This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Before replacing the electrodes and/or shroud, the GTA Furnace workhead will need to be disassembled. After replacing the electrodes and shroud, the GTA Furnace workhead will need to be reassembled.
Steps to Follow
Disassemble the workhead before the removal and replacement of the furnace electrodes and graphite shroud. This step allows the relevant tooling and procedures to be completed.
|
Warning: Always turn off the furnace and disconnect the power supply. Allow the furnace to cool completely before handling it.
|
Disassembling the furnace
Figure 1. Replacing quart window assemblies
2. Using the lever, open the workhead. The right electrode assembly should fall away from the center block (Figure 2).
Figure 2. Opening the workhead
3. Pull the center cooling block away from the left electrode housing. Position the cooling block over the rear of the unit to maximize access to the electrodes (Figure 3). Be careful not to damage the tubing.
Figure 3. Pulling cooling block and positioning it over the rear
4. Replace the electrode and shroud, with the center block positioned (Figure 3). The tubing can remain in place when removing and replacing the graphite shroud in the center block.
Tips: See Removing and Replacing GTA Furnace Electrodes and Replacing the GTA Shroud for more information.
|
Reassembling the furnace
Follow the steps outlined to reassemble the workhead to normal operating position.
Figure 4. Reconnecting the cooling block
2. Reconnect the right side electrode to the workhead using the lever (Figure 5).
Figure 5. Reconnecting the right side electrode
Tips: With the quartz windows removed, check the O-ring on the base of the assembly. It should be supple and free from kinks or cuts. Check the silicone rubber seal on the window. If either are damaged or worn, replace them.
|
3. Reinstall both quartz window assemblies (Figure 6).