Hello,
Error 1213 typically comes with other errors that truly describe the issue, like the 2150 error you are getting.
First, check the cone orifices. Use a microscope or powerful magnifying glass. If they are clogged or salted over, clean. The opening…
This may be due to several factors. First of all, make sure that all plasma monitor connections and cables are properly connected and not damaged. Also check that there is sufficient gas flow (eg argon) to maintain the plasma. Check the cleanliness and…
Hello all,
I´m facing a Silicium contamination in ICP-MS analysis, I have used a 7800 ICP-MS.
I have digested my samples with HF, HCL and HNO3 and neutralized it with boric acid, after the analysis I have Silicium CPS has increased around 30%.
…
Hi Layden,
your source has died (normal ratio Amps/Volts should be starting at 1.9/2.7). Each year you will find a decline in Amps until it has reached 1.6, then it will give a red bar. In your case it's 0, so the filament is probably broken - your…
I have almost the same system as shown in the picture for over a year now. The application was supplied by Da Vinci Laboratory Solutions in the Netherlands. It is a Mettler TGA 2, with a Gerstel MPS autosampler on a Agilent GCMS(MS) with a CIS injector…
Thank you. I contacted the support. I just had to login into agilent subscribenet and download the latest version after entering the authorization key.
Hello,
That looks like a standard IS calibration using area percent. The purity of the standard can be entered as the compound multiplier. I would start by searching the help for calibration, internal standard, and mass percent calculations. You can…
Hello Lucas_S ,
Based on your follow-up explanation in this discussion -
(+) Macro for Batch Export of CSV in OpenLab CDS C.01.10 to a unique archive. - Forum - Chromatography Software - Agilent Community
I don't believe this macro could be easily…
This Information Applies To: Agilent MassHunter 11 and later
Issue
Generating Acquisition Desktop and Quantitative Analysis shortcuts.
Steps to Follow
Double-click the Control Panel icon on the desktop or select Windows Start menu > Agilent...
This Information Applies To: Agilent MassHunter Workstation for LC/TQ and MassHunter Workstation for LC/TOF and LC/Q-TOF
Issue
How to view the MassHunter Data Acquisition Software Logs entries, change settings, and export the Logbook for archiving...
This Information Applies To: ICP-MS MassHunter with User Access Control (UAC)
Issue
It is necessary to update the database folder path in Change User Access Control:
After creating a new database in SDA with ICP-MS MassHunter.
To view a batch...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Over time, the electrodes may become pitted or eroded and users may see a decrease in precision or begin to experience the "Tube fault" error when new tubes are installed. If...
This Information Applies To: Agilent GTA Furnace and Flame Instruments, Agilent 240FS/240GFAA/GTA120/PSD120
Issue
There may be occasions when users need to remove GTA furnace center block workhead (instruments with both Flame and Furnace capabilities...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
The shroud should be replaced annually, or when there is evidence of damage caused by mechanical shock or contamination from liquid spillage.
Items Required
Shroud Removal...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Before replacing the electrodes and/or shroud, the GTA Furnace workhead will need to be disassembled. After replacing the electrodes and shroud, the GTA Furnace workhead will...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer
Issue
After starting a run, the run is aborted with the following error: Current = 0.0uA, Expected range: 5 - 3000.
Steps to Follow
Check...
This Information Applies To: Agilent OpenLab CDS 2.x in file system configuration
Issue
Sometimes, you will need to change the Projects Root Path / Project Storage Path . This article outlines the procedure on how to change it.
Background
...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7
Issue
Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES
Tip : These steps should be completed after the Subsystem Communication...