Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
Articles Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC/MS Portal Home
  • GC/MS Portal Articles
  • GC/MS Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC/MS Systems
  • Component :  GC/MS Single Quadrupole
  • Product Name :  5975 Series GC/MSD 5977 Series GC/MSD
  • Task :  Maintenance Operation
Related Articles
  • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD

    This Information Applies To: Agilent 5975 and 5977 series GC/MSD Issue: If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant. Inspect the calibration vial on the back of the flow controller. If necessary, fill the vial with PFDTD according to the instructions in this article. Steps to Follow : Materials needed Wrench, open-end, 1/4-inch × 5/16-inch (p/n 8710–0510 ) PFDTD calibrant, for GC/MS – 1 mL (p/n 8500–8510 ) Syringe...
  • Reviewing the Autotune Report for Agilent 5977 GC/MSD Systems in EI Mode

    This Information Applies To: All Agilent 5977 GC/MSD systems working in EI mode Issue: You can evaluate the 5977 MSD tune performance by reviewing the autotune report. Follow the steps in this article to review MSD tune performance. Steps to follow: 1. Obtain the hardcopy tune report or open the PDF tune report file in the path X/MassHunter/GCMS/1/5977/PDF (see Figure 1 ). Figure 1. PDF tune file in X/MassHuner/GCMS/1/5977/PDF 2. The mass peak profiles for the tune...
  • GC/MSD Autotune Stopped: Gain and Offset Error

    This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems Issue The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow! Resolution The MSD autotune program is run after routine maintenance and at customer set intervals when analyzing samples. See How to run autotune for 5977 series GC/MSD . Possible causes of Tune action stopped error include: The instrument was not assembled correctly after a routine procedure...
  • How to Create, Edit and Run a Sequence in Agilent OpenLab CDS

    This Information Applies To: Agilent OpenLab CDS Software (version 2.x) Issue: This article will give instructions on how to create, save, and run sequences in the Agilent OpenLab CDS Software. Steps to follow: Launch the acquisition software through the OpenLab Control Panel, selecting Instruments in the bottom-left corner menu, then selecting the instrument from the list and clicking Launch . Alternatively, double-click the shortcut on the desktop (if it is created). In the...
  • How to Calibrate the Agilent 7693A Sample Tray

    This Information Applies To: Agilent 7693A Automatic Liquid Sampler Issue: The Automatic Liquid Sampler (ALS) system calibration process aligns the sample tray with the injector turret position to enable vial transfers without incident. Calibration should be performed if a calibration does not exist, as a routine maintenance procedure and is recommended if any ALS components are moved. Do not confuse ALS alignment, with ALS calibration. During alignment, the ALS tray gantry will basically...

Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD

Answer

This Information Applies To: Agilent 5975 and 5977 series GC/MSD


Issue:

Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article explains how to refill the calibration vial.


Steps to follow:

Materials required

  • GC/MS Tuning Standard - PFTBA (p/n 05971-60571)
  • Syringe or pipette

To Refill the Calibrant vial

  1. Stop any tuning or data acquisition.

  2. Turn off the analyzer and the vacuum gauge controller (if installed). (See Figure 1).

    1. In the Tune and Vacuum Control view, under the Execute menu, click MS OFF.

    2. Under the Vacuum menu, click Tune Vacuum Gauge on/off...

    3. Select the off radio button, and click OK.


      Figure 1.
      Turning off vacuum gauge and analyzer

  3. After removing the front analyzer window cover, loosen the calibration vial collar

  4. Without removing the collar, remove the calibration vial. (See Figure 2)



    Figure 2. Removing the EI Calibration Vial

  5. Using a syringe or pipette, fill about 70 µL of PFTBA into the vial (Figure 3). With the vial vertical, the liquid should be just below the end of the internal tube. In the horizontal position, the liquid should not touch the internal tube (Figure 4).


    Figure 3. Filling the Calibration vial


    Figure 4. PFTBA level in the calibration vial.

  6. Push the calibration vial into the vial, as far as possible, and then withdraw the vial 1 mm.

  7. Turn the collar clockwise to tighten it. The collar should be snug but not overly tight.

  8. Reinstall the analyzer window cover (Figure 5).


    Figure 5. Reinstalling the EI Calibration Vial

 Caution 
Caution: After removing a calibration vial, you must purge the calibration valve. Failure to do so may result in damage to the filaments and the electron multiplier.

To purge the calibration valve

  1. If the vacuum gauge controller is on, turn it off.

  2. In Tune and Vacuum Control view, under the Vacuum menu, select Purge Calibrant Valve (Figure 6).


    Figure 6.
    Purging the Calibrant Valve

The EI Calibration will open for several minutes with all analyzer voltages turned off.

 

 Tip 
Learn how to effectively maintain your Agilent 5977B GC/MSD:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning path available from Agilent education.
 