Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Consumables
  • Technical Areas
  • More
Consumables
Wiki Pre-Columns - the forgotten art of using retention gaps
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • New
Consumables requires membership for participation - click to join
  • Wiki
  • A Beginner’s Guide to Hydrophobic Interaction Chromatography
  • A simple tip to protect your columns
  • A Tip for Preparing Robust and Consistent Mobile Phases
  • Achieve accurate and repeatable gas flow meter measurements
  • AdvanceBio Columns Blog Series
  • Agilent 123 Meter Sodium probe maintenance
  • Agilent Collection of Columns, Supplies, and Standards Resources
  • Analysis of microplastics in the environment
  • Analyzing ADCs by HIC
  • Avoiding downtime in the lab: top tips for GC/MS success
  • Best Practices for Aqueous Mobile Phases
  • Best Practices for Making Good Connections
  • Bursting Tubing and Columns (GC and HPLC)
  • Calculating Column Volume
  • Cannabis Potency Testing: a Reliable, Cost-Effective Method
  • Checking Your Intuition - Sub 2 µm vs Superficially Porous
  • Choosing the right GC Injection Technique
  • Choosing the right pore size for size exclusion chromatography
  • Columns, Supplies, and Standards Knowledgebase
  • Consumables Applications and Workflows
  • Consumables Recommended Supplies Lists for Agilent.com
  • Custom Product Request
  • Extreme Makeover – Derivatizations in Chromatography – Part 1 GC
  • Extreme Makeover – Derivatizations in Chromatography – Part 2 LC
  • Fake It Until You Make It: When BioInert Isn’t an Option
  • Flipping Amino Acid Analysis on Its Head
  • Glycans at a glance:  Analyzing therapeutic glycoproteins
  • Handle and Care of Syringes
  • Help! My Peaks Look Strange - Fronting and Tailing in GC
  • Help! My Peaks Look Strange - Saddle Points - LC/GC Troubleshooting
  • Hydrophobic Interaction Chromatography of Proteins and mAbs
  • Importance of Silica Particle Strength for Sub-2 µm SEC Columns
  • KB: Ferrules recommended for GC self-tightening column nut
  • LC Method Translation - the Dwell Volume
  • Minimize spectroscopy workflow disruptions
  • Minimizing Metals for Best HILIC Results
  • More than just a drink: Analyzing the elemental composition of beer
  • Multi-Attribute Methods – Peptide Mapping Part IV
  • Must See Webinars
  • Nomenclature of CFC's/Freons/Halons/Coolants
  • Oligonucleotide Analysis - Unexpected Details Matter
  • Optimizing Bonding Chemistry for Sub-2 µm SEC Particles
  • Pass the Salt, Please – Mobile Phase Preparation for HIC
  • Pesticides and their stability during GC analyses
  • Pre-Columns - the forgotten art of using retention gaps
  • Problematic polar analytes? Hello HILIC…
  • Protecting your laboratory productivity
  • Recommended Reading
  • Sample Prep Pointers - Peptide Mapping Part I
  • Save your results with sample filtration
  • Simplified cone inspection with the new Agilent LED measuring magnifier
  • Software tool for the ADM Flow Meter (G6691A)
  • Software-Supported Method Development - the Scanview Program
  • Stay Safe: A Win-Win for Solvent Storage
  • Streamline your sample processing
  • The importance of chemical composition for vial performance
  • Tips & Tricks for Amino Acid Analysis – Part I
  • Tips & Tricks for Amino Acid Analysis – Part II
  • Tips & Tricks for Amino Acid Analysis – Part III
  • Tips & Tricks for Amino Acid Analysis – Part IV
  • Tips for Smooth Sailing with HIC
  • Troubleshooting HPLC autosamplers
  • Troubleshooting HPLC degassers
  • Troubleshooting Sequence Coverage – Peptide Mapping Part III
  • UV, MS, TFA, and Formic Acid – What to use? Peptide Mapping Part II

Pre-Columns - the forgotten art of using retention gaps

Pre-columns, guard columns, retention gaps, DuraGuards, EZ-Guards - these are just a few of the names used in capillary GC to indicate a certain type of inlet section on the capillary column. From a technical point of view they are all the same. They consist of a small piece of fused silica, usually 1 - 5 meter in length, installed in front of the analytical column. The coupling with the analytical column is done using various types of connectors. In case of EZ-Guard columns the pre-column is an integrated part of the analytical column without the need for an extra coupling. All of these pre-columns are made of uncoated but deactivated fused silica tubing. More precisely they are almost uncoated. They are coated and deactivated using a thin layer of stationary phase. The film thickness of this layer is only a fraction of that to be found on the analytical column even when compared to the thinner film columns of 0.10 µm.

 

The difference between pre-columns

The main difference is only their naming, which is directly linked to their usage.

Retention gaps are used to introduce a liquid sample into the analytical column via (cold) on-column or splitless injection techniques. Retention gaps are essential to obtain small injection bands, narrow sample peaks and maintain column efficiency if these injection techniques are applied.

