This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due to the flow cell follow the procedure below.
Steps to follow
Before proceeding:
Caution: If the column is left in the flow path, it may be damaged during the flushing procedure.
|
Steps to follow for reversed phase application:
Confirmation: If the clog was removed, and the pressure should be stable and no longer rising. You can proceed with the system flush, otherwise contact Agilent Technical Support.
|
Steps to follow for normal phase application:
Confirmation: If the clog was removed, and the pressure should be stable and no longer rising. You can proceed with the system flush, otherwise contact Agilent Technical Support.
|
Learn how to effectively troubleshoot your Agilent UV Detector:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent education
|