Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Chromatography Software
  • Technical Areas
Chromatography Software
Wiki How to ... align uncalibrated peaks in 2 or more signals
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • +Software Collection of Resources
  • 7694B Headspace Sampler Software Support
  • Another calibration report for OpenLAB CDS ChemStation
  • Another calibration report for OpenLAB CDS EZChrom
  • Autostart macro failed error 999 in OpenLAB CDS ChemStation
  • Belarc Advisor for PC Profiling
  • Calculate date and time differences - Save date and time in IR
  • Calibration History Report (OpenLAB CDS)
  • Chemstation .CSV generation Decimal Places
  • Chemstation Chemmain Error During Startup
  • ChemStation Scheduler Commands
  • OpenLab CDS and Windows 7 Podcast
  • Computer Maintenance
  • Configuring OpenLAB CDS Chemstation for a 6890 with 3rd party detector
  • Connection to selected server failed Error 1068 - dependency service failed
  • DAMethod Register Export
  • Error Message "No Signals Available" appears when loading data
  • Error Message: Vial Range -1 to -1 - OpenLAB CDS Chemstation
  • Error: No Export File Selections Made Yet - OpenLAB CDS Chemstation
  • How to .. hide a table row based on a calculated value (OpenLAB CDS Intelligent Reporting)
  • How to ... add a percent sign to a PRSD number.
  • How to ... add LC diagnostic data to Intelligent Reports (OpenLAB CDS Intelligent Reporting)
  • How to ... add several comments to your report template without generating extra space on the report (OpenLAB CDS Intelligent Reporting)
  • How to ... align text boxes to the editor's grid (OpenLAB CDS Intelligent Reporting)
  • How to ... align uncalibrated peaks in 2 or more signals
  • How to ... auto-extract signals from a UV iso-absorbance plot
  • How to ... calculate summary statistics on computed numbers
  • How to ... calculate with a specific mutiplier or dilution factor generated by OpenLAB CDS EZChrom
  • How to ... filter for EZChrom run types QC Standard  (OpenLAB CDS Intelligent Reporting)
  • How to ... populate the template editor with your data
  • How to ... print ChemStation's live system values (OpenLAB CDS Intelligent Reporting)
  • How to ... report the number of repeated injections with OpenLAB CDS EZChrom
  • How to ... report the output of a category aggregator  (OpenLAB CDS Intelligent Reporting)
  • How to ... store a date stamp as a variable? (OpenLAB CDS Intelligent Reporting)
  • How to ... use external report parameters (OpenLAB CDS Intelligent Reporting)
  • How to ...  setup an array of compound specific parameters
  • How to configure Agilent GC headspace system within Waters Empower and ICF
  • How To Overlay Chromatograms in OpenLab ChemStation
  • How to... report sequence related customized parameters (OpenLab ChemStation Intelligent Reporting)
  • HowTo : Report the n Biggest Peaks of a sample analysis (OpenLab CDS Custom Calculator)
  • KB: Useful Free Third-Party Software Tools
  • Little script to print Text and Numbers as Results ( e.g. below detection limits ) (OpenLAB CDS Intelligent Reporting)
  • LookInMyPC - PC Profiling and Diagnostics
  • Make your own custom way of displaying the Date and Time in your reports
  • Migrating from OpenLAB CDS EZChrom Edition to OpenLab CDS Version 2.3 – Workstation to Workstation
  • MSD Error: HPMSAXQ0.DLL failed upon launching Openlab - OpenLAB CDS Chemstation
  • Notepad++
  • Openlab CDS C.01.07 and 6890 classic driver
  • Reprocessing data using a new calibration table
  • Sequence Aborted Due to File Access Error - OpenLAB CDS Chemstation
  • Sequence container/result set data will not load in navigation panel - OpenLAB CDS Chemstation
  • Video Series: OpenLab CDS for ChemStation Users
  • When trying to run a sequence, user gets error "macro has no name"
Still Need Help?

Post your question in our User Forum or Contact Support.

How to ... align uncalibrated peaks in 2 or more signals

Created by Agilent Agilent over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh over 1 year ago

Use case

2 signals; DAD and MSD TIC with many uncalibrated/unknown peaks.

 

Challenge

How to print UV and MS peak apex spectra for each pair of corresponding peaks

 

Due to the setup of detectors, the MS signals are shifted and the retention times could not be used to identify corresponding pairs of peaks.   Even after correcting the retention time shifts the numbers do not match precisely and results of two corresponding peaks cannot reported together.

 

Solution

Since retention times do not match precisely, the retention time values can be rounded to such an extend that the rounded retention times become equal for both corresponding peaks.


Example:

  • Drag a peak result table, a UV spectra plot and a MS plot on the template
  • Select these 3 items and combine them in a new group

 

The definition of the group repetition will take care of the alignment, similar to the retention time windows in an peak identification table

=Peak_RetentionTime   (repeats the result per individual peak retention times)

=(1/60)*int(60*Peak_RetentionTime)  (repeats the result per peak retention times rounded to 1 sec)

=(1/n)*int(n*60*Peak_RetentionTime)  (repeats the result per peak retention times rounded to n sec)

 

By introducing a report parameter n (granularity factor, time window), the user could modify the granularity of the peak alignment in the  report preview.  This is likely a method specific parameter that needs to be adjusted. Once a suitable value is found, the report parameter could be hidden to the template user.

 

The approach described above generates a fix time window for peak recognition and does not take the increasing peak width over run times into account. Adding the peak width as an additional factor could take this effect into account.

 

In case of larger signal shifts, the alignment of the peak results might be achieved by an IIF condtion in the group repetition expression.

 

The following expression is just an example and has not been tested, yet:

=iif(Signal_name="MSD",(1/n)*int(n*60*(Peak_RetentionTime-0.3)),(1/n)*int(n*60*Peak_RetentionTime))

 

 

(originally posted in OpenLAB Forum: https://zohodiscussions.com/openlabcdsforum#Topic/68030000000179003

2015 May 4 by betrich)

  • OpenLab CDS
  • OpenLab Intelligent Reporting
  • alignment
  • OpenLab
  • uncalibrated
  • Intelligent Reporting
  • reporting
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help