Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Cell Analysis
  • Technical Areas
Cell Analysis
Wiki Agilent Seahorse XF Imaging and Normalization FAQs
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • Collection of Cell Analysis Resources
  • Cell Analysis FAQs and Troubleshooting
  • Agilent Cell Analysis Video Library
  • Cell Analysis
  • Agilent MitoXpress Intra Intracellular Oxygen Assay FAQs
  • Agilent MitoXpress Xtra and pH-Xtra Assay Data Visualization Tool User Guide FAQs
  • Agilent MitoXpress Xtra Oxygen Consumption Assay (HS Method) FAQs
  • Agilent pH-Xtra Glycolysis Assay FAQs
  • Agilent Seahorse Calibration Error Troubleshooting Tips
  • Agilent Seahorse Error Message: Door Not Closing Troubleshooting Tips
  • Agilent Seahorse Error on Wave desktop: “Instrument not installed”: Troubleshooting Tips
  • Agilent Seahorse Error Unloading Cartridge Troubleshooting Tips
  • Agilent Seahorse File Save Error: “File cannot be saved”: Troubleshooting Tips
  • Agilent Seahorse Group Name(s) from Wave Don’t Match Excel: Troubleshooting Tips
  • Agilent Seahorse Licensing Issue with Software: Troubleshooting Tips
  • Agilent Seahorse Mito Fuel Flex Report Generator FAQs
  • Agilent Seahorse Mito Stress Test Single-File and Mulit-File XF Report Generator FAQs
  • Agilent Seahorse Motor error/plate tray immovable: Troubleshooting Tips
  • Agilent Seahorse Plate error: Plate Not Compatible with Assay: Troubleshooting Tips
  • Agilent Seahorse Plate error: “This cell plate is not assigned to this sensor cartridge.  Please replace the cell plate and try again”: Troubleshooting Tips
  • Agilent Seahorse Prism Error when opening .Pzfx file exported from Wave Software: Troubleshooting Tips
  • Agilent Seahorse Wave Crashes When “Design” Clicked: Troubleshooting Tips
  • Agilent Seahorse Wave Error exporting to GraphPad Prism: Troubleshooting Tips
  • Agilent Seahorse Wave File Save Error: “We cannot run this experiment.  Some groups have the same name”: Troubleshooting Tips
  • Agilent Seahorse Wave Hypoxia Data Does Not Recalculate after Excluding and then Reincluding the Sulfite Group: Troubleshooting Tips
  • Agilent Seahorse Wave Software Crash or Freeze: Troubleshooting Tips
  • Agilent Seahorse XF Cell Energy Phenotype Test Kit FAQs
  • Agilent Seahorse XF Cell Energy Phenotype Test Report Generator FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Kit FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Single-File and Multi-File Report Generator FAQs
  • Agilent Seahorse XF Imaging and Normalization FAQs
  • Agilent Seahorse XF Mito Fuel Flex Test Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Report Generator FAQs
  • Agilent Seahorse XFp: Cartridge failed to eject after assay: Troubleshooting Tips
  • Agilent Seahorse XFp: Unable to set up wireless connection with adapter: Troubleshooting Tips
  • Agilent Seahorse: Unable to open Wave Software/data files, Unable to save data files/ Unable to view all assay templates: Troubleshooting Tips
  • Seahorse Barcode-Related Error: Troubleshooting Tips
Still Need Help?

Post your question in our User Forum or Contact Support.

Agilent Seahorse XF Imaging and Normalization FAQs

Created by Courtney Watts Courtney Watts over 3 years ago | Last modified by kristen.sensabaugh kristen.sensabaugh over 1 year ago

What additional steps (other than FSE/CPS visit) are necessary to set up the Cytation/XFe/Controller system?

When BioTek’s Gen5 is set up, the image library must be on either an external hard drive or a network location. XF Imaging and Cell Counting software (ICC) and Wave will use a database that is automatically generated. Upon installation to link images to cell plates – the database location should not require attention, but can be changed using the help menus in both Applications.

 

Can I use ICC and Wave at the same time? How about Gen5 and Wave?

The ICC software utilizes Gen5 to run image collection protocols. Gen5 will be running while ICC is open, but without an open window and as a background process. Wave can be used at the same time. Gen5 cannot be opened in addition to ICC software. Also, you may need to navigate to the task manager and close the background instance of Gen5 after closing ICC, before you can open Gen5 separately.

 

How can I ensure best image quality?

It is highly recommended that users wipe the bottom of their cell plates with isopropyl/ethanol before imaging. Debris will affect autofocus and image quality.

 

Can I adjust the images that have been collected by ICC?

Zoom, brightness, and contrast can be adjusted during image review both in ICC and in Wave. These adjustments will not change the raw pixel information.

 

Can I adjust the image collection protocols for ICC?

No, not currently – the image protocols have been optimized for XF plates and assays. Custom protocols and cell counts can be created and run in Gen5 software. If a custom cell counting protocol is created in Gen5, the results can be copied and pasted into Wave’s normalization mode. The imagery will only be imported in Wave when ICC is used to acquire them.

 

Well View is not available for my Wave results, but I collected images. How can

I import them?

The images must be collected using the ICC software (not Gen5). Also, confirm that Wave is using the same image database as ICC. If it needs to be changed, you may need to restart the software. If problems persist, contact Agilent Technical Support.

 

How are cell counts generated from only the center field of view?

Using only the center image field of view, the cell count per well is calculated based on the area of the well upon importing into the Wave Normalization tab.

 

Can I use a different fluorescent dye for cell counting?

In ICC, only dyes compatible with the DAPI filter set may be used.

 

How can I confirm cell count accuracy?

Upon completion of the fluorescence image acquisition, you will be able to review the images. Currently, there is a check box for viewing the cell counting “mask.” The mask highlights all counted objects within the image. The count is only for the field of view shown. Upon importing cell counts into Wave, an area factor is applied to generate a cell count for the entire well.

 

My cells are not consistently stained with Hoechst dye. What can I do?

We recommend optimizing the concentration of Hoechst dye – just like you would do for cell density or FCCP concentration. We have used a 20 μM dilution successfully in all validation testing.

 

Which versions of Cytation are compatible?

Only the Cytation 1 and  5 are compatible with the XF Imaging and Cell Counting software.

 

How can I export the full resolution images?

Full resolution images can be exported from the XF Imaging and Cell Counting software interface. Navigate to the plate menu and either export an entire plate of images, or open to the review screen and select specific wells to export. Also, “masked” images may only be exported when they are being reviewed – upon completion of fluorescence image acquisition.

 

Where do I find the User Manual?

The XF Imaging and Cell Counting Software User Manual can be found on the

Agilent website, here: https://www.agilent.com/cs/library/usermanuals/public/S7810AA_Seahorse_XF_Imaging_and_Cell_Counting_User_Manual.pdf 

 

Do I need to set up my cell plate in a particular manner?

ICC software requires two specific wells to be seeded with cells to accurately perform initialization; this process includes preliminary autofocus, auto-exposure, and white balance steps. Ensure wells D5 and E8 are seeded in XF96 plates and wells B3 and C4 in XF24 plates.

 

How many image sets can I import into Wave?

When importing into Wave, you can import a maximum of four image sets.

 

More FAQs: Cell Analysis FAQs and Troubleshooting

 

Keywords: Normalization, Cell Imaging, Cytation, Hoechst, ICC, Gen5, Seahorse, XFe96, XFe24, FAQ

  • XFe96
  • cytation
  • XFe24
  • Seahorse
  • gen5
  • hoechst
  • normalization
  • Normalization Solution
  • cell imaging
  • icc
  • faq
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help