This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer
Issue
After starting a run, the run is aborted with the following error: Current = 0.0uA, Expected range: 5 - 3000.
Steps to Follow
Figure 1. Drawers and compartments of a parallel capillary electrophoresis system. The example shown is a 5200 Fragment Analyzer.
The top compartment and top two drawers must be closed while the instrument is running. The lower drawers and side drawers can be opened.
1. Top compartments: capillary array and optics, 2. Side compartments: pump and reagent bottles, 3. Drawers.
Figure 2. Instrument top compartment. High-voltage supply cable. The example shown is a 5300 Fragment Analyzer.
1. Holder with high-voltage plug, 2. High-voltage cable connection.
After performing the troubleshooting steps, perform a sample run again to check if the correct current is now reached. If these steps do not help, contact Agilent support. To find the correct contact, visit the Contact Us page and choose your country.