Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
Articles Replacing the 5977/7010 HES Ion Source Filaments
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC/MS Portal Home
  • GC/MS Portal Articles
  • GC/MS Portal Videos
  • More
  • Cancel
  • Created :  17 Jan 2023
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC/MS Systems
  • Component :  GC/MS Ion Sources and Samplers GC/MS Single Quadrupole GC/MS Triple Quadrupole
  • Product Name :  5977 Series GC/MSD 7010 Series Triple Quadrupole GC/MS HES EI
  • Task :  Maintenance
Related Articles
  • Labeling the Chromatogram Peaks with MS Library Search Result Using MassHunter Qualitative Analysis

    This Information Applies To: Agilent MassHunter qualitative analysis software Content: MassHunter can be configured to label TIC chromatographic peaks with the best match compound name from an MS library search. Steps to follow: On the top tool bar click Compounds View to open the Method Editor window (Figure 1). Figure 1. Location of Compounds View button Click Method Automation > Workflow , in the Options tab select Workflow > Compound Discovery , Compound mining > Find by Integration...
  • How to Install a Capillary Column on 5977 Series GC/MSD

    Multiple Results
    This Information Applies To: 5977A and 5977B Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut. Warning : a. The GC and MS operate at high temperatures...
    • Found In:
    • Articles
    • ,
    • Articles
  • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)

    This Information Applies To: Agilent 5977 GC/MSD Issue GC/MSD maintenance or repair tasks will require removal and reinstallation of the ion source within the analyzer assembly. Filament exchange Replacement of insulators Ion source cleaning. Types of Sources It is important to identify the ion source type correctly when removing and re-installing an ion source to ensure that the electrical leads are reconnected correctly. Ion source type label is on the heater/sensor part....
  • Introduction to the Layout of Navigator View in MassHunter Qualitative Analysis Software

    This Information Applies To: Agilent MassHunter Qualitative Analysis software version 10.0 Issue Introduction to MassHunter Qualitative Analysis for Agilent GC/MS – Navigator View. Steps to follow Note : Use the demonstration data files evaldemo.d, alkdemo.d, barbdemo.d at C:\GCMS\msdemo . 1. The user interface is separated into three main sections (see Figure 1 ): a. The menu bar. b. The main toolbar. c. Individual windows. An individual window can also have a toolbar...
  • How to Disassemble and Assemble the HES (High Efficiency Source)

    This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD Issue: During normal operation of the GC/MSD, the ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or replacement of consumable parts (dual filament assembly and ceramic insulators). The ion source will then need to be assembled prior to re-installing. Proper abrasive cleaning, assembly, and installation are essential to robust and reliable operation. Tools Needed: Gloves...

Replacing the 5977/7010 HES Ion Source Filaments

Answer

This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems


Issue

The filaments are replaced as part of routine maintenance, or when they break (zero output current).


Items Required

  • Gloves, nylon, lint-free, clean (p/n large 8650–0030, medium G8135-80210, small 8650–0029)
  • Cloth, 100% cotton, lint-free, clean (p/n 05980–60051)
  • T6 Torx driver (p/n 8710–2548)
  • Filament assembly, high efficiency dual filament (p/n G7002-60001)
  • Ion source stand, supplied with instrument.

Steps to follow

 Use Figure 1 as reference to identify the HES source parts in this procedure.


 Figure 1
Figure 1. HES ion source parts
1. Filament block securing screw, 2. Ion source mount, 3. Finger grip on the filament block assembly, 4. Dual filament

Removing the old filament

  1. Vent the MS, see How To Vent (Shut Down) the GC/MSD - EI - MassHunter

       Warning: The analyzer, GC/MS interface, and other components in the analyzer chamber operate at high temperatures. Do not touch any part until you are sure it is cool.
  2. Open the analyzer chamber, remove the ion source, and place it in the ion source stand on a clean cloth

      
    Caution: To prevent contamination when working in the analyzer chamber always wear clean, lint free gloves.
  3. Remove the screw securing the filament block to the source mount (Figure 2)



    Figure 2. Removing the securing screw
  4. Hold the finger grip on the filament block, invert the ion source, and lift the source body off the filament block, and remove the filament block from the source.

      
    Caution: Dual filaments are brittle. Handle with extreme care.
  5. Remove the dual filament from the filament block (see Figure 3)



    Figure 3. Removing the dual filament

Installing the new filament

  
Caution: Caution: Dual filaments are brittle. Handle with extreme care.

 

  1. Insert the three pins on the dual filament through the rear of the filament block (see Figure 4)



    Figure 4. Installing the dual filament
  2. Holding the ion source upside down, place the filament block in the source mount (see Figure 5)



    Figure 5
    . Installing filament block
  3. Place the source in the ion source stand

  4. Use a T6 Torx screwdriver to tighten the screw securing the filament block to the source mount (Figure 6)



    Figure 6. Tightening the securing screw
  5. Reinstall the ion source

  6. Close the analyzer chamber and pump down the MS. See How To Pump Down (Start) the GC/MSD - EI - MassHunter

  7. The MS will need an autotune before running samples. 

  Learn how to effectively maintain your Agilent GC/MS systems:
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 5977 GC/MSD with OpenLab CDS e-learning paths available from Agilent education
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help