Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: 4150 TapeStation, 4200 TapeStation
Issue
How to deal with a TapeStation Electrode Cartridge change is due warning message.
Background
The Agilent TapeStation controller software will notify users to replace the...
This Information Applies To: Agilent OpenLab CDS
Issue
Sometimes the integration result is not ideal for all injections. This article outlines the step-by-step instruction to manually integrate peaks.
Background
There are 2 ways to integrate...
This Information Applies To: Agilent OpenLab CDS , Agilent OpenLab ChemStation
Issue
Calculation of Area percent or Height percent for uncalibrated compounds in OpenLab software. If all components respond equally in the detector, then Area% and...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
How to power off the Agilent ICP-MS Foreline Pump to perform maintenance, or when fully powering off the Agilent ICP-MS System.
Background
The ICP-MS MassHunter...
This Information Applies To: Agilent ICP-MS (7500,7700,8800)
Issue
How to check the E2M18 or E2M18F foreline pump to determine if the pump needs maintenance?
Background
The oil cleanliness and oil level in the E2M18 or E2M18F foreline pump...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the foreline pump oil in the DS 402.
Background
The DS 402 oil should be replaced at least every 3 months or when the oil color indicates it needs to be changed...
This Information Applies To: Agilent 8890/8860 GC Systems
Issue
How to delete data that is stored on the GC instrument.
Steps to follow
Caution : Before deleting data from your GC, Agilent recommends extraction of data as a backup...
This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer
Issue
When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to check the MS40+ or MS40-S foreline pump to determine if the pump needs maintenance?
Background
The oil's cleanliness and level, in the MS40+ or MS40-S foreline...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Powering on Agilent ICP-MS Systems after maintenance, long-term shutdown, or for power cycling.
Steps to Follow
Confirm that the ICP-MS power plug...