Wiki Agilent Seahorse XF Glycolytic Rate Assay Single-File and Multi-File Report Generator FAQs
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • New
Cell Analysis requires membership for participation - click to join
  • Wiki
  • Agilent Cell Analysis Video Library
  • Agilent MitoXpress Intra Intracellular Oxygen Assay FAQs
  • Agilent MitoXpress Xtra and pH-Xtra Assay Data Visualization Tool User Guide FAQs
  • Agilent MitoXpress Xtra Oxygen Consumption Assay (HS Method) FAQs
  • Agilent pH-Xtra Glycolysis Assay FAQs
  • Agilent Seahorse Calibration Error Troubleshooting Tips
  • Agilent Seahorse Error Message: Door Not Closing Troubleshooting Tips
  • Agilent Seahorse Error on Wave desktop: “Instrument not installed”: Troubleshooting Tips
  • Agilent Seahorse Error Unloading Cartridge Troubleshooting Tips
  • Agilent Seahorse File Save Error: “File cannot be saved”: Troubleshooting Tips
  • Agilent Seahorse Group Name(s) from Wave Don’t Match Excel: Troubleshooting Tips
  • Agilent Seahorse Licensing Issue with Software: Troubleshooting Tips
  • Agilent Seahorse Mito Fuel Flex Report Generator FAQs
  • Agilent Seahorse Mito Stress Test Single-File and Mulit-File XF Report Generator FAQs
  • Agilent Seahorse Motor error/plate tray immovable: Troubleshooting Tips
  • Agilent Seahorse Plate error: Plate Not Compatible with Assay: Troubleshooting Tips
  • Agilent Seahorse Plate error: “This cell plate is not assigned to this sensor cartridge.  Please replace the cell plate and try again”: Troubleshooting Tips
  • Agilent Seahorse Prism Error when opening .Pzfx file exported from Wave Software: Troubleshooting Tips
  • Agilent Seahorse Wave Crashes When “Design” Clicked: Troubleshooting Tips
  • Agilent Seahorse Wave Error exporting to GraphPad Prism: Troubleshooting Tips
  • Agilent Seahorse Wave File Save Error: “We cannot run this experiment.  Some groups have the same name”: Troubleshooting Tips
  • Agilent Seahorse Wave Hypoxia Data Does Not Recalculate after Excluding and then Reincluding the Sulfite Group: Troubleshooting Tips
  • Agilent Seahorse Wave Software Crash or Freeze: Troubleshooting Tips
  • Agilent Seahorse XF Cell Energy Phenotype Test Kit FAQs
  • Agilent Seahorse XF Cell Energy Phenotype Test Report Generator FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Kit FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Single-File and Multi-File Report Generator FAQs
  • Agilent Seahorse XF Imaging and Normalization FAQs
  • Agilent Seahorse XF Mito Fuel Flex Test Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Report Generator FAQs
  • Agilent Seahorse XFp: Cartridge failed to eject after assay: Troubleshooting Tips
  • Agilent Seahorse XFp: Unable to set up wireless connection with adapter: Troubleshooting Tips
  • Agilent Seahorse: Unable to open Wave Software/data files, Unable to save data files/ Unable to view all assay templates: Troubleshooting Tips
  • Cell Analysis FAQs and Troubleshooting
  • Collection of Cell Analysis Resources
  • Seahorse Barcode-Related Error: Troubleshooting Tips

Agilent Seahorse XF Glycolytic Rate Assay Single-File and Multi-File Report Generator FAQs

What rate measurements are used to calculate the parameters in the Report Generator?

Parameter equations are described in Table 1 on page 6 and Table 2 on page 17 of the Single File RG user guide.

Parameter equations are described in Table 1 on page 6 and Table 2 on page 21 of the Multi-File RG User Guide.

 

Incompatible elements with the XF Glycolytic Rate Assay The following items are not supported with the XF Glycolytic Rate Assay:

  • Sensor cartridges for the XF24 and XF24-3 Analyzer.
  • Spheroid, Islet, and V28 plate types.

 

How do I remove outlier wells in the Report Generator?

Outlier assay wells must be turned OFF or reassigned in Wave 2.4 before Excel export. See “Exclude Outliers/Groups from Analysis” on page 23 or the Wave user guide for more information.

 

I need to change Buffer Factor for a group, where can I find this information?

Buffer Factor values are automatically imported from each Excel File per group. Modifying the Buffer Factor in the Multi-File XF Report Generator is not supported presently. To change the Buffer Factor value for a group, open the assay result file in Wave 2.4 software, change the Buffer Factor value of the assay media and the background wells, and re-export the data file to Excel.

 

Can I use the Excel file exported from the XFp, XF96 Analyzer, or an earlier version of Wave software?

This Multi-File XF Report Generator supports import and analysis of Excel files exported from Wave 2.4 (Desktop or Controller) or later. Excel files exported from the XFp Analyzer, XF96 Analyzer, XF Reader software, or earlier versions of Wave software are not supported.

 

If you receive an error message about Instrument Protocol (XFe96; XFe24; XF96 only)

Custom Cycles are not part of the standardized assay template for the XF Glycolytic Rate Assay and not supported in Report Generator analysis. A Custom Cycle refers to an axtra 'Mix' or 'Wait' command step in the Instrument Protocol of an assay. Please contact Agilent Seahorse Technical Support if you have any additional questions regarding Custom Cycles.

 

Can I use baseline rate data (%) to calculate assay parameters?

Normalized (or nonnormalized) ECAR data is used for parameter calculations, parameter calculations using baseline rate data (%) is not supported currently.

 

Will I see Warning message(s) when the % PER from Glycolysis is less than 50% and less than 10%

When the % of PER from Glycolysis for a selected group is less than 50% or less than 10%, the Report Generator will display a warning message in the legend on the Summary Printout tab (below the Normalize button). For more information, see “Advanced Options” on page 19.

 

More FAQs: Cell Analysis FAQs and Troubleshooting

 

Keywords: Glycolytic Rate Assay, phenol-red free, Buffer Factor, HEPES, Report Generator, Warning Message, % PER, Report Generator Error, Outlier, Parameter Equations, Seahorse, XFe96, XFe24, XFp, FAQ

  • XFe96
  • % per
  • warning message
  • XFe24
  • outlier
  • hepes
  • Seahorse
  • XFp
  • glycolytic rate assay
  • parameter equations
  • phenol-red free
  • report generator error
  • buffer factor
  • report generator
  • faq
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help