Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Pumps
  • Product Name :  1100 Series LC System 1120 Compact LC 1200 Series LC System 1220 infinity II LC System 1260 Infinity II Bio-Inert LC System 1260 Infinity II LC System 1260 Infinity II Prime Bio LC System 1260 Infinity LC System 1260 Infinity II Prime LC System 1290 Infinity II Bio LC System 1290 Infinity II LC System 1290 Infinity LC System
  • Task :  Troubleshooting
Related Articles
  • How to Troubleshoot Low Pressure or No Flow/Solvent Coming Out of 1100/1200 Series Isocratic/Quaternary/Binary Pumps

    This Information Applies To: Agilent 1100/1200 Series G1310A Isocratic Pump, G1311A Quaternary Pump, and G1312A/B Binary Pump modules Issue This article shows different options to troubleshoot low pressure or no flow issues in 1100/1200 series Isocratic/Quaternary/Binary pumps. Items Required Open-end wrench 14 mm (p/n 8710–1924 ) Open-end wrench 1/4 - 5/16 inch (p/n 8710–0510 ) Pair of Tweezers Steps to follow The following steps will guide you to find out which part is causing...
  • Modifying Solvent Compressibility for LC Pumps in OpenLab CDS

    This Information Applies To: Agilent OpenLab CDS, G7110B Isocratic Pump, G7111A/B, G5654A Quaternary Pumps, and G7112B Binary Pump. Issue Incorrectly applied solvent compressibilities can result in an increased pressure ripple, or inaccurate solvent composition in LC pumps. Background In HPLC, liquids under high pressures are pumped through the system, which can result in measurable liquid compression. Though the amount of compression is very small, it is enough to change the flow rate...
  • How to Change the Active Inlet Valve (AIV) and cartridge on a 1260 Isocratic, Binary or Quaternary Pump

    This Information Applies To: Agilent Isocratic, Binary and Quaternary Pumps with AIV : G1310A/B,G1311A/B/C, G1312A/B, G7110B, G7111A/B Issue The inlet valve is defective (leaking) and causing backflow (shown by a flow/bubble going back and forth). The leak test in Lab Advisor suggests the inlet valve may require replacement. A large pressure fluctuation is observed (>2 bar) Requirements Switch off the pump and unplug the power cable. Use an optional solvent shutoff valve or...
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules

    This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules Issue: Reinstallation of the pump head on the pump metering drive during maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module Steps to follow 1. Ensure that your Agilent 1260 Infinity Pump module is turned off at the front power switch. 2. Place your pump head on the pump metering drive. 3. Insert the two pump head bolts through the two holes in the pump head...
  • How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump

    This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pumps. Issue: Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module. Steps to follow Caution : When loosening the pump head blots, make sure that the bolts are loosening evenly left and right. Do not loosen one pump bolt fully while leaving the other bolt tight as it may cause misalignment of the pump head...

How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX

Answer

This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps


Issue:

How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.


Resolution:

The following information should help you find what part of the pump is causing a flow or pressure problem.

Figure 1 and Figure 2 show the hydraulic paths of the pumps.


Figure 1 - Hydraulic path for the Isocratic and quaternary pump
1. From solvent bottles, 2. Inlet valve, 3. Outlet valve, 4. Damper, 5. To sampling unit and column, 6. To waste

 



Figure 2. The Hydraulic Path of the Binary Pump with Damper and Mixer

The most common places for a no flow/no pressure issue in a pump are:

  1. Large leak in pump head.
  2. Total blockage of the solvent bottle inlet filters (p/n 5041–2168)
  3. Blocked or stuck outlet valve or inlet valve.
  4. Blockage in the tube from outlet valve to the pressure damper- it is possible while reinstalling the pump head that the capillary can get crushed.

