Operating like this puts sub-optimal flow into the GCMS ion source of more than 2 ml/min. Please make sure that you have a system with a turbomolecular high vacuum pump and not a diffusion pump as 2 ml/min exceeds the maximum allowable flow into a diffusion…
Exciting news in the world of FTIR spectroscopy!
Agilent has just released MicroLab 5.8 , MicroLab Pharma 5.8 , and MicroLab Expert 1.3 —the software that powers the Cary 630 FTIR spectrometer and Agilent’s handheld and mobile FTIR systems . MicroLab…
Hi Margot,
It may well be that the column performance has deteriorated so much within the 7 months that the EasiVial standards are measured with poorer PDI values. The column performance can be tested with regular plate count determinations. It is a…
Hi sojeong,
every test report, delivered with aour columns, contains multiple specifications, but they are all measured with specialized systems and software. That means you might not be able to reproduce them identically. BTW for plate count you also…
Hi, I cannot configure the column compartment (CC) as it came integrated with the vial sampler on LC1290. We have the G7129B model for the integrated vial sampler and CC. I tried adding the column compartment details from the instrument configuration…
When the tuning is done at 3200 (+) everything is fine. When I change it to tune at 1700 (+) the masses are reasonably accurate. After it gets to the third mass (620) it appears that the entire spectrum shifts by about 1.5 AMU and the tuning fails. There…
Hi Siddhartha,
Thank you for contacting Cell Analysis Technical Support. Can you please go ahead and send us your System Files so that we can see what might be going on? These can be accessed in the Help Section of the Wave software. You can select…
Hello thank you very much for your response, thanks for the firmware data, the degasser worked perfect, the firmware change no longer, and I can not find the previous version. and are from the 6.80, but how is it possible that a team of thousands of dollars…
Thanks for the reply. I am monitoring at 254 nm. Higher wavelengths do not show that drift. It must be related to the THF (I am using unstabilized) but I do not see how.
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items...
This Information Applies To: Agilent GC Systems with flame ionization detector (FID)
Issue
Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items required...
This Information Applies To: Micro Electron Capture Detector (uECD)
Issue
Baking the micro electron capture detector (µECD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Warning ...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue
This article will describe how an operator can configure the makeup gas on both the instrument keypad and the software driver (Method Editor).
Background...
This Information Applies To: Agilent HPLC systems
Issue
The presence of ghost peaks can be a symptom of a system contamination issue, caused by accumulation of sample and/or foreign substances in the flow path. It is critical for high sensitivity...