Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Cell Analysis
  • Technical Areas
Cell Analysis
Wiki Agilent Seahorse Wave Error exporting to GraphPad Prism: Troubleshooting Tips
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • Collection of Cell Analysis Resources
  • Cell Analysis FAQs and Troubleshooting
  • Agilent Cell Analysis Video Library
  • Cell Analysis
  • Agilent MitoXpress Intra Intracellular Oxygen Assay FAQs
  • Agilent MitoXpress Xtra and pH-Xtra Assay Data Visualization Tool User Guide FAQs
  • Agilent MitoXpress Xtra Oxygen Consumption Assay (HS Method) FAQs
  • Agilent pH-Xtra Glycolysis Assay FAQs
  • Agilent Seahorse Calibration Error Troubleshooting Tips
  • Agilent Seahorse Error Message: Door Not Closing Troubleshooting Tips
  • Agilent Seahorse Error on Wave desktop: “Instrument not installed”: Troubleshooting Tips
  • Agilent Seahorse Error Unloading Cartridge Troubleshooting Tips
  • Agilent Seahorse File Save Error: “File cannot be saved”: Troubleshooting Tips
  • Agilent Seahorse Group Name(s) from Wave Don’t Match Excel: Troubleshooting Tips
  • Agilent Seahorse Licensing Issue with Software: Troubleshooting Tips
  • Agilent Seahorse Mito Fuel Flex Report Generator FAQs
  • Agilent Seahorse Mito Stress Test Single-File and Mulit-File XF Report Generator FAQs
  • Agilent Seahorse Motor error/plate tray immovable: Troubleshooting Tips
  • Agilent Seahorse Plate error: Plate Not Compatible with Assay: Troubleshooting Tips
  • Agilent Seahorse Plate error: “This cell plate is not assigned to this sensor cartridge.  Please replace the cell plate and try again”: Troubleshooting Tips
  • Agilent Seahorse Prism Error when opening .Pzfx file exported from Wave Software: Troubleshooting Tips
  • Agilent Seahorse Wave Crashes When “Design” Clicked: Troubleshooting Tips
  • Agilent Seahorse Wave Error exporting to GraphPad Prism: Troubleshooting Tips
  • Agilent Seahorse Wave File Save Error: “We cannot run this experiment.  Some groups have the same name”: Troubleshooting Tips
  • Agilent Seahorse Wave Hypoxia Data Does Not Recalculate after Excluding and then Reincluding the Sulfite Group: Troubleshooting Tips
  • Agilent Seahorse Wave Software Crash or Freeze: Troubleshooting Tips
  • Agilent Seahorse XF Cell Energy Phenotype Test Kit FAQs
  • Agilent Seahorse XF Cell Energy Phenotype Test Report Generator FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Kit FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Single-File and Multi-File Report Generator FAQs
  • Agilent Seahorse XF Imaging and Normalization FAQs
  • Agilent Seahorse XF Mito Fuel Flex Test Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Report Generator FAQs
  • Agilent Seahorse XFp: Cartridge failed to eject after assay: Troubleshooting Tips
  • Agilent Seahorse XFp: Unable to set up wireless connection with adapter: Troubleshooting Tips
  • Agilent Seahorse: Unable to open Wave Software/data files, Unable to save data files/ Unable to view all assay templates: Troubleshooting Tips
  • Seahorse Barcode-Related Error: Troubleshooting Tips
Still Need Help?

Post your question in our User Forum or Contact Support.

Agilent Seahorse Wave Error exporting to GraphPad Prism: Troubleshooting Tips

Created by Courtney Watts Courtney Watts over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh 9 months ago

 

If the Wave software used with Seahorse data displays an error stating: “The file cannot be opened since it is damaged or truncated”, this document can be used to help troubleshoot.

 

Here is a list of the most common causes for this error:

  • No background correction groups assigned. For any Seahorse assay, background wells are critical for acquiring reliable data.  Background correction wells are assigned to the plate map by default.  Although it is possible to modify the default background wells, at least one well must be assigned to the background correction group

 

Try these steps for a possible resolution:

  • If assigning at least one background correction well does not remedy the issue, contact Agilent Cell Analysis technical support at:
    • Email contact:
      • cellanalysis.support@agilent.com
    • Phone Contact:
      • US Contact: 1 800 227 9770 option 3, 8
      • UK Contact: 0 800 096 7632 option 3
      • DE Contact: 0 800 180 66 78 option 3
      • NL Contact: 0 800 022 7243 option 3
      • Other European Countries Contact: 45 3136 9878 option 3

 

More Troubleshooting Tips: Cell Analysis FAQs and Troubleshooting

 

Keywords: Prism error, Prism, GraphPad, GraphPad Error, Export Error, Seahorse, XFe96, XFe24, XFp, XF24, XF96, Troubleshooting

  • XFe96
  • Troubleshooting
  • XFe24
  • prism
  • Seahorse
  • graphpad
  • XFp
  • xf24
  • graphpad error
  • export error
  • xf96
  • prism error
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help