Wiki Agilent Seahorse Calibration Error Troubleshooting Tips
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • New
Cell Analysis requires membership for participation - click to join
  • Wiki
  • Agilent Cell Analysis Video Library
  • Agilent MitoXpress Intra Intracellular Oxygen Assay FAQs
  • Agilent MitoXpress Xtra and pH-Xtra Assay Data Visualization Tool User Guide FAQs
  • Agilent MitoXpress Xtra Oxygen Consumption Assay (HS Method) FAQs
  • Agilent pH-Xtra Glycolysis Assay FAQs
  • Agilent Seahorse Calibration Error Troubleshooting Tips
  • Agilent Seahorse Error Message: Door Not Closing Troubleshooting Tips
  • Agilent Seahorse Error on Wave desktop: “Instrument not installed”: Troubleshooting Tips
  • Agilent Seahorse Error Unloading Cartridge Troubleshooting Tips
  • Agilent Seahorse File Save Error: “File cannot be saved”: Troubleshooting Tips
  • Agilent Seahorse Group Name(s) from Wave Don’t Match Excel: Troubleshooting Tips
  • Agilent Seahorse Licensing Issue with Software: Troubleshooting Tips
  • Agilent Seahorse Mito Fuel Flex Report Generator FAQs
  • Agilent Seahorse Mito Stress Test Single-File and Mulit-File XF Report Generator FAQs
  • Agilent Seahorse Motor error/plate tray immovable: Troubleshooting Tips
  • Agilent Seahorse Plate error: Plate Not Compatible with Assay: Troubleshooting Tips
  • Agilent Seahorse Plate error: “This cell plate is not assigned to this sensor cartridge.  Please replace the cell plate and try again”: Troubleshooting Tips
  • Agilent Seahorse Prism Error when opening .Pzfx file exported from Wave Software: Troubleshooting Tips
  • Agilent Seahorse Wave Crashes When “Design” Clicked: Troubleshooting Tips
  • Agilent Seahorse Wave Error exporting to GraphPad Prism: Troubleshooting Tips
  • Agilent Seahorse Wave File Save Error: “We cannot run this experiment.  Some groups have the same name”: Troubleshooting Tips
  • Agilent Seahorse Wave Hypoxia Data Does Not Recalculate after Excluding and then Reincluding the Sulfite Group: Troubleshooting Tips
  • Agilent Seahorse Wave Software Crash or Freeze: Troubleshooting Tips
  • Agilent Seahorse XF Cell Energy Phenotype Test Kit FAQs
  • Agilent Seahorse XF Cell Energy Phenotype Test Report Generator FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Kit FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Single-File and Multi-File Report Generator FAQs
  • Agilent Seahorse XF Imaging and Normalization FAQs
  • Agilent Seahorse XF Mito Fuel Flex Test Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Report Generator FAQs
  • Agilent Seahorse XFp: Cartridge failed to eject after assay: Troubleshooting Tips
  • Agilent Seahorse XFp: Unable to set up wireless connection with adapter: Troubleshooting Tips
  • Agilent Seahorse: Unable to open Wave Software/data files, Unable to save data files/ Unable to view all assay templates: Troubleshooting Tips
  • Cell Analysis FAQs and Troubleshooting
  • Collection of Cell Analysis Resources
  • Seahorse Barcode-Related Error: Troubleshooting Tips

Agilent Seahorse Calibration Error Troubleshooting Tips

If the software used with Seahorse analyzers displays a calibration error, this document can be used to help troubleshoot.

 

Here is a list of the most common causes for a calibration-related issue:

  • Inadequate hydration time
  • Passed expiration date
  • Overexposure of cartridge to light
  • Fluorophore adhesion issue
  • Previous cartridge left loaded on probes (either after assay or after shipping)
  • Not running the most recent version of software
  • A hardware component could have become damaged

 

 

 

Try these steps for a possible resolution:

  • Ensure the steps outlined in the XF Sensor Cartridge hydration protocols (XFp, XF24, XFe24, XF96 and XFe96) are being followed. The XF Sensor Cartridge must be hydrated for at least 4 hours before your XF assay. Overnight hydration is preferred.
  • Check expiration date and lot number – see photo above for example of expiration date location.
  • Ensure XF Sensor Cartridge is not exposed to light; for example leaving the cartridge on the lab bench or under UV light in hood for long periods of time. Cartridge should be left in plastic shipping case until you are ready for overnight hydration – see photo above for example of cartridge tub.
  • Turn over XF Sensor Cartridge and view fluorophore spots are not rippled or decaying.
  • Open instrument cover to ensure that the metal probes can be seen (rather than a green cartridge)
  • Verify that the most recent version of software is installed on controller
  • If issue is still occurring after all of the above has been verified, contact Agilent Cell Analysis technical support at:
    • Email contact:
      • cellanalysis.support@agilent.com
    • Phone Contact:
      • US Contact: 1 800 227 9770 option 3, 8
      • UK Contact: 0 800 096 7632 option 3
      • DE Contact: 0 800 180 66 78 option 3
      • NL Contact: 0 800 022 7243 option 3
      • Other European Countries Contact: 45 3136 9878 option 3

       

    More Troubleshooting Tips: Cell Analysis FAQs and Troubleshooting

     

    Keywords: Calibration error, calibration, calibration issue, calibration failure, pH calibration failure, O2 calibration failure, Seahorse, XFe96, XFe24, XFp, XF24, XF96, Troubleshooting

    • XFe96
    • troubleshooting
    • XFe24
    • Seahorse
    • XFp
    • calibration failure
    • calibration issue
    • xf24
    • calibration error
    • calibration
    • xf96
    • o2 calibration failure
    • ph calibration failure
    • Share
    • History
    • More
    • Cancel
    Anonymous
    Related

    Agilent Community Feedback

    Agilent Community Feedback

    ×
    We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
    Submit Cancel
    Submit Cancel
    Recommended
    Privacy Statement
    Terms of Use
    Contact Us
    Site Help