This Information Applies To: Agilent HPLC systems
Issue
The presence of ghost peaks can be a symptom of a system contamination issue, caused by accumulation of sample and/or foreign substances in the flow path. It is critical for high sensitivity LC/MS and LC/MS/MS.
In addition to the presence of ghost peaks, some increase in pressure may have been noticed.
Background
One common source of contamination and carry-over of peaks from previous sample runs can be caused by worn or contaminated:
These parts are all on the sample and flow path, as seen on Figure 1:
Figure 1. Flow Path of Sampler in Mainpass
1. Sample 2. From Pump to Injection Valve, 3. Metering device, 4. Sample Loop, 5. Needle, Needle Seat, 6. To Column from Injection Valve, 7. Solvent, 8. Wash pump, 9. Wash Port
Requirements
Disconnect the column from the instrument and replace it with a restriction capillary, then perform a blank run on your LC system.
Confirmation: A blank run is a run with only injecting the solvent that samples are dissolved in.
|
A comparison of column vs. no-column blank runs should identify whether the noted contamination is due to carry over from the column. If the ghost peaks still appear on the blank run, then the contamination comes from the system and not the column.
The key system components should be checked thoroughly – such as the column inlet or the autosampler.
Items Required
When facing a contamination, one of its origins can be found in the autosampler, the following parts should be exchanged in this order:
A full list of the required parts for the particular model of autosampler can be found here.
Tips: This troubleshooting should be carried out with freshly made mobile phase and samples (with new vials). If a needle wash is installed, please flush the wash solvent bottle, clean the wash port, and replace the wash solvent.
|
Steps to Follow
After completing each of the following steps, a blank run should be performed. If the contamination is still present, proceed to the next step.
Tips: See the specific User Manual for instructions on changing the sample loop/rotor seal for your model of Autosampler.
|
If you cannot locate the source of contamination after having been through all the steps, please contact your local Agilent Support representative for further technical support.