This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Error 45 is displayed on the AriaMx instrument touch screen.
Background
Error 45 is shown when the temperature recorded by one or more thermistors is outside the expected range. The operating temperature range for the AriaMx is 20 to 30 °C.
Steps to Follow
Make sure that the instrument is installed in a room within the operating environment temperature range (20 to 30 °C). If the temperature of the room is too hot or cold, adjust the room temperature to within the required range. If you would prefer to move the AriaMx to a different location that is within the correct temperature range.
Tip: The performance of the AriaMx could be adversely affected by bumping it hard or shaking it vigorously. Moving the instrument between labs (on the same floor) can be handled on a sturdy cart. For longer moves, contact your local Agilent sales and support center for assistance in moving the Aria and assuring proper operation afterwards.
|
The temperature sensors can be sensitive to environmental conditions, so make sure that the instrument is not in direct sunlight, by an open window, near a heater or an air conditioning vent.
Check the block temperature shown in the top left of touch screen of the AriaMx (Figure 1) to verify that the instrument is operating within the indicated temperature range.
Figure 1. AriaMx touch screen with the block temperature in the top-left corner highlighted with a yellow square.
Tip: To run the diagnostic check, log in as an administrator account (the default password for the Administrator user name is ADMIN). On the instrument touch screen, select Settings > Instrument Diagnostics. Deselect the observational test section as these tests are not required to remove error 45. Leave all other tests selected. Select Run. This is also described in chapter 2 of the AriaMx Real-Time PCR System Setup and User Guide. |
Root cause