Troubleshooting steps:

 Warning 
Warning: Put on suitable gloves and other PPE for the solvents contained in your LC solvent bottles before starting this procedure.
Opening fittings while the solvent filters are submerged in solvent will lead to solvents leaks. Ensure that enough lab wipes or other absorbent material is used to soak up any spills. Dispose of lab waste appropriately.
  1. If you cannot see a leak at the bottom of the pump head when you try to purge, then most likely a leak is not the cause of the issue. Move to step 2.
    If you see a major leak from the bottom of the pump head or in the gap between the two pump head, please contact your local Agilent support representative.
    Also see the following knowledge base articles for detailed instructions on how to remove the pump head, change the piston seals, and then reinstall the pump head on the pump module:
     How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module 
     How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
     Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules    
  2. Put on gloves and cover the pump leak sensor/tray with paper towels to soak up any spills. Then disconnect the solvent line going into to the inlet valve at the bottom of the pump head. 
  3. Does it drip freely?
    If the answer is yes, solvent filters are okay, and there seems to be no blockage from the solvent bottles to the inlet valve of the pump head, please move on to step 5.
    If no, remove the solvent filters and see if it now drips freely. If it does, replace solvent filters and go to step 5. If changing filters does not fix the issue, go to step 4.
  4. In a quaternary pump, it may be the Multichannel Gradient Valve (MCGV) that has a blockage limiting your pump flow. Please bypass the MCGV.
    To bypass it, connect a solvent bottle of HPLC-grade IPA (with a clean solvent filter installed) to the active inlet valve on the bottom of the pump head using the Adapter Part Number 0100-1847 (see Figure 4). Attach this adapter to the Active Inlet Valve (AIV) shown by the red arrow in Figure 5, then attach the solvent inlet tube from the bottle of IPA to this adapter.
    If your pump has a passive inlet valve (PIV) installed, then you can connect the solvent inlet line from the IPA bottle to the PIV without the adapter, as shown in Figure 3.
     


    Figure 3. Passive inlet valve connection
    1. Passive inlet valve, 2. Solvent inlet tube



    Figure 4. Adapter, PEEK, female 1/4-28 to male 10-32

    Figure 5. AIV Fitting
  5. Purge the pump.
  6. Loosen the capillary coming off the top of the outlet ball valve (OBV).
  7. Do you see any air or solvent?
    If yes, wait to see if you get a steady flow of solvent. Once established, then tighten the line to see if you have pressure and flow.
    If no, remove the outlet ball valve from the top of the pump head and go to step 8.
  8. Do you see liquid filling up the well when the OBV was removed?
    If yes, sonicate the OBV in methanol or IPA as it may be sticking. (see Tip)
    If no, the inlet valve is the issue. PIVs and AIV cartridges (not the active inlet valve itself, Figure 6) can be sonicated in HPLC-Grade methanol or IPA (see Tip).


     Caution 
    Caution: Only sonicate the AIV cartridge if sonication of an AIV is required. Do not sonicate the body of the AIV as it will damage the electrical components of the valve.


    Figure 6. Active Inlet Valve (AIV)
    1. AIV Cartridge, 2. Active Inlet Valve body


  9. Re-attach and tighten the valves on the pump head, reconnect capillaries, and try purging the pump again.
  10. If the pump is still having issues, then it is time to replace the inlet and/or outlet valves as determined in the preceding troubleshooting steps. Please call you local Agilent Service Representative to arrange a service engineer visit to fix the issue.
  11. Close the purge valve and make sure that the pump reaches pressure and is stable.  
 Tip 
Tips: Note: The solvent used for sonicating depends on the mobile phase. If your mobile phase contains a salt buffer, or old acetonitrile being used in the organic phase then use warm (Less than 60 °C) HPLC grade water for sonication.

It is recommended to have replacement parts on hand to fix the pump issue. If sonicating the valve fixes the issue it is likely to stick again, so users should prepare to replace the valve soon.

 Tip 
Learn more on how to effectively troubleshoot your Agilent LC Pump pump:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent Education
 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help