This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to check the MS40+ or MS40-S foreline pump to determine if the pump needs maintenance?
Background
The oil's cleanliness and level, in the MS40+ or MS40-S foreline pump, should be checked monthly. Perform maintenance as needed, based off the monthly check. The oil should be replaced at least once every six months, but the type of samples and number of samples analyzed may shorten the life of the oil; adjust the oil change frequency accordingly.
|
Warning: Never add pump oil while the foreline pump is on.
|
|
Caution: Only use Agilent Recommended pump oil, appropriate for the MS40+ or MS40-S foreline pump. Using a different oil can substantially reduce pump life and invalidates the pump warranty.
|
Steps to Follow
Figure 1. Checking the Oil Color
Oil cleanliness Chart: 1. Ok (yellow), 2. Replace soon (orange), 3, Replace immediately (brown)
Tips: Since the oil level changes depending on where the vane stops inside the pump, make sure you check the oil level during pump operation (vacuum on).
|
Figure 2. Checking the Oil Level, desired upper and lower ranges indicated with red lines and red arrow.
Learn how to effectively maintain your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter e-learning course available from Agilent education |