I am trying to run a Injector Precision PV test using the conditions described in the software: caffeine standard (125 ug/mL) and HP ODS Hypersil 5 um, 125 x 4 mm part 7982618-564 . The method uses 85:15 Water:ACN mobile phase at 0.8 mL/min. The test…
Hello mkaadan ,
What I meant by a "low" failure rate (notice the quotation marks), is that it is likely not an indication of a leak, as that would result in a lot more samples failing, or potentially all of them. By no means I wanted to minimize the…
Hello Paul, it looks to me that there is too much acetylene in the gas mixture coming to the burner. This is causing the carbon build-up.
I'd suggest you first remove the absorption cell from the holder and clean this by soaking in dilute nitric acid…
l have lc 1260 pump type G1311c and it is in residual mode due to installing a degasser board on the pump board. I used the Lab Advisor program to update the firmware, but the process was not successful and messages appeared. Connection failed instrument…
Hi, I have a GC/MS (8890 GC with 5977b HES MSD) that is also configured with an FID and PDHID. I would like to create an FID/PDHID-only method, where the MSD remains off the entire run and does not collect data. I haven't figured out how to do this so…
Can SimDis data processing be performed manually (i.e. without the aid of additional SimDis software produced by Agilent or a third-party) with a GC running Chemstation?
I'm looking for a low-cost solution to perform SimDis analysis of jet-fuel samples…
Hi everyone. I have a problem with my 8890 GC-MS as I do not have pressure in the inlet. The issue began during the IT outage, the EPC module turned off itself hence all heated zones turned off as well. When I managed to re-establish the connection I…
Hi ISH, yes all was fine with column installation and pump down was efficient. The issue seems to be with the entrance lense offset - I loaded a previous tune and managed to run an autotune however it is failing due to too high 503/502 ratio. I've seen…
HI Lars,
Yes, you can run the test on a VI. Specific to your system, a failing result could indicate a leak within the inlet or at any of the in-line connections of your GSV, so the results may not be clear without a point leak detector. I would also…
This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems
Issue
The filaments are replaced as part of routine maintenance, or when they break (zero output current).
Items Required
Gloves, nylon,...
This Information Applies To: Agilent 5977A/B/C GC/MSD
Issue
The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source.
Items Required
Electron multiplier horn (5977B/C:...
This Information Applies To: Agilent 5973/ 5975/5977 GC/MSD with rotary vane foreline pump
Issue:
The oil mist filter is recommended to be replaced annually. If the color of the filter has changed or there is smell from the filter, the oil mist...
This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments...
This Information Applies To: Agilent G7129A/B/C Vialsampler
Issue:
Replacing the needle and needle seat of the Vialsampler is required when:
The needle is visibly damaged.
Leaks or blockages in the vialsampler are observed, especially between...
This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps
Issue:
How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.
...
This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due...
This Information Applies To: Agilent G1312X and G7112X binary pumps.
Issue:
A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages...
This Information Applies To: Agilent 1100, 1200, 1260 LC Pumps.
Issue:
The following procedure allows the cleaning of residue from the pump's check valves, often caused by polymerization of aged acetonitrile or salt-based solutions that may prevent...
This Information Applies To: Agilent LC Systems
Issue
After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to...