Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Bake Out the Agilent Thermal Conductivity Detector
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Systems
  • Component :  GC Detectors
  • Product Name :  6850 GC System 6890 GC System 7820 GC System 7890 GC System 8860 GC System 8890 GC System Intuvo 9000 GC System
  • Task :  Maintenance
Related Articles
  • Capillary Column Installation Quick Reference Guide

    This Information Applies To: All Agilent Gas Chromatographs, 6890, 6850, 7890, 7820, 8890, and 8860. Issue: A quick reference guide on the capillary column installation for the GC. Inlets Detectors Split/splitless inlet Purged packed inlet Multimode Inlet Cool on Column (COC) Inlet PTV Inlet Volatiles Interface (VI) Inlet FID / NPD TCD - Standard Low Leakage Ferrules TCD - Original Design Column Adapter µECD FPD Steps to follow Warning : The...
  • Agilent Thermal Conductivity Detector Troubleshooting

    T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector) Issue Problems observed with thermal conductivity detector (TCD) results can have different underlying causes. Resolution Check that the TCD settings are within the recommended range, then assess the instrument for the following errors/problems: Hardware Fault Error No signal and/or the GC front panel screen displays Front/Back Detector Error. Cause: Filament is...
  • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC

    This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet Issue Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet degrades over time due to accumulated contamination (e.g., baseline noise, ghost / artifact peaks, loss of reproducibility, carry over, loss of compound response or no compound response). Items Required The following Gold seals are available...
  • Agilent GC Detector Sampling (Data) Rate

    This Information Applies To: Agilent GC Systems Issue The data rate determines the number of sampling points across the chromatographic peak and the sample background. The data rate is based on the width of the analyte speak (in minutes) and the ratio of the analyte signal to the background noise. Background The data rate unit is hertz (Hz), where 1 Hz is one cycle per second. The data rate determines the number of data points taken across the width of the peak, and in the background...
  • Bake Out the Agilent Flame Ionization Detector

    This Information Applies To: Agilent GC Systems with flame ionization detector (FID) Issue Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces. Items required Ferrule, no hole 10/pk (p/n 5190-4054 ) Column nut, any of the following: Column nut, universal (p/n 5181-8830 ) Blanking plug, finger-tight style (p/n 5020-8294 ) Finger-tight column nut (p/n 5020-8292 ) Column nut, collared, self...

Bake Out the Agilent Thermal Conductivity Detector

Answer

This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)


Issue

Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces. 


Items Required 

  • Ferrule, no hole (p/n 5190-4054) 
  • Column nut, any of the following:
    • Column nut, universal (p/n 5181-8830)
    • Blanking plug, finger-tight style (p/n 5020-8294)
    • Finger-tight column nut (p/n 5020-8292)
    • Column nut, collared, self-tightening, inlet/detector (p/n G3440-81011) 
  • TCD capillary adapter and fittings (for 7890 and 7820A GC systems)
    • Capillary column adapter (p/n G1532-80540)
    • Nut, 1/8 inch, brass (p/n 5180-4103)
    • Ferrule, 1/8 inch, 15% graphite/85% Vespel (p/n 0100-1332)

Steps to follow

 Caution 
Caution: If the column is not installed, you must turn off the TCD filament and cap the detector column fitting to prevent irreparable damage to the filament caused by oxygen entering the detector.
  1. Load the normal GC method and wait for the GC to become ready. 

  2. Note the value of the FID signal output for comparison with value after the bake-out. 

    • 7890 GC Systems: On the instrument front control panel, press [Front Det] or [Back Det] then scroll to output.
    • 7820A GC Systems: On the Agilent software keypad (remote controller), click [Front Det] or [Back Det] then scroll to output.
    • 8890 GC Systems: On the touch screen or the browser interface, go to Home > Status listing > +Add. Select the detector’s output from the drop-down list, then select Add.
    • 8860 GC Systems: On the browser interface, go to Home > Status listing > +Add. Select the detector’s output from the drop-down list, then select Add.
  3. Determine if you need to remove the column. The detector bake-out temperature is 375 °C and the column temperature is 250 or 25 °C above the normal maximum operating temperature. If the upper temperature for the column is less than the bake-out temperature, remove the column, otherwise, jump directly to step 4:

    1. Turn off the TCD filament
    2. To cool the oven and detector, load the GC maintenance method.
    3. Remove the column from the detector.
    4. Cap the detector connection with the column nut and the no-hole ferrule. For the 7890 and 7820A TCD detectors, make sure that the capillary adapter is installed.  
    5. Turn off the carrier gas and remove the column from the GC inlet and go to step 5.
       Warning 
      Warning: If using hydrogen as a carrier gas, turn off the hydrogen supply and remove the column to prevent an oven explosion.
       
  4. Turn off the TCD filament 
  5. Load the normal operating flows. If the column has not been removed from the GC inlet, maintain inert carrier gas flow through the column.
  6. Set the reference gas flow rate between 20 and 30 mL/min.
  7. Set the temperature zones to the bake out temperatures:

     Caution 
    Caution: Do not exceed the column manufacturer's recommended temperature, as it will damage the column. If the bake out temperature exceeds the recommended temperature, remove the column and turn off the inlet gas flows.
    1. Set the detector temperature to 350 to 375 °C, or 25 °C higher than the normal operating temperature

    2. Set the oven temperature to 250 or 25 °C above the normal maximum operating temperature. Do not exceed the inlet temperature. 

    3. If the column is still installed, set the inlet temperature to 250 or 25 °C above the normal maximum operating temperature. Do not exceed the column maximum temperature. To bake out the GC inlet see How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems .
  8. Hold at the bake out temperature for several hours

  9. Allow the instrument to cool and reinstall the column if appropriate.

  10. Load the normal analytical method and allow the system to equilibrate.

  11. Check the TCD output value. It should be lower than the first reading. Without a column installed, a clean system baseline should be < 20 pA. If it is not, contact your Agilent service center.

 Tip 

Learn how to effectively operate your Agilent GC Detector:
GC-0GEN-1013s - GC Detectors Theory and Operation e-learning course available from Agilent education

Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help