Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles How to Change the Needle and Needle Seat in an Agilent G7129x Vialsampler
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Injectors and Sample Thermostats
  • Product Name :  1260 Infinity II Bio-Inert LC System 1260 Infinity II LC System 1260 Infinity II Prime Bio LC System 1260 Infinity LC System 1260 Infinity II Prime LC System
  • Task :  Maintenance
Related Articles
  • How to Change the Needle on a G7167x Multisampler

    This Information Applies To: Agilent G7167A/B Multisampler. Issue: Required when: The needle is visibly damaged. Leaks or blockages in the vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of yearly maintenance Steps to follow: Tools Required: 8710–0510 - Open-end wrench 1/4 - 5/16 inch Parts Required: G4267-87201 - Needle assembly OR G4267-87210 -...
  • How to Remove and Reinstall the Needle in Agilent G4226X/G1367E High Performance (HiP) Autosamplers

    This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosamplers Issue Replacing the Needle Assembly of the Autosampler is required when: The needle is visibly damaged. Leaks or blockages in the Autosampler are observed from the needle. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of yearly maintenance. Tool Required Wrench, open 1/4 - 5/16 inch (p/n 8710-0510 ) Parts Required...
  • How to Change the Needle Seat on a G7167x Multisampler

    This Information Applies To: Agilent G7167x Multisampler modules Issue: Required when: The needle seat is visibly damaged, blocked, or is leaking. Leaks or blockages in the Vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of the yearly maintenance procedures. Steps to follow Tools Required: 8710–0510 - Open-end wrench 1/4 - 5/16 inch Flathead screwdriver...
  • Removing and Reinstalling the Needle Seat in Agilent G4226X/G1367E High Performance (HiP) Autosamplers

    This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosampler Issue Replacing the Needle Seat of the Autosampler is required when: The needle seat is visibly damaged. Leaks or blockages in the Autosampler are observed from the needle seat. Replacement the needle seat is performed as part of yearly maintenance. Tool Required Wrench, open end 1/4 - 5/16 inch (p/n 8710-0510 ) Parts Required For an Agilent 1290 Infinity (G4226A...
  • Backflushing the Needle Seat of the Agilent Vialsampler

    This Information Applies To: Agilent Infinity II Vialsamplers G7129A/B/C. Issue: High pressure in the Vialsampler is oftentimes due to a blockage in the needle seat assembly. The needle can pick up sample particulates or plugs from the vial caps when the Vialsampler is drawing the sample. When the valve switches back to mainpass, they are forced into the seat capillary and cause obstruction and high pressure. Steps to follow: To backflush the needle seat according to the following procedure...

How to Change the Needle and Needle Seat in an Agilent G7129x Vialsampler

Answer

This Information Applies To: Agilent G7129A/B/C Vialsampler


Issue:

Replacing the needle and needle seat of the Vialsampler is required when:

  • The needle is visibly damaged.
  • Leaks or blockages in the vialsampler are observed, especially between the needle and needle seat.
  • The limit for the needle interaction EMF counter is exceeded.
  • The needle needs to be replaced as part of yearly maintenance.

Over time, deposits can block the needle and the needle seat causing pressure issues and leaks. Also wear and tear of the needle against the seat during normal operation eventually damages both assemblies, requiring replacement. 


Steps to follow:

Tools Required:

  • Open-end wrench 14 mm (p/n 8710-1924)
  • Screwdriver Torx TX-10 (p/n 8710-2140)

Parts Required:

One of the following needle assemblies depending on the Vialsampler configuration:

  • Needle assembly, 1260 Vialsampler (p/n G7129-87200)
  • Needle assembly, 1290 II Vialsampler (p/n G7129-87201)
  • Needle assembly, slotted, for high injection volumes (p/n G7129-87202) 

And one of the following needle seats depending on the Vialsampler configuration:

  • Seat Assembly, 1260 Vialsampler, PEEK, 0.17 mm (p/n G7129-87017)
  • Seat Assembly, 1290 Vialsampler, PEEK, 0.12 mm (p/n G7129-87012)
  • Seat Assembly, 1260 Vialsampler, Vespel, 0.17 mm (p/n G7129-87117)
  • Seat Assembly, 1290 Vialsampler, Vespel, 0.12 mm (p/n G7129-87112) 

See also: Needle & Needle Seats for HPLC

 Warning 
Warning: This procedure risks exposure to hazardous, toxic, or flammable solvents and reagents. When working with these substances observe safety procedures (for example by wearing personal protective equipment) as described in the material handling and safety data sheet supplied by the chemical vendor, and always follow good laboratory practice.

 Tip 
Tips: It is recommended to always exchange the needle assembly and the needle seat at the same time to prevent premature leakage.
  1. Finish any pending or running acquisition job.

