MassHunter Qual TIC and EIC Area Differences

Hi all,

I've recently encountered an issue in MassHunter Qual v.10 when creating a custom workflow. I have added "extract additional chromatograms" to the workflow and noticed that the peak area it outputs is different based on whether I am extracting a TIC or an EIC. For reference, I am running on our 6475 QQQ, and acquiring my data in precursor ion scan mode. I monitor 2 primary product ions that are characteristic of the compounds I am studying (nitazenes), which are m/z 100 and m/z 72. An example of the values I received for one of my compounds, metodesnitazene, extracted from the same data file:

m/z 100 peak area extracted with TIC: 3,658,123

m/z 100 peak area extracted with EIC: 2,418,212

Obviously those are completely different values but there's nothing I can find that would make sense as to why they are so different. Additionally, when looking at the spectra, the abundance of the precursor ion for m/z 100 is the same for both processing methods, but the abundance of m/z 72 is different. I have attached screenshots of the "additional chromatograms" section of the method editor for reference as well (note that the EIC method only includes the extraction for one of my compounds as I was just testing the changes). Any help about this is greatly appreciated since I am hesitant to process/collect more data until I know where these numbers are being pulled from.

Parents
  • Hello  ,

    The reason for the difference in response between the TIC and the EIC is because when you extract the TIC of a given precursor you are extracting the ion signal from all precursors that create a 72 product ion, using your example. If you extract an EIC of 338->72 then you are looking at a chromatogram made up of only the signal due to a specific precursor (338) to a specific product ion (72). It is essentially an MRM trace, though not as sensitive of course. 

    I think it is easier to conceptualize if you look at the trace labels for the chromatograms themselves. This is just an example using background signal, but even so you can see the difference, and the trace labels, (** -> 200.0) vs (375.0 -> 200.0) makes it easier to explain.

Reply
  • Hello  ,

    The reason for the difference in response between the TIC and the EIC is because when you extract the TIC of a given precursor you are extracting the ion signal from all precursors that create a 72 product ion, using your example. If you extract an EIC of 338->72 then you are looking at a chromatogram made up of only the signal due to a specific precursor (338) to a specific product ion (72). It is essentially an MRM trace, though not as sensitive of course. 

    I think it is easier to conceptualize if you look at the trace labels for the chromatograms themselves. This is just an example using background signal, but even so you can see the difference, and the trace labels, (** -> 200.0) vs (375.0 -> 200.0) makes it easier to explain.