Good day . i have another issue to my 8890. after replacing the fused silica in transfer line. tha air flow in FID is suddenly shutting down. then i try the detector diagnostic in the system it said test failed and need to replace the jet.
after replacing…
Hello,
this look like your plasma is not align with the cones. For the EM to failed it's difficult. You have to check for noise when it's doing it and if the counts are low when it's using mass 80. You can try to tune EM with mass 37 if you know where…
Try to ignite the plasma, without fixing the shield plate. Due to poor quality of argon gas or leak on gas line, reflected power will be above the limit and prevent the plasma sustain. Sometimes, the plasma sensor fiber optics tip is damaged, needs to…
Hello,
I am currently developing a residual solvent analysis method by Gas chromatography equipped with a flame ionization detector. My sample is an API with a phosphate salt attached. One of the residual solvents of interest is triethylamine. When…
8/7/19
Hello,
I have had a fair amount of success with posted questions in this forum, so here is another one. I am interested in knowing more about whether the flame ionization detector I have equipped on my HP/Agilent 6890 GC is the correct detector…
We have a lot of 1260s and 1290s in our lab, and as part of our sustainability initiatives are looking for ways to reduce our energy usage. Currently, all the instruments are left on (standby), and most of them also have the lamps on when not in use.…
When I need information about detectors I use the following books:
Raymond P. W. Scott "Chromatographic Detectors - Design, Function, and Operation", Chromatographic Science Series Volume 73, Marcel Dekker, New York (1996). ISBN 0-8247-9779-5.
Raymond…
Hi Guys,
I have a GC that does not seperate natural gas(maily C1 and C2 with small traces of C3+), it comes as almost one peak, and for condensate (C3+) it seperates C6+, and does not seperate C3÷C5,
I Have,
1: Readjusted the split ratio Many Many…
Hello Inês,
Typically we use the Performance report as a general health report of the instrument. Changing plasma parameters may cause the test to fail, unless you correct the limits. Moving the sample depth will decrease sensitivity and can increase…
Hello
I'm trying to setup report to have values in column "PASS/FAIL" green or red depends on result.
Can I do it as conditional formatting for column "PASS/FAIL" to mark FAIL as red based on text?
This column "PASS/FAIL" is custom field and…
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system run ends prematurely and the following error appears in the logbook: Run aborted on port x. Instrument error occurred on port x, Unusual...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system electropherograms in the Agilent 2100 Expert software have a wavy, uneven baseline ( Figure 1 ).
Figure 1. Wavy, uneven baseline on Bioanalyzer...
This Information Applies To: Agilent AriaMx systems.
Issue
When IP addresses of an AriaMx instrument are reassigned, it will appear under the new and the previous IP address in the AriaMx Instrument Explorer. This situation can be prevented using...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Sometimes, for support purposes, it is necessary to send the diagnostic test result file to Agilent support representatives.
This file contains detailed information about the...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx Real-Time PCR software
Issue
How to change the discovery protocol port in the AriaMx software.
Steps to Follow
The AriaMx software uses the UDP Broadcast protocol for...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The instrument diagnostic tests should be run whenever it is suspected that there could be a fault with the instrument hardware.
Background
The instrument diagnostics consist...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Error 45 is displayed on the AriaMx instrument touch screen.
Background
Error 45 is shown when the temperature recorded by one or more thermistors is outside the expected...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx showing a series of errors after delivery, guide for troubleshooting and resolving temperature-related error.
Background
When an AriaMx is delivered, it could take...
How to Access Manual Tune in GC/MS MassHunter and MSD Productivity ChemStation
Multiple Results
This Information Applies To: MassHunter acquisition software for GC/MS systems and MSD Productivity ChemStation for all Agilent GC/MS systems
Issue
Manual tuning lets you interactively set the MSD parameters, such as lens voltages and tuning masses...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to find the AriaMx software and firmware version that you have installed.
Steps to Follow
There are two different ways to find out AriaMx software version:
In the...