Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
Articles Error Occurs in Autotune After Filament Switch on Agilent GC/MSD
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC/MS Portal Home
  • GC/MS Portal Articles
  • GC/MS Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC/MS Systems
  • Component :  GC/MS Ion Sources and Samplers GC/MS Single Quadrupole GC/MS Triple Quadrupole
  • Product Name :  5973 Series GC/MSD 5975 Series GC/MSD 5977 Series GC/MSD 7000 Series Triple Quadrupole GC/MS 7010 Series Triple Quadrupole GC/MS CI HES EI HydroInert Inert EI InertPlus EI Stainles Steel EI
  • Task :  Troubleshooting
Related Articles
  • No Emission Current During Tuning on GC/MSD Systems

    This Information Applies To: Agilent GC/MSD systems Issue: When performing MS tuning, there is error message "No Emission Current" and tune is aborted. Steps to Follow This error normally means that the selected filament is not working. Perform the following checks: In the Manual Tune screen, make sure that the PTFBA status is Open . If not, double click to switch it to Open . (See Figure 1 for MSD productivity ChemStation and Figure 2, Figure 3 , and Figure 4 for MassHunter) Click...
  • Replacing the 5977/7010 HES Ion Source Filaments

    This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems Issue The filaments are replaced as part of routine maintenance, or when they break (zero output current). Items Required Gloves, nylon, lint-free, clean (p/n large 8650–0030 , medium G8135-80210 , small 8650–0029 ) Cloth, 100% cotton, lint-free, clean (p/n 05980–60051 ) T6 Torx driver (p/n 8710–2548 ) Filament assembly, high efficiency dual filament (p/n G7002-60001 ) Ion source...
  • How to Disassemble and Assemble the CI Source

    This Information Applies To: Agilent CI Source Issue: During normal operation of the GC/MS, the CI ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filament and ceramic insulators). The ion source will then need to be assembled prior to re-installing. Correct abrasive cleaning, assembly, and installation are essential to robust and reliable operation. Tools Needed: Gloves, clean, lint-free ( 8650-0030...
  • Error Message "The ChemStation did not Complete its Initialization Cycle" when Starting MassHunter GC/MS Acquisition Software

    This Information Applies To: Agilent MassHunter GC/MS Acquisition software Issue The MassHunter software performs several checks during the startup sequence. If it detects a problem, an alert message box will be displayed and the message bar on the bottom left of the software window sometimes shows the valuable clue for error. Resolution The Alert dialog box appears with the message: "The ChemStation did not complete its initialization cycle". Click OK to display the instrument log file...
  • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources

    This Information Applies To: Agilent EI Source Issue: During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments and ceramic insulators). The ion source will then need to be assembled prior to re-installing. Correct abrasive cleaning, assembly, and installation are essential to robust and reliable operation. Tools Needed: • 597X Tool Kit and Cleaning...

Error Occurs in Autotune After Filament Switch on Agilent GC/MSD

Answer

This Information Applies To: Agilent GC/MSD systems


Issue:

After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then tuning the MS, may result in various error messages.

One of the error messages is “Tune action stopped: -310, "({sp00} :rmp:strt;): timeout waiting 1078760 usec for AE to process 1 scan(s)".


Resolution:

This error may be resolved by one or more actions in the following order:

Rebooting / Power Cycling the MS

Resetting the tune file and/or GC/MS instrument configuration

Replacing the filaments

Rebooting / Power Cycling the MS

Option [1] – Reboot from the MS instrument control panel

1. Close the acquisition software.

2. Reboot the MS. 

  • For 5977A, and 5975:

From the MS instrument control panel press Menu several times until Controller is displayed. 
Press Item until Reboot Controller? is displayed.
Press Yes/Select.

  • For 5977B, with LVDS communication enabled, reboot the MS from the GC instrument control panel.

    If LVDS communication is disabled see Option [2] to reboot the MS.

    • For 7890B GC with 5977B MSD, on the 7890B Keypad, press Config then press MS/Aux Det then scroll down and press Request MSD Reboot?  then press On/Yes. 
    • For 8890 GC with 5977B MSD, on the 8890 Touchscreen Display, click Setting > Configuration > Detectors then scroll down and click Reboot. 
    • For 8860 GC with 5977B MSD, on the 8860 Touchscreen Display, click Setting > MSD Info > Reboot.

3. Once the MS reboot completes, re-open the acquisition software and re-tune the MS.

Option [2] – Vent the MS and then power cycle to restart

1. Vent the system, see How To Vent (Shut Down) the GC/MSD - EI - MassHunter

Tip
Tips: To reboot the MS, do not open the Vent Valve to physically vent the vacuum. If the valve is opened, will significantly reduce the time to re-establish a low air and water background in the MS.

2. When the vent is finished, close the acquisition software and power off the MS.

3. Wait 60 seconds and then power on the MS to start a Pump Down of the MS, see How To Pump Down (Start) the GC/MSD - EI - MassHunter.

4. Once the power cycle is completed, start the acquisition software and monitor the progress of the Pump Down.

5. When the High Vacuum is ready, set the MS Temperatures to their usual operating values.

6. When the MS temperatures have reached their setpoints wait at least one hour, and then tune the MS.

Resetting the tune file and/or GC/MS instrument configuration

If rebooting the instrument does not resolve the issue, first follow the instructions from the article: How to Reset the Agilent GC/MSD Tune File to Default Values, then tune the MS.

Replacing the filaments

If the rebooting/power cycling the MS, resetting the tune file, and reconfiguring the instrument does not resolve the issue, then you will need to vent the MS and replace the filaments. It is recommended to also perform an ion source clean at this time.

 

If the error continues, contact your local Agilent support representative.

 

 Tip 
Learn how to effectively operate your software, please see the course below:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning course available from Agilent education.
 