Operating like this puts sub-optimal flow into the GCMS ion source of more than 2 ml/min. Please make sure that you have a system with a turbomolecular high vacuum pump and not a diffusion pump as 2 ml/min exceeds the maximum allowable flow into a diffusion…
Exciting news in the world of FTIR spectroscopy!
Agilent has just released MicroLab 5.8 , MicroLab Pharma 5.8 , and MicroLab Expert 1.3 —the software that powers the Cary 630 FTIR spectrometer and Agilent’s handheld and mobile FTIR systems . MicroLab…
Hi Margot,
It may well be that the column performance has deteriorated so much within the 7 months that the EasiVial standards are measured with poorer PDI values. The column performance can be tested with regular plate count determinations. It is a…
Hi sojeong,
every test report, delivered with aour columns, contains multiple specifications, but they are all measured with specialized systems and software. That means you might not be able to reproduce them identically. BTW for plate count you also…
When the tuning is done at 3200 (+) everything is fine. When I change it to tune at 1700 (+) the masses are reasonably accurate. After it gets to the third mass (620) it appears that the entire spectrum shifts by about 1.5 AMU and the tuning fails. There…
Hi Siddhartha,
Thank you for contacting Cell Analysis Technical Support. Can you please go ahead and send us your System Files so that we can see what might be going on? These can be accessed in the Help Section of the Wave software. You can select…
Hello thank you very much for your response, thanks for the firmware data, the degasser worked perfect, the firmware change no longer, and I can not find the previous version. and are from the 6.80, but how is it possible that a team of thousands of dollars…
Thanks for the reply. I am monitoring at 254 nm. Higher wavelengths do not show that drift. It must be related to the THF (I am using unstabilized) but I do not see how.
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...