This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Over time, the electrodes may become pitted or eroded and users may see a decrease in precision or begin to experience the "Tube fault" error when new tubes are installed. If any of these instances occur, the electrodes should be replaced.
Requirements
Items Required
Steps to Follow
Removing the Furnace Electrodes
Warning: Ensure that the control unit is switched off and unplugged. Ensure that the workhead has been allowed to cool completely before proceeding.
|
Before replacing the electrodes, disassemble the GTA Furnace workhead.
Figure 1. Direction of removal
Figure 3. Placing the spacer over the electrode, follow with the collar and washer on top of the spacer
8. Repeat steps 1 through 6 on the left side electrode (Figure 8)
Figure 8. Repeating steps 1 through 6 on the left side electrode
Replacing the Furnace Electrodes
Caution: Always replace the left side electrode first.
|
Figure 9. Direction of installation.
Warning: Hot Surface
Contact with hot furnace atomizer surfaces can cause serious burns. Always switch off the control unit, disconnect the power supply cable, and allow time for the workhead to cool to a safe handling temperature before handling. |
Figure 12. Installing the bolt, adapter, and electrode into the electrode housing
Figure 15. Electrode is pulled into the housing evenly
Caution: Once you feel resistance, stop tightening the nut. Over-tightening the nut will crack or damage the electrode.
|
Figure 17. Repeating steps 1 through 7 on the right side electrode
Tips: To ensure that the electrodes are fitted correctly, place a graphite tube between the two electrodes. Without the center block (the shroud) in place, clamp the electrodes together. You should be able to turn the graphite tube with your fingers. If there is too much resistance, the electrodes may need to be pushed into the furnace bore further
|
The workhead can now be re-assembled. See Removing the GTA Furnace Workhead.