This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
How to power off the Agilent ICP-MS Foreline Pump to perform maintenance, or when fully powering off the Agilent ICP-MS System.
Background
The ICP-MS MassHunter Software controls the pump’s ON/OFF control, which will transition the ICP-MS System from standby mode to shutdown mode.
If software communication is not possible, it is possible to shift from standby to shutdown mode by pressing the vacuum ON/OFF switch on the ICP-MS. The switch is located under the ICP-MS large cover.
Steps to Follow
3a. For ICP-MS MassHunter 4.6 and below: Click the drop-down arrow next to the Hardware button and select Vacuum Off (Figure 1)
Figure 1. Vacuum Off for MassHunter 4.6 and below
3b. For MassHunter 5.1 and above: Click the Dashboard tab on the left, then click on the Mainframe image. Select Vacuum from the menu (Figure 2)
Figure 2.Vacuum Off for MassHunter 5.1 and above
3c. Using the switch located under the ICP-MS large cover
|
Warning: When the plasma is lit, do not open the top covers. Confirm that the ICP-MS is in standby mode before continuing.
|
Figure 3. Front View of ICP-MS
1. Large Cover, 2. Manual Vacuum ON/OFF Switch, 3. LED Status Indicator
Figure 4. Top Down View of ICP-MS with Large Cover Open
1. Manual Vacuum ON/OFF Switch
Figure 5. Rear View of ICP-MS
1. Foreline Pump Breaker Switch. 2. Main Power Breaker Switch
The Foreline Pump is now completely powered off and ready for maintenance. See the instructions to Power On the ICP-MS Foreline Pump.
Learn how to effectively troubleshoot and maintain your Agilent ICP-MS -
Agilent 7800/7850/7900 ICP-MS with MassHunter e-learning path available from Agilent education |