 

Pre-columns are referred to as guard columns when their main purpose is to protect the analytical column from contamination and it allows the injection of relatively „dirty samples“ on the column. Obviously pre-columns are often used both as retention gap and as guard column at the same time e.g. in case of on-column injections of dirty samples. Also the DuraGuard/EZ-Guard columns can be used either way.

 

 

Which pre-column polarity do I need?

As mentioned, the pre-column is coated and deactivated with a thin layer of stationary phase with a certain polarity. A coating with a 100 % methylpolysiloxane phase will result in a non-polar pre-column; polyethylene glycol phases in polar ones. The stationary phase used for deactivation of the guard column piece for EZ-guard columns is always identical to the stationary phase of the analytical column. In case of separately installed pre-columns the polarity of the analytical column can be different from the polarity of the pre-column. In practice this will rarely happen. It is the choice of sample solvent, which will dictate the choice of pre-column polarity.

 

Apolar SolventPolarityApolarMedium
Polar
Polar
Squalane-0.8✓
Isooctane-0.4✓
n-Decane-0.3✓
Pentane0.0✓
Cyclohexane0.0✓
Hexane0.0✓
Carbon disulfide1.0✓
Carbon tetrachloride1.7✓
Triethylamine1.8✓
Toluene2.3✓✓
p-Xylene2.4✓✓
Diethyl ether2.9✓✓

 

 

If the polarity-columns are checked the appropriate polarity-
retention gap can be used

Medium-polar SolventPolarityApolarMedium
Polar
Polar
Benzene3.0✓✓
n-Octanol3.2✓✓
Dichloromethane3.4✓✓
Dichloroethene3.7✓✓

tert.-Butanol

3.9✓✓
n-Propanol3.9✓✓
Methyl isobutyl keton4.2✓✓
Tetrahydrofuran4.2✓✓
Isopropanol4.3✓✓✓
Ethyl acetate4.3✓✓
Chloroform4.4✓✓
Butanone4.5✓✓
Cyclohexanone4.5✓✓
Methyl ethylketon4.5✓✓
Dioxane4.8✓✓
2-Picoline4.8✓✓
Polar SolventPolarityApolarMedium
Polar
Polar
Diethylene glycol5.0✓✓
Ethanol5.2✓✓
Pyridine5.3✓✓
Ethylene glycol5.4✓✓
Acetone5.4✓✓
Methoxyethanol5.7✓
Acetonitrile6.2✓
Dimethylformamide6.4✓
Dimethylsulfoxide6.5✓
N-Methylpyrrolidon6.5✓
HMPA6.6✓
Methanol6.6✓
Nitromethane6.8✓
Formamide7.3✓
Water9.0✓
Tetrafluoropropanol9.3✓

 

The table above indicates pre-column polarity and solvent compatibility. In general non-polar pre-columns should be combined with apolar solvents, polar solvents with polar pre-columns. There is however some overlap possible in the intermediate polar solvents. A fairly polar solvent like e.g. Isopropanol can be combined with an either apolar, polar or medium polar pre-columns. However, it is important to note that especially in cases where the pre-column is used as retention gap for on-column injections this compatibility issue is crucial. The liquid sample solvent should be evenly spread on the entire phase layer of the pre-column in order to obtain good sample focusing and prevent peak splitting. This can only occur when there is a good polarity match between solvent and pre-column. The nature of the analytes as such plays a minor role in the pre-column polarity choice.

 

Which pre-column length do I need?

The determination of the length of the pre-column depends on application, injection technique and some other factors. For those applications where it is used as guard column the length is not very critical. In most cases the interfering contaminants will be trapped in the first meter of the guard column. One meter is often enough. However, it is more practical to work with a longer length guard column, as is the case for EZ-guards columns. The front-end of the EZ-Guard can be cut off when it becomes too contaminated. Monitoring of the column efficiency and peak symmetry will indicate the moment to remove the first section of the guard column. Cutting off one meter will be sufficient in most cases.

 

Retention gap lengths are determined differently. During a splitless injection the retention gap usually functions as an extension of the injector liner. It prevents the injector from getting overloaded. As a result the volume of the retention gap for splitless is linked directly to the total vapor volume of the sample solvent. The analytes are „cold trapped“ in the stationary phase at the front of the analytical column. The length of the retention gap needed depends on injected volume, type of solvent, injector temperature, column head pressure and internal diameter of the retention gap. The table below shows some retention gaps lengths for 1 µl of solvents at 250 °C and 100 kPa (14.5 psi). It is clear that the 530 µm retention gaps are the most practical because of their shorter length. Water in particular has a large vapour volume and that has its effect on the retention gap length. In practice one installs 25 % extra retention gap length just to be on the safe side.

 

Our GC Calculator can help calculating those volumes.

  • gc columns
  • on-column
  • General
  • injections
  • retention gaps
  • Fittings
  • GC
  • solvent focussing
  • splitless
  • Tubing
  • Tools
  • pre-columns
  • and Supplies
  • Capillaries
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help