Hello APacana ,
The reason for the difference in response between the TIC and the EIC is because when you extract the TIC of a given precursor you are extracting the ion signal from all precursors that create a 72 product ion, using your example. If…
Hello AJohnson13 ,
Here is the PC setup guide for older Windows 10 versions of MassHunter. It includes the PC specifications.
Windows10_Setup.book (agilent.com)
The software installation instructions would be included as part of your original media…
At 1 mL/min pumping water, the pressures are as follows:
Pump + Column = 53 bar
Pump + Column + 0.1 mm ID 30 cm tubing = 70 bar
Pump + Column + Tubing + Flow Cell = 75 bar
I have also confirmed that this pressure fluctuation has nothing to do…
It all depends on your following steps. You can even use Methanol or Acetonitrile for the protein precipition: Protein Precipitation for Biological Fluid Samples Using Captiva EMR-Lipid
See also the chapter on protein precipitation in Agilent's Handbook…
The Digest of upcoming webinars has been updated as of April 30.
Did you miss April's GC Live Session Recording and Discussion Thread? Check it out here: ( +) Diagnostics From Your Desk, Troubleshooting From Your Kitchen Table - Follow-up Discussion…
Hello cjwwfd ,
B.06.00 is out of support at this point, but I can offer some general suggestions for current revisions and hopefully you can apply them to your version. Version 12.0 is the current release. If possible, I would investigate your options…
I'm experimenting the same error. Logbook displays <MSQQQ_1: QqQ fault detected: 1.23 vacuum system is not ready. This is typical during a pump down cycle or when the high vacuum pump is turned off during a vent cycle or fault condition>. Does anyone…
Hello DET ,
Sorry for the late response on this.
What sort of spiking do you see usually? What is the output of the FID, in pA?
Are you using any UPS systems, or are the GCs plugged directly into the wall?
Hi, I think it would be a good idea to upload your startup tune report in order to help you better.
Also I think you could take advantage of reading this post: https://community.agilent.com/technical/atomic-spec/f/forum/470/icp-ms-manual-tuning
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system run ends prematurely and the following error appears in the logbook: Run aborted on port x. Instrument error occurred on port x, Unusual...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system electropherograms in the Agilent 2100 Expert software have a wavy, uneven baseline ( Figure 1 ).
Figure 1. Wavy, uneven baseline on Bioanalyzer...
This Information Applies To: Agilent AriaMx systems.
Issue
When IP addresses of an AriaMx instrument are reassigned, it will appear under the new and the previous IP address in the AriaMx Instrument Explorer. This situation can be prevented using...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Sometimes, for support purposes, it is necessary to send the diagnostic test result file to Agilent support representatives.
This file contains detailed information about the...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx Real-Time PCR software
Issue
How to change the discovery protocol port in the AriaMx software.
Steps to Follow
The AriaMx software uses the UDP Broadcast protocol for...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The instrument diagnostic tests should be run whenever it is suspected that there could be a fault with the instrument hardware.
Background
The instrument diagnostics consist...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Error 45 is displayed on the AriaMx instrument touch screen.
Background
Error 45 is shown when the temperature recorded by one or more thermistors is outside the expected...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx showing a series of errors after delivery, guide for troubleshooting and resolving temperature-related error.
Background
When an AriaMx is delivered, it could take...
How to Access Manual Tune in GC/MS MassHunter and MSD Productivity ChemStation
Multiple Results
This Information Applies To: MassHunter acquisition software for GC/MS systems and MSD Productivity ChemStation for all Agilent GC/MS systems
Issue
Manual tuning lets you interactively set the MSD parameters, such as lens voltages and tuning masses...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to find the AriaMx software and firmware version that you have installed.
Steps to Follow
There are two different ways to find out AriaMx software version:
In the...