenylspiro(furan-2(3H),1'-phthalan)-3,3'-dione

FMOC

(9-fluorenyl)methyl chloroformate

HCPI

2-(4-Hydrazinocarbonylphenyl)-4,5-diphenylimidazole

NBD-Cl

7-chloro-4-nitrobenzyl-2-oxa-1,3-diazole

NBD-F

7-Fluoro-4-nitrobenzyl-2-oxa-1,3-diazole

NDA

Naphthalene dicarboxaldehyde

OPA

o-Phthaldialdehyde

Table 2: List of common fluorotag reagents.

 

(9-Fluorenyl)alkyl chloroformates, like the methyl derivative, FMOC, or the ethyl derivative, FLEC, are often used for amines and amino acids. The ethyl derivative carries an asymmetrically substituted/chiral carbon atom and is also available enantiomerically pure in both enantiomers, so that it is possible to check for both enantiomers of, for example, an amino acid due to the formation of diastereomeric derivates on an HPLC column with an achiral phase:

 

Figure 3: Derivatization of (S)-alanine with (S)-1-(9-fluorenyl)ethyl chloroformate (FLEC).

 

A typical procedure for chloroformate reagents is as follows:

In the absence of water: Add 5 µL of the chloroformate to 2-10 µL of the sample in acetone in a mix-ture of acetonitrile/pyridine/alcohol (22:2:1, alcohol: methanol, ethanol or chloroethanol) and shake for some seconds. Inject.

In the presence of water: To 50 µL of a pyridine-water solution (5:1), 50 µL acetonitrile/ethanol (2:3) and 5 µL chloroformate are added. Extraction is performed by 100 µL hexane and 200 µL water. In-ject an aliquot of the hexane layer.

 

Fluorotags are often also chromotags due to extended aromatic systems in the reagent’s structure, but chromotags typically are not also fluorotags.

Volatility Reagents for LC/MS

This is about acylation reactions with perfluorinated organic acid anhydrides, like trifluoroacetic acid anhy-dride or the homologous pentafluoropropanoic and heptafluorobutanoic acid anhydrides. Acylation reactions are described in part one of this document, which was about gas chromatography.


Post-Column Derivatizations/Online Derivatizations

Post-column derivatizations are online chemical reactions, carried out after the separation has taken place and before the component enters the detector:

 

Figure 4: Flow path of a standard HPLC analysis with post-column derivatization.

 

An example for the method used in Figure 4 is the post-column fluorotag derivatization of sulfonamides [1]:

 

Figure 5: Post-column derivatization of Sulfanilamide (blue) with benzene-1,2-dicarbaldehyde (OPA) and 2-sulfanylethanol. The exitation can be measured at 302 nm and the emission at 412 nm wavelengths.

 

Another possibility for post-column reactions are immobilized enzyme reactors (IMERs), where an enzyme is used to catalyze certain reactions before the component goes to the detector. An example is the reaction of acetylcholine to choline and the oxidation from choline to carboxy-N,N,N-trimethylmethanaminium, an acid and hydrogen peroxide, that can be easily detected by an electrochemical detector down to low picomole/femtomole levels [2]. The first reaction (1) is catalyzed by acetylcholine esterase and the second (2) by choline oxidase, two enzymes immobilized in the IMER reactor.

 

Figure 6: enzyme-catalzed derivatization from  acetylcholine to choline and the corresponding acid.

 

 


Drawbacks

If you inject a sample together with an unreacted  derivatization reagent or their reaction products, it can happen that the  baseline noise increases, but normally the detection is so specific that the  sensitivity still increases.

 

For post-column systems you will need another pump for the  reagent and the mixing/reaction chamber.

 

Compared to the amount of standard reagents in gas  chromatography, the amount for HPLC reagent is enormously higher and you have  lots of specific reactions for specific components. In fact, every reaction in  organic chemistry may be a derivatization reaction for HPLC components.


Conclusions

Derivatizations in liquid chromatography are used to:

 

  • Insert UV active groups
  • Insert fluorescent groups
  • Insert or generate electro-chemically measurable  groups
  • Enhance volatility for LC/MS applications

 


References

  • P.  Viñas, C. L. Erroz, N.  Campillo, M. Hernández-Córdoba; ”Determination of sulfonamides in foods by  liquid chromatography with post-column fluorescence derivatization”, J.  Chromatogr. A, 726, 1996, 125-131.
  • G.  Damsma, D. Lammerts van Bueren, B. H. C. Westerink, A. S. Horn; “Determination  of acetylcholine and choline in the femtomole range by means of HPLC, a  post-column enzyme reactor, and electrochemical detection“, Chromatographia,  24(1), 1987, 827-831.

 


Recommended Literature

  • G. Lunn, L. C. Hellwig; “Handbook of  Derivatization Reactions for HPLC”, Wiley-Intersciece, New York 1998. ISBN  978-0471164585.
  • D. R. Knapp; “Handbook of Analytical  Derivatization Reactions”, Wiley-Intersciece, New York 1979. ISBN  978-0471034698.
  • K. Blau, J. M. Halket; “Handbook of Derivatives  for Chromatography”, Wiley-Interscience, New York 1993. ISBN 978-0471926993.
  • T. Toyo'oka; “Modern Derivatization Methods for  Separation Science”, Wiley-Interscience, New York 1999. ISBN 978-0471983644.
  • S. C. Moldoveanu, V. David; “Sample Preparation  in Chromatography” (J. Chromatogr. Library Vol. 65), Elsevier Scientific,  Amsterdam 2002. ISBN 978-0444503947.
  • LC
  • General
  • LC Columns
  • derivatization
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help