When the tuning is done at 3200 (+) everything is fine. When I change it to tune at 1700 (+) the masses are reasonably accurate. After it gets to the third mass (620) it appears that the entire spectrum shifts by about 1.5 AMU and the tuning fails. There…
Hi Siddhartha,
Thank you for contacting Cell Analysis Technical Support. Can you please go ahead and send us your System Files so that we can see what might be going on? These can be accessed in the Help Section of the Wave software. You can select…
Hello thank you very much for your response, thanks for the firmware data, the degasser worked perfect, the firmware change no longer, and I can not find the previous version. and are from the 6.80, but how is it possible that a team of thousands of dollars…
Thanks for the reply. I am monitoring at 254 nm. Higher wavelengths do not show that drift. It must be related to the THF (I am using unstabilized) but I do not see how.
Hello, we have an mp aes 4210, we cannot turn it on and we get this error "PLASMA_IGNITION_SHER_GAS_PRESSURE. POP gas pressure out of range". We performed the gas test but it does not pass.
We contacted an Agilent agent and he told us that the problem…
Thanks for the info! What I was thinking about is that some instruments, like GC/MS units sometimes come with recommendations to not put the unit on a UPS due to problems it can cause with the oven. So, I was wondering if Agilent had anything specific…
This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems
Issue
The filaments are replaced as part of routine maintenance, or when they break (zero output current).
Items Required
Gloves, nylon,...
This Information Applies To: Agilent 5977A/B/C GC/MSD
Issue
The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source.
Items Required
Electron multiplier horn (5977B/C:...
This Information Applies To: Agilent 5973/ 5975/5977 GC/MSD with rotary vane foreline pump
Issue:
The oil mist filter is recommended to be replaced annually. If the color of the filter has changed or there is smell from the filter, the oil mist...
This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments...
This Information Applies To: Agilent G7129A/B/C Vialsampler
Issue:
Replacing the needle and needle seat of the Vialsampler is required when:
The needle is visibly damaged.
Leaks or blockages in the vialsampler are observed, especially between...
This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps
Issue:
How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.
...
This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due...
This Information Applies To: Agilent G1312X and G7112X binary pumps.
Issue:
A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages...
This Information Applies To: Agilent 1100, 1200, 1260 LC Pumps.
Issue:
The following procedure allows the cleaning of residue from the pump's check valves, often caused by polymerization of aged acetonitrile or salt-based solutions that may prevent...
This Information Applies To: Agilent LC Systems
Issue
After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to...