This Information Applies To: Agilent G1312X and G7112X binary pumps.
Issue:
A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages due to the pump drive experiencing too much resistance during the piston movement:
If your LC system shows any of these errors, then please follow these steps to try to locate and rectify the issue.
Steps to follow:
After each step, turn on the pump, start a normal pump flow, and then check if one of the errors listed earlier re-occurs or not.
Figure 1. Different parts to be checked for a blockage:
Capillary between OBV and Piston 2 (purple), Y-Capillary from outlet to mixing chamber (blue), OBV (green), AIV, and its cartridge (yellow)
If the OBV or the AIV are causing the error, you may try to clean them. Sonicate the OBV or the AIV cartridge (see Figure 1) first in isopropanol, then water and then in isopropanol, each for 10 minutes (do not sonicate the whole AIV, only the cartridge).
|
Caution: Remove the AIV cartridge from the AIV and only sonicate the AIV cartridge. Do not sonicate the AIV body.
|
If this procedure does not solve the issue, you may proceed with checking the pump head interior:
Figure 2. Piston Housing of the Pump Head Assembly: The red arrows show the springs pushing back the pistons. In the right part of the image, the glass piston is visible.
If the procedures described do not solve your issue or enable you to localize a blockage in your flow path, contact your local Agilent support representative for assistance.
|
Learn how to effectively maintain your Agilent HPLC:
Agilent 1260 Infinity LC Systems with OpenLab CDS Learning Path available from Agilent Education
|