Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Consumables
  • Technical Areas
  • More
Consumables
Wiki Optimizing Bonding Chemistry for Sub-2 µm SEC Particles
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • New
Consumables requires membership for participation - click to join
  • Wiki
  • A Beginner’s Guide to Hydrophobic Interaction Chromatography
  • A simple tip to protect your columns
  • A Tip for Preparing Robust and Consistent Mobile Phases
  • Achieve accurate and repeatable gas flow meter measurements
  • AdvanceBio Columns Blog Series
  • Agilent 123 Meter Sodium probe maintenance
  • Agilent Collection of Columns, Supplies, and Standards Resources
  • Analysis of microplastics in the environment
  • Analyzing ADCs by HIC
  • Avoiding downtime in the lab: top tips for GC/MS success
  • Best Practices for Aqueous Mobile Phases
  • Best Practices for Making Good Connections
  • Bursting Tubing and Columns (GC and HPLC)
  • Calculating Column Volume
  • Cannabis Potency Testing: a Reliable, Cost-Effective Method
  • Checking Your Intuition - Sub 2 µm vs Superficially Porous
  • Choosing the right GC Injection Technique
  • Choosing the right pore size for size exclusion chromatography
  • Columns, Supplies, and Standards Knowledgebase
  • Consumables Applications and Workflows
  • Consumables Recommended Supplies Lists for Agilent.com
  • Custom Product Request
  • Extreme Makeover – Derivatizations in Chromatography – Part 1 GC
  • Extreme Makeover – Derivatizations in Chromatography – Part 2 LC
  • Fake It Until You Make It: When BioInert Isn’t an Option
  • Flipping Amino Acid Analysis on Its Head
  • Glycans at a glance:  Analyzing therapeutic glycoproteins
  • Handle and Care of Syringes
  • Help! My Peaks Look Strange - Fronting and Tailing in GC
  • Help! My Peaks Look Strange - Saddle Points - LC/GC Troubleshooting
  • Hydrophobic Interaction Chromatography of Proteins and mAbs
  • Importance of Silica Particle Strength for Sub-2 µm SEC Columns
  • KB: Ferrules recommended for GC self-tightening column nut
  • LC Method Translation - the Dwell Volume
  • Minimize spectroscopy workflow disruptions
  • Minimizing Metals for Best HILIC Results
  • More than just a drink: Analyzing the elemental composition of beer
  • Multi-Attribute Methods – Peptide Mapping Part IV
  • Must See Webinars
  • Nomenclature of CFC's/Freons/Halons/Coolants
  • Oligonucleotide Analysis - Unexpected Details Matter
  • Optimizing Bonding Chemistry for Sub-2 µm SEC Particles
  • Pass the Salt, Please – Mobile Phase Preparation for HIC
  • Pesticides and their stability during GC analyses
  • Pre-Columns - the forgotten art of using retention gaps
  • Problematic polar analytes? Hello HILIC…
  • Protecting your laboratory productivity
  • Recommended Reading
  • Sample Prep Pointers - Peptide Mapping Part I
  • Save your results with sample filtration
  • Simplified cone inspection with the new Agilent LED measuring magnifier
  • Software tool for the ADM Flow Meter (G6691A)
  • Software-Supported Method Development - the Scanview Program
  • Stay Safe: A Win-Win for Solvent Storage
  • Streamline your sample processing
  • The importance of chemical composition for vial performance
  • Tips & Tricks for Amino Acid Analysis – Part I
  • Tips & Tricks for Amino Acid Analysis – Part II
  • Tips & Tricks for Amino Acid Analysis – Part III
  • Tips & Tricks for Amino Acid Analysis – Part IV
  • Tips for Smooth Sailing with HIC
  • Troubleshooting HPLC autosamplers
  • Troubleshooting HPLC degassers
  • Troubleshooting Sequence Coverage – Peptide Mapping Part III
  • UV, MS, TFA, and Formic Acid – What to use? Peptide Mapping Part II

Optimizing Bonding Chemistry for Sub-2 µm SEC Particles

In the previous two blog posts (Choosing the right pore size for SEC and Importance of silica particle strength for Sub-2 µm SEC columns), we discussed how the physical properties of the silica, such as pore size, pore volume, and particle strength, affect SEC performance. In this post, we would like to discuss the effect of bio-inertness of bonding chemistry on SEC performance.