  2. Stop the flow at the pump and remove the solvent lines from the eluent bottles to avoid spilling solvent.

  3. Close the shutoff valves at the pump if available on your pump module.

  4. In the Agilent Lab Advisor Software, connect to your LC system then click Service and Diagnostics>Maintenance Positions>Change Needle, Loop and Seat.

  5. Click Start to start the maintenance task. Confirm the part number for the needle and needle seat listed in the pop-up configuration dialog box match the new needle and needle seat being installed in the Vialsampler. (See Figure 1)


    Figure 1. Opening the Change Needle, Loop and Seat task in Lab Advisor

  6. If using the handheld local controller then start the maintenance mode on the controller and select the Change Needle, Loop and Seat function.

  7. Open the doors of the Vialsampler

     Warning 
    Warning: Risk of injury by uncovered needle. Do not open the safety cover of the needle station during normal operation. Wear safety goggles and safety gloves when removing the needle assembly.
  8. Remove the safety cover by loosening the Torx T-10 screw on the side of the safety cover and titling it up and then away from the Vialsampler. (See Figure 2).



    Figure 2. Loosen the Torx T-10 screw then tilt the safety cover up and then away from the Vialsampler.

     Caution 
    Caution: It is important to move up the wash port and then rotate it to the left until it locks in the service position before servicing the needle and needle seat. Failure to perform this step may lead to alignment issues, and bent needles and wash port damage if the needle hits the top of the wash port during movement to the needle seat after servicing. Always ensure that the following step is performed before moving on with needle and needle seat exchange.


  9. Move the wash port into the service position by slightly lifting it up and then rotating it to the left until it locks in place. (See Figure 3)



    Figure 3. Lift up the wash port, and then to the right until it locks in place.

  10. In Lab Advisor, click Next to move the needle in the down position (approx. 2 mm above the seat). If using the handheld local controller, move the needle down until the needle tip is approx. 2–4 mm above the seat. This will allow access to the needle assembly to remove the sample loop

  11. Disconnect the sample loop from the needle, and, if needed, also from the top of the Analytical Head. (See Figure 4)


    Figure 4. Disconnect the loop capillary from the needle, and if necessary, the analytical head.

  12.  Loosen the fixing screw holding the needle assembly in place, then gently lift out the needle assembly sliding it up off the alignment pin.

  13. Disconnect the seat capillary from the injection valve (port 5) and ease out the needle seat from the leak tray with the help of a flathead screwdriver. (See Figure 5)


    Figure 5. Removing the needle seat from the leak tray.

  14. Insert the new needle seat and press it firmly into position. Then connect the seat capillary fitting to port 5 of the injection valve. (See Figure 6)


    Figure 6. Installing the new needle seat.

  15. Mount and align the new replacement needle assembly (see Parts Required section) onto the alignment pin so that it is directly over the needle seat. Then tighten the fixing screw firmly to secure the needle in place over the needle seat.

     Warning 
    Warning: In the following step, take care when adjusting the needle to avoid any hand injuries with the sharp needle tip.
  16.  Align the needle tip with the seat and ensure that they are concentric. If needed, carefully bend the needle into the right position with your fingers. (See Figure 7)


    Figure 7. Aligning the needle tip with the center of the needle seat.

     Caution 
    Caution: Do not overtighten the loop capillary fitting in the next step. Incorrect positioning and installation of the needle/loop connection can result in damaging or breaking the sample loop.
  17. Reconnect the sample loop to the needle and to the metering head if removed. Make sure of the correct position of the loop capillary - the uncoated part of the capillary must be horizontal. See the beginning of Figure 4 for the correct installation orientation of the needle assembly. Do not over tighten this fitting.

  18.  In Lab Advisor, click Next to lift the needle slowly into the up position. If using the local controller, move the needle up to the uppermost position.

  19. When instructed in Lab Advisor, move the wash port back down into its original starting position over the needle seat by turning it right and down. (See Figure 8)


    Figure 8. Move the wash port back to the operating position over the needle seat by turning it right and down.

  20. Reinstall the safety cover removed in step 8, and tighten the Torx T-10 screw to hold the safety cover in place.

  21. In Lab Advisor, click Next to move the needle into the needle seat, and then click Back to finish the maintenance procedure. If using the local controller, select Continue to move the needle into the needle seat and then Exit to finish the maintenance procedure.

  22. Close the doors of the Vialsampler.

If possible, run a system pressure test with the port 6 of the injection valve blanked using the appropriate blanking nut to ensure that the Vialsampler flow path is pressure tight to a pressure higher than your application require.

 Tip 

Learn more on how to effectively troubleshoot your Agilent G7129x Vialsampler:
Agilent 1260 Infinity LC Systems with OpenLab CDS e-learning path available from Agilent Education

 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help