Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Bypassing the AVS for the 5000 Series ICP-OES
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Spectroscopy Accessories ICP OES Systems
  • Component :  Atomic Spectroscopy Sample Introduction Accessories
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES AVS 4 AVS 6/7
  • Task :  Maintenance Operation
Related Articles
  • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X Issue The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping) the AVS and running a worksheet during can identify if the AVS is the root cause of the issue. Other root causes could be the method application, or an instrument fault. Items Required If an internal standard line is required, install the...
  • AVS 6/7 Tubing Connections and Part Numbers

    This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7 Issue This article shows the required t ubing connections and part numbers between the AVS 6/7 and ICP-OES ICP-OES sample introduction system: Figure 1. Sample introduction system and AVS 7 (shown as 1) Figure 2. Connections to AVS: 1 to AVS Pump, 2 from Autosampler probe (rinse), 3 Argon gas, 4 Rinse/Carrier Figure 3. AVS (showing tubing removed): (A) AVS valve, (B) Sample loop holder, (C) AVS pump inlet, (D) AVS...
  • Replacing the AVS 6/7 Rotor

    This Information Applies To: Agilent 5000 Series ICP-OES Systems, AVS 6/7 Issue Over time the AVS 6/7 rotor can become worn or damaged by chemicals. Agilent recommends inspecting and/or replacing the rotor every 5000 AVS switches or when analytical results or precision fall outside expected limits. Steps to Follow: Tip : Refer to the Help section of the Agilent ICP Expert Instrument Software for further detailed instructions. Warning : Always wear appropriate...
  • AVS 6/7 Maintenance Checklist

    This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7 Issue The AVS 6 and AVS 7 require routine maintenance. This article provides a maintenance checklist. Figure 1. 1 AVS 7 Warning : Always rinse all tubing and the valve before disconnecting the tubing from the switching valve to avoid contact with potentially dangerous chemicals. Appropriate personal protective equipment (PPE) should be worn when handling components contaminated with chemicals. ...
  • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments

    This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert Issue How to Reset the AVS 6/7 after maintenance or when a red LED is visible Steps to follow Tips : Instructions for changing the AVS switching valve can be found in the ICP Expert Help under “Hardware How to” and “Replace the AVS 4/6/7 Tubing”. Figure 1. ICP Expert Help After replacing the AVS valve and tubing, a red LED is showing on the AVS plate. Figure 2 shows the location...

Bypassing the AVS for the 5000 Series ICP-OES

Answer

This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7


Issue

Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES


 Tip 
Tip: These steps should be completed after the Subsystem Communication, Air Flow, Water Flow, Gas Flows, RF Generator, Camera, and Optics Tests have been completed with a status of ‘Done’ or ‘Pass’

Bypass (or skip) the AVS to run a worksheet and troubleshoot instrument performance issues such as sensitivity and precision. The root cause may be related to an AVS component, the method application, or an instrument fault and bypassing the AVS helps identify the root cause.

With the AVS 6/7 installed on the 5000 series ICP-OES, the sample introduction area will look like Figure 1 (AVS 6) or Figure 2 (AVS 7). The setup difference between the AVS 6 and AVS 7 is the seventh port on the AVS 7 valve for the internal standard. An internal standard tube must be installed on the peristaltic pump.


Figure 1. AV6 connections
1 To Autosampler, 2 To carrier, 3 To waste


Figure 2. AVS 7 connections
1 To AVS, 2 To waste

Steps to follow

  1. Disconnect the UniFit sample connector from the nebulizer. Unscrew the threaded flangeless fitting from port 2 of the AVS, remove, and set this fitting aside. It will not be used when bypassing the AVS (Figure 3):


    Figure 3. UniFit sample connected to the nebulizer and Port 2 of the AVS

  2. Disconnect the barbed end (labeled “Peri Pump Rinse”) of the T-piece from the rinse tubing on the peristaltic pump (Figure 4):


    Figure 4. Peri-Pump rinse connector 

  3. If an internal standard line is installed on the AVS 7, unscrew the threaded internal standard fitting from port 7 of the AVS and remove the tubing from the peristaltic pump. Set the internal standard tubing and connection aside. It will not be used when bypassing the AVS (Figure 5):


    Figure 5. Internal Standard Line connector

  4.  Locate the UniFit sample connector, 0.75 mm id, for U-Series SeaSpray concentric nebulizer. This connector is included with a new nebulizer (Figure 6):

     

     Tip 
    Tip: This connector is also available to order as a pack of 10 (Agilent part number G8010-80035).


    Figure 6. UniFit connector

  5. Attach the UniFit sample connector to the nebulizer and insert the capillary tube end into the rinse tubing on the peristaltic pump (Figure 7):


    Figure 7. Connector and Rinse Tubing connected

  6. Remove the autosampler threaded fitting from port 5 on the AVS that connects the autosampler probe tubing to the valve (Figure 8):


    Figure 8. Port 5 of the AVS

  7. Locate the ICP-OES connection kit, Autosampler to Peristaltic Pump (p/n G8010-60348). Find the small rectangular block with a barb on one end and a threaded port on the other end. If you:
    1. Have this block: Attach the autosampler threaded fitting into the threaded port of the small rectangular block and insert the barb into the free end of the rinse tubing on the peristaltic pump. This peristaltic pump tubing will no longer act as a rinse line but will now be the sample line and will deliver the solution to the nebulizer, bypassing the AVS valve.
    2. Do not have this block: Pull the threaded autosampler connection fitting off the autosampler capillary and insert the autosampler capillary end into the free end of the rinse tubing on the peristaltic pump. This peristaltic pump tubing will now be the sample line and will deliver the solution straight to the nebulizer, bypassing the AVS valve. It is recommended to have spare autosampler connection fittings available (p/n G8010-60306).
  8. Clamp the peri-pump tubing, send the autosampler to a vial containing DI or rinse, turn on the peristaltic pump, and confirm that the solution is being delivered to the nebulizer at a smooth steady rate. Confirm that the spray chamber is draining. Adjust the roller arm tension for the sample and drain if needed. Turn off the peristaltic pump when completed.

  9. Perform an ICP-OES ICP-OES Detector and Instrument Calibration 
 Tip 
Learn how to effectively troubleshoot your Agilent AVS7 on the 5000 Series ICP-OES:
Agilent 5000 Series ICP-OES Maintenance and Tips for Optimal Performance e-learning course available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help