Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Cell Analysis
  • Technical Areas
  • More
Cell Analysis
Wiki Agilent Seahorse XF Mito Fuel Flex Test Kit FAQs
  • Announcements
  • Forum
  • Files
  • Wiki
  • More
  • Cancel
  • New
Cell Analysis requires membership for participation - click to join
  • Wiki
  • Agilent Cell Analysis Video Library
  • Agilent MitoXpress Intra Intracellular Oxygen Assay FAQs
  • Agilent MitoXpress Xtra and pH-Xtra Assay Data Visualization Tool User Guide FAQs
  • Agilent MitoXpress Xtra Oxygen Consumption Assay (HS Method) FAQs
  • Agilent pH-Xtra Glycolysis Assay FAQs
  • Agilent Seahorse Calibration Error Troubleshooting Tips
  • Agilent Seahorse Error Message: Door Not Closing Troubleshooting Tips
  • Agilent Seahorse Error on Wave desktop: “Instrument not installed”: Troubleshooting Tips
  • Agilent Seahorse Error Unloading Cartridge Troubleshooting Tips
  • Agilent Seahorse File Save Error: “File cannot be saved”: Troubleshooting Tips
  • Agilent Seahorse Group Name(s) from Wave Don’t Match Excel: Troubleshooting Tips
  • Agilent Seahorse Licensing Issue with Software: Troubleshooting Tips
  • Agilent Seahorse Mito Fuel Flex Report Generator FAQs
  • Agilent Seahorse Mito Stress Test Single-File and Mulit-File XF Report Generator FAQs
  • Agilent Seahorse Motor error/plate tray immovable: Troubleshooting Tips
  • Agilent Seahorse Plate error: Plate Not Compatible with Assay: Troubleshooting Tips
  • Agilent Seahorse Plate error: “This cell plate is not assigned to this sensor cartridge.  Please replace the cell plate and try again”: Troubleshooting Tips
  • Agilent Seahorse Prism Error when opening .Pzfx file exported from Wave Software: Troubleshooting Tips
  • Agilent Seahorse Wave Crashes When “Design” Clicked: Troubleshooting Tips
  • Agilent Seahorse Wave Error exporting to GraphPad Prism: Troubleshooting Tips
  • Agilent Seahorse Wave File Save Error: “We cannot run this experiment.  Some groups have the same name”: Troubleshooting Tips
  • Agilent Seahorse Wave Hypoxia Data Does Not Recalculate after Excluding and then Reincluding the Sulfite Group: Troubleshooting Tips
  • Agilent Seahorse Wave Software Crash or Freeze: Troubleshooting Tips
  • Agilent Seahorse XF Cell Energy Phenotype Test Kit FAQs
  • Agilent Seahorse XF Cell Energy Phenotype Test Report Generator FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Kit FAQs
  • Agilent Seahorse XF Glycolytic Rate Assay Single-File and Multi-File Report Generator FAQs
  • Agilent Seahorse XF Imaging and Normalization FAQs
  • Agilent Seahorse XF Mito Fuel Flex Test Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Kit FAQs
  • Agilent Seahorse XF Real-Time ATP Rate Assay Report Generator FAQs
  • Agilent Seahorse XFp: Cartridge failed to eject after assay: Troubleshooting Tips
  • Agilent Seahorse XFp: Unable to set up wireless connection with adapter: Troubleshooting Tips
  • Agilent Seahorse: Unable to open Wave Software/data files, Unable to save data files/ Unable to view all assay templates: Troubleshooting Tips
  • Cell Analysis FAQs and Troubleshooting
  • Collection of Cell Analysis Resources
  • Seahorse Barcode-Related Error: Troubleshooting Tips

Agilent Seahorse XF Mito Fuel Flex Test Kit FAQs

What if all three inhibitors only cause a small decrease in total OCR?

Processes other than oxidation of these three fuels may contribute to baseline OCR. These processes may be broken down further into mitochondrial and nonmitochondrial oxygen consuming processes: Other mitochondrial respiration: respiration dependent on an alternative substrate(s) being oxidized to support mitochondrial respiration. This may include (but not limited to) short and medium chain fatty acids and amino acids other than glutamine. Nonmitochondrial oxygen consumption: consumption of oxygen by other biochemical processes in the cell. This includes (but is not limited to) very long chain fatty acids that get partially oxidized in the peroxisomes and other cellular enzymatic processes that consume oxygen. The nonmitochondrial fraction of total oxygen consumption can be measured using the Seahorse XF Cell Mito Stress Test.

 

Why is Dependency reported as zero?

If Dependency is not significantly above zero or negative due to well to well variability, there is no dependency on that particular substrate. If the cells are not dependent on the target fuel pathway, OCR may slightly increase following injection of inhibitor. When this occurs, Dependency is automatically set to zero (no dependence), and Flexibility will be equal to Capacity.

 

If I have negative flexibility values, what does it mean?

When changes in OCR are small, well to well variability might lead to negative flexibility values. Negative flexibility values of less than 5 % are generally attributable to noise in the assay. If you detect significant negative flexibility, contact Technical Support.

 

How can I further diagnose or troubleshoot Mito Fuel Flex Test results?

The preceding potential issues described (apparent low response to inhibitors, 0 % dependency, or negative flexibility values) may be further diagnosed by performing (as a separate test or added group) a Mito Fuel Flex Test with only media injections (no inhibitors) to establish if baseline respiration changes significantly over the course of the assay. If the absolute respiration rates (OCR) in the absence of inhibitors trend significantly upward or downward throughout the assay, then the parameters of this test (relative % dependency, capacity and flexibility) will be either underestimated or exaggerated, respectively. This depends on the magnitude and direction of any baseline trends versus the magnitude of change due to added inhibitors. To limit any upward/downward baseline trending OCR, ensure that cell culture and technical parameters of the assay have been thoroughly optimized.

 

Will these inhibitors and concentrations work with all cells?

Yes, the test uses all three compounds at concentrations well above their EC50 values for inhibition in mammalian cells. These values have been validated in many cell lines and primary isolates. While most cell types or cell lines have an appreciable response to at least one inhibitor, not all cells will respond to all inhibitors. If the cells are not responsive to a particular inhibitor, they may not depend on that particular fuel pathway (that is, they are flexible regarding the fuel used for oxidative phosphorylation).

 

How do I interpret ECAR and glycolysis in this assay?

Using combinations of inhibitors can confound interpretation of ECAR data with this test due to shifts in cellular ATP production and demand. For directly measuring glycolytic function, we recommend using the Seahorse XF Glycolysis Stress Test.

 

The recommended assay medium does not include fatty acid, can I add it?

Although not required, long chain fatty acid may be added to the medium. We recommend using a single species of long chain fatty acid, such as Seahorse XF palmitate-BSA FAO substrate, when testing exogenous fatty acid oxidation. NOTE: only oxidation of long-chain fatty acid, such as palmitate, is sensitive to inhibition by etomoxir.

 

More FAQs: Cell Analysis FAQs and Troubleshooting

 

Keywords: Mito Fuel Flex, Dependency, Capacity, Flexibility, Fatty Acid, Negative Flexibility, Seahorse, XFe96, XFe24, XFp, XF24, XF96, FAQ

  • XFe96
  • fatty acid
  • XFe24
  • capacity
  • mito fuel flex
  • Seahorse
  • flexibility
  • XFp
  • xf24
  • xf96
  • negative flexibility
  • faq
  • dependency
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help