Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Checking for External Leaks on 7697A Headspace
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Sample Preparation and Introduction
  • Component :  Headspace Samplers
  • Product Name :  7697A Headspace Sampler
  • Task :  Troubleshooting
Related Articles
  • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler

    This Information Applies To: Agilent 7697A headspace sampler. Issue: The headspace sampler may present a flow shutdown or poor chromatography with decreased area counts or shifted retention times, caused by leaks or restrictions in the sampler. The hardware has a built-in restriction and pressure decay test that can be utilized to help with troubleshooting. Things to know about the test: Runs two tests successively, the restriction and pressure decay test and the crossport leak test...
  • Replacing a Broken Transfer Line on 7697A Headspace Sampler

    This Information Applies To: Agilent 7697A Headspace Sampler Issue The transfer line will need to be replaced if it breaks at the valve, the inlet or in the middle of the transfer line. The previously broken liner will need to be removed and new one installed. You may experience leak-like symptoms or poor/no chromatography. Items required 3/16- inch open end wrench Column cutting wafer 1/4- inch open end wrench Wrench, angled, septum nut (GC inlet wrench) Polyimide ferrule corrected...
  • Agilent GC Firmware Update Procedure

    This Information Applies To: All Agilent GC systems, Headspace Samplers, and Automatic Liquid Samplers. Issue: Sometimes there maybe a need to upgrade the Agilent GCs, ALSs, or headspace samplers firmware versions to meet the requirement of a software upgrade, Agilent recommendation, or interoperability issues. Steps to follow: Caution : Always use complete downloaded firmware file. Do not interrupt communication, or turn off GC and computer during firmware update process ...
  • Inlet Liners for GC Trace Analyses

    This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet Issue Different liners are required for different splitless trace analysis techniques including: General use liners Pressure, temperature, volume (PTV) Hot inlet High recovery Cold multimode Headspace (HS) and purge and trap (P&T) Solid phase micro extraction (SPME) . Inlet liner choice is part of method validation, and the same liner type must be used consistently while running that method...
  • Troubleshooting Remote/Cluster Partner Missing Error

    This Information Applies To: All Agilent LC systems. Issue: A cluster is partnership between modules. Most modules are linked to other modules automatically, without intervention from the user, but sometimes the cluster is configured manually. Examples: 1. To work as a ‘one module’ unit, independent modules must be clustered (e.g.: Two G1361A prep pumps are clustered to make one gradient pump). 2. A valve pressure cluster between the 1290 Infinity pump and external valves allows for correct...

Checking for External Leaks on 7697A Headspace

Answer

This Information Applies To: Agilent 7697A Headspace Sampler


Issue:

There is a potential leak external to the Headspace unit that can cause loss of pressure.


Steps to Follow:

This article is a troubleshooting guide for when you suspect there is an external leak in the headspace sampler (HS) when it is installed on the inlet of a GC. (Figure 1)

Figure 1: External Leaks can occur in multiple locations: Gas supply bulkhead fittings, Gas cylinder fittings, Regulator fittings, Traps, Shut- off valves (if installed, not included in ship kit), Unions and tee fittings (if installed, not included in ship kit), transfer line connection to the GC

 Warning 
Warning: A pressure decay test cuts off carrier flow through the HS and GC. To prevent damage to the flow path components (column, inlet liner, and so forth), cool the HS and GC before beginning.
  1. Cool the HS sample loop, transfer line, and GC inlet and column oven to prevent damage during the test.
  2. Turn off the HS.
  3. Set the regulator pressure to 415 kPa (60 psi).
  4. Fully turn the regulator pressure adjustment knob counterclockwise to shut the valve.
  5. Wait 5 min.
    1. If there is a measurable drop in pressure, there is a leak in the connections or an internal leak. Utilize a leak detector to locate any external leaks. 
    2. No drop in pressure indicates that the external connections are not leaking 
 Tip 
Learn how to effectively maintain your 7697A Headspace Sampler:
SI-7697-2100sV3 - Agilent 7697A Headspace Sampler Operation and Maintenance e-learning available from Agilent education.

 

 

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help