This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments and ceramic insulators). The ion source will then need to be assembled prior to re-installing. Correct abrasive cleaning, assembly, and installation are essential to robust and reliable operation.
Tools Needed:
• 597X Tool Kit and Cleaning Supplies p/n G7077-60566 (contains)
Figure 1: Parts Breakdown of the EI Source
Item Number | Item Description | Part Number |
1 | Gold plated set screw | G1999-20022 |
2 | Gold plated screw | G3870-20021 |
3 | Interface socket | G1099-20136 |
4 | Source body | G1099-20130 |
5 | Drawout cylinder | G1072-20008 |
6 | Drawout plate, 3 mm | 05971-20134 |
7 | 4-turn filament | G7005-60061 |
8 | Spring washer | 3050-1374 |
9 | Lens insulator | G3170-20530 |
10 | Entrance lens | G3170-20126 |
11 | Ion focus lens | 05971-20143 |
12 | Repeller insulator | G1099-20133 |
13 | Repeller | G3870-60172 |
14 | Flat washer | 3050-0982 |
15 | Belleville spring washer | 3050-1301 |
16 | Repeller nut | 0535-0071 |
17 | Source heater block assembly* | * |
18 | Repeller block insert | G3870-20135 |
* If replacement is required then contact Agilent for repair options
To Disassemble the EI source:
Figure 2. Removing the filaments, repeller and insulators from the source heater block assembly.
Figure 3. Removing the lens, drawout plate, drawout cylinder and insulators from the source body.
To Assemble the EI Source:
Caution: Do not overtighten the repeller nut or the ceramic repeller insulators will break when the source heats up. The nut should only be finger-tight
|
Figure 4. Assembling the lens, insulators, drawout plate and drawout cylinder into the source body.
Caution: Do not overtighten the interface socket. Overtightening could strip the threads.
|
Figure 5. Assembling the repeller, insulators, and filaments into the source heater block assembly.
Learn how to effectively operate your 5977:
Agilent 5977 GC/MSD with MassHunter Workstation Agilent 5977 GC/MSD with OpenLab CDS e-learning paths available from Agilent education |