 

The effect of silica particle properties usually seen in chromatograms is related to peak width and resolution. The effect of phase chemistry on the chromatogram is related to peak shape, protein adsorption, and retention time shift of small molecules. One good indication of whether the surface is bio-inert or not is to compare the retention time of uridine, a small neutral molecule, to that of vitamin B12, a small biomolecule. If the B12 retention time is longer than that of uridine, it usually indicates some surface silanol interaction.

 

 

 

 

For SEC, the characteristics of bonding phase should exhibit:

  1. Very hydrophilic – No hydrophobic adsorption of proteins nor nonspecific binding
  2. Full surface coverage – No surface silanol activity, no ion exchange interaction
  3. Stable phase – Long chemical lifetime, no retention time shift over time

For AdvanceBio SEC 200 Å 1.9 µm, we did bonding optimization to achieve the above goals:

  1. It is not diol phase. Many SEC column manufacturers use a diol phase. We have found that diol phase is not ideal for ADCs, as the peaks tail. We developed a new hydrophilic bonding chemistry using another hydrophilic group for our 2.7 µm AdvanceBio SEC column, and found it to give better peak shape. 1.9 µm AdvanceBio SEC is based on similar phase chemistry.
  2. The surface coverage is further optimized for 1.9 µm AdvanceBio SEC. Total coverage of the surface is good, but too much coverage will result in greater loss of pore volume. We optimized the bonding process to balance surface coverage and pore volume reduction after bonding so that we still have inert surface but maintain high pore volume after bonding.
  3. The phase is crosslinked to ensure chemical stability.

 

The figure below shows overlays of chromatograms over a lifetime test of 400 injections. There is no protein adsorption and no vitamin B12 shift over 400 injections. If the phase is unstable and changes during the test, B12 will usually start to shift to the right.

 

 

If the phase is too hydrophobic and has surface silanol activity, the large proteins will be adsorbed to surface, and the column has to be so-called “primed”, where the few first injections have to be made to saturate the surface activity before the column can be used. The figure below shows the first five injections of BioRad sample on a fresh column of AdvanceBio SEC 200 Å 1.9 µm and other competitor’s SEC 1.8 µm. There is no need for “priming” the AdvanceBio SEC 200 Å 1.9 µm column, while the competitor’s SEC 1.8 µm column had protein adsorption for the first couple injections and needed priming.

 

 

In summary, for BioSEC, bonding chemistry should provide:

  1. Hydrophilic and bio-inert surface for good peak shape and no nonspecific binding
  2. Stable phase for column lifetime

 

---

 

Thanks, Wu! I always feel more confident about evaluating and troubleshooting my chromatography when I have a better understanding of the “why” behind how my column works. Hopefully this series of posts from Wu help you to make more informed SEC column and method decisions, too.

Until next time -

Anne

 

Keywords: mAb, monoclonal antibody, ADC, antibody drug conjugates, SEC, size exclusion chromatography, biologics, biosimilars, insulin, molecular weight, aggregation, fragment, quality control, biopharmaceutical, CQA, critical quality attributes, pore size, AdvanceBio, column lifetime, silica, biomolecule, secondary interactions, non specific binding

  • mab
  • molecular weight
  • secondary interactions
  • adc
  • quality control
  • insulin
  • size exclusion chromatography
  • fragment
  • biosimilars
  • critical quality attributes
  • monoclonal antibody
  • column lifetime
  • antibody drug conjugates
  • aggregation
  • biomolecule
  • pore size
  • Biopharma LC Columns & Consumables
  • non specific binding
  • sec
  • silica
  • advancebio
  • biopharmaceutical
  • cqa
  • biologics
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help