Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles No Data From One or More Capillary on Parallel Capillary Electrophoresis Instruments
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  15 Mar 2023
  • Modified :  10 May 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  Capillary Arrays Femto Pulse Systems Fragment Analyzer Systems ZAG DNA Analyzer Systems
  • Product Name :  5200 Fragment Analyzer System 5300 Fragment Analyzer System 5400 Fragment Analyzer System Femto Pulse System ZAG DNA Analyzer system
  • Task :  Troubleshooting
Related Articles
  • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II Issue: This article explains the steps to take after installing a brand new array in the Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, and Oligo Pro II systems. Steps to follow Capillary array replacement steps are detailed in chapter 8 of the Fragment Analyzer user manual , Femto Pulse user manual , ZAG DNA analyzer user manual , and chapter 9 of the Oligo Pro II user manual...
  • Agilent Parallel Capillary Electrophoresis Compatible Plasticware

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG Issue: Only plasticware from the approved consumables list should be used in the Agilent Parallel Capillary Electrophoresis systems. The use of plasticware with different dimensions could lead to decreased injection quality and consistency as well as damage to the tips of the instrument capillaries. Resolution: Compatible plates and tubes are listed in the following manuals: 5200, 5300...
  • Insert Disk Error in the Agilent Fragment Analyzer Controller Software

    This Information Applies To: 5200/5300/5400 Fragment Analyzer controller software, Femto Pulse controller software, ZAG DNA Analyzer controller software, Oligo Pro II controller software Issue: The Agilent Fragment Analyzer, Femto Pulse, ZAG, or Oligo Pro II controller software does not open when the desktop icon is double-clicked. Instead, one of the following errors come up: "Insert the 'Fragment Analyzer' disk and click OK" "The feature you are trying to use is on a network resource...
  • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software. Issue Failed to build pressure error comes up shortly after starting a separation. Resolution Failure to build pressure error occurs when the system is trying to build pressure when liquid is being pumped during conditioning or flushing of the instrument. Possible causes and suggestions to address the issue are: 1. Empty reagent...
  • Run Aborted and Target Z Position Error on Parallel Capillary Instruments

    This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Agilent Femto Pulse, Agilent ZAG DNA Analyzer Issue: The instrument encounters a run error, Run Aborted, Position fault: Did not reach target Z position ( Figure 1 ) . This error can occur if the drawer gets opened when the stage is still moving. The error occurs soon after starting the run. Figure 1. Run Error:CtrlTaskStage Error Resolution: System Reboot: Perform an entire system reboot as described in the...

No Data From One or More Capillary on Parallel Capillary Electrophoresis Instruments

Answer

This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer


Issue

When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned, blocked, require extra conditioning, or the electrodes might be encrusted.


Steps to Follow

Check the capillary alignment

Check the alignment of the capillaries in ProSize. Checking the capillary alignment is briefly described in this section, but is described in full in Chapter 9 of the ProSize Data Analysis Software User Manual.

  1.  Click Analysis>View Capillary Positions.
  2. In the View Capillary Positions window, if the alignment is correct, the red dots will be at the top of each blue peak.
  3. If the alignment is not correct but a blue peak is present for each capillary array (e.g there will be 12 peaks for a 12 capillary array), drag and drop the red dots onto the top of the blue peaks.
  4. If there are fewer blue peaks than capillaries in the capillary array, review the subsequent section. 
  5. Once you are happy with the capillary alignment, click Accept to confirm the current capillary positions and return to the main ProSize menu. This will correct the alignment for this data file only.

If you have corrected the capillary alignment and you now have data for the missing capillary or capillaries, you have identified the root cause of the problem. Correct the alignment using the instrument controller software to avoid the alignment being incorrect in future data files. Check chapter 4 of the Fragment Analyzer System Manual, Femto Pulse User Manual, or ZAG DNA Analyzer System Manual for instructions. It is recommended and easiest to follow the Capillary Alignment from a File method. The capillary alignment should be done with the new data file, as only this latest run will have the current capillary positions. 

If the signal from one or more capillary is missing

First, check for the presence of a blockage. A blockage can be confirmed by running a full conditioning method with a deep-well plate in the waste drawer. The conditioning method requires gel and Conditioning Solution. To start the conditioning method, from the operation tab in the controller software, click Add to queue under the heading Capillary Array - Conditioning. The Select Conditioning Method window will open and the Full conditioning.mthdc can be selected from the drop-down menu (Figure 1). If a capillary is blocked, no waste, or less waste compared to the functioning capillaries, will be present in its corresponding well.

 Figure 1

Figure 1. Running a full conditioning method in the controller software.

If a capillary is blocked

  
Caution: To prevent blocked capillaries, the capillary array should be stored in Capillary Storage Solution. The Capillary Storage Solution should be changed at least once every four weeks. However, it is recommended to change the Capillary Storage Solution every one to two weeks due to evaporation, especially in low humidity/high temperature environments. It is best practice to run a full conditioning run weekly.

To unblock the affected capillaries, perform a Method B - hot water soak followed by a Method C - 0.5N NaOH flush. How to perform these methods is described in the appendix of the 5200, 5300, and 5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer system manuals.

 During the Method C flush, the NaOH concentration can be increased to 1 N but no further. Flushing the array repeatedly with 1 N NaOH may damage the capillaries and should therefore be avoided if possible. If the Method B and Method C flush do not unblock the affected capillaries, a new capillary array may need to be purchased.

If no capillaries are blocked

The electrode of the affected capillary could be encrusted with, for example, dried Storage Solution, or the capillary might require extra conditioning.

  1. To remove any encrusted solution, perform a Method B - hot water soak followed by a Method C - 0.5N NaOH flush. These methods are described in the appendix of the 5200, 5300, and 5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer system manuals.
  2. Before the next run, run a full conditioning method. 

If the Method B and Method C flush and extra conditioning does not restore the signal of affected capillaries, a new capillary array may need to be purchased.

For Research Use Only. Not for use in diagnostic procedures.

PR7001-0840

 Tip 
Learn how to effectively maintain and operate your parallel capillary electrophoresis instrument:
5200, 5300, and 5400 Fragment Analyzer System Manual on Agilent.com
Femto Pulse User Manual on Agilent.com
ZAG DNA Analyzer System Manual on Agilent.com
ProSize Data Analysis Software User Manual on Agilent.com

 

Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help