This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer
Issue
When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned, blocked, require extra conditioning, or the electrodes might be encrusted.
Steps to Follow
Check the capillary alignment
Check the alignment of the capillaries in ProSize. Checking the capillary alignment is briefly described in this section, but is described in full in Chapter 9 of the ProSize Data Analysis Software User Manual.
If you have corrected the capillary alignment and you now have data for the missing capillary or capillaries, you have identified the root cause of the problem. Correct the alignment using the instrument controller software to avoid the alignment being incorrect in future data files. Check chapter 4 of the Fragment Analyzer System Manual, Femto Pulse User Manual, or ZAG DNA Analyzer System Manual for instructions. It is recommended and easiest to follow the Capillary Alignment from a File method. The capillary alignment should be done with the new data file, as only this latest run will have the current capillary positions.
If the signal from one or more capillary is missing
First, check for the presence of a blockage. A blockage can be confirmed by running a full conditioning method with a deep-well plate in the waste drawer. The conditioning method requires gel and Conditioning Solution. To start the conditioning method, from the operation tab in the controller software, click Add to queue under the heading Capillary Array - Conditioning. The Select Conditioning Method window will open and the Full conditioning.mthdc can be selected from the drop-down menu (Figure 1). If a capillary is blocked, no waste, or less waste compared to the functioning capillaries, will be present in its corresponding well.
Figure 1. Running a full conditioning method in the controller software.
If a capillary is blocked
Caution: To prevent blocked capillaries, the capillary array should be stored in Capillary Storage Solution. The Capillary Storage Solution should be changed at least once every four weeks. However, it is recommended to change the Capillary Storage Solution every one to two weeks due to evaporation, especially in low humidity/high temperature environments. It is best practice to run a full conditioning run weekly.
|
To unblock the affected capillaries, perform a Method B - hot water soak followed by a Method C - 0.5N NaOH flush. How to perform these methods is described in the appendix of the 5200, 5300, and 5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer system manuals.
During the Method C flush, the NaOH concentration can be increased to 1 N but no further. Flushing the array repeatedly with 1 N NaOH may damage the capillaries and should therefore be avoided if possible. If the Method B and Method C flush do not unblock the affected capillaries, a new capillary array may need to be purchased.
If no capillaries are blocked
The electrode of the affected capillary could be encrusted with, for example, dried Storage Solution, or the capillary might require extra conditioning.
If the Method B and Method C flush and extra conditioning does not restore the signal of affected capillaries, a new capillary array may need to be purchased.
For Research Use Only. Not for use in diagnostic procedures.
PR7001-0840