When you run startup, does it use the Standard Tune or User Tune? In the new D rev of software (5.X) you can go to Startup/Startup Configuration and just above the Hardware settings table select Configure ignition sequence. For C rev (4.X) click the narrow…
I have Open Lab CDS Chemstation Edition for GC systems and 7890A with FID. I get in vial concentration of an analyte in a sample from calibration curve as ppm. Need to generate an Intelligent report, I get concentration in sample (with Sample weight,…
I am using Agilent 8890. I am trying to set the standard for my GC-FID, I am injecting 5mL of methane gas standard with a gas-tight syringe through a switching valve which is set up to the left of my GC. However, I am getting different peaks and areas…
Good Afternoon,
I'm developing a processing method using Mass Hunter Quant. 10.2 for ethanol (quantitate) and other related volatiles (ID only). The instrument setup is a single column that splits to two detectors FID and MS (FID for quant purposes…
Hello taub .
I think it's a corner case issue and you need a support visit from agilent authorized engineer so please request a service visit as its seems that the error is hidden and not straight forward as you mentioned that you changed alot of things…
Hi
I am trying to install the G7201D version of ICP MassHunter on a 1909 Windows 10 laptop.
It goes through the initial process and then errors with a code of 0xe0434352.
Some additional research implies this is a .Net error but a repair install…
Hello Guys, I am using Agilent 7890a GC with an FID detector. I using this GC in online mode to analyze reactor exit gases. I am currently studying the dehydrogenation of Methylcyclohexane over Pt catalyst. Recently I noticed that the FID response is…
First off, it is not the barcode reading that causes the vial leaking, but my best guess is that the barcode reading introduces a variable time, that then can cause a scheduling conflict. This scheduling conflict is then not handled, and the 7697A is…
I am new to GC. I am trying to make a calibration table for a set of fatty acid standards using Agilent 8860 (GC-FID)
After the derivitization step, analysis of the resulting methyl esters shows only the solvent peak. Usually a very high solvent peak…
I have a MicroGC 990. Seemingly out of no where, the following error has popped up after hitting Control -> Single Run on a method we've been using for months:
Failed to submit "D:\ .... path name .... (GMT -04-00).seq" for acquisition.
Please resubmit…
This Information Applies To: Agilent MassHunter Qualitative Analysis Software
Issue
Error message prompted “Access to the path C:\Users\AgilentMad\AppData\Local\Temp\FILESPEC.TXT is denied" during search using NIST MS Program in MassHunter Qualitative...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
When sending a run configuration from the AriaMx PC software, no run configuration is transmitted. The expected blue confirmation icon does not appear at the bottom of the touch...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
Can the AriaMx software be used on Windows 11?
Resolution
The AriaMx a nalysis software 2.0 has not been validated for Windows 11, so Agilent...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to control more than one AriaMx with one PC.
Resolution
If you connect the PC and instruments using an ethernet cable, you will not be able to establish multiple connections...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
In the Export Data menu, after exporting data in a LIMS format, the output file's name is incomplete.
Background
An example of everything after...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
The AriaMx Data saving behavior when priming experiments from the PC is dependent upon whether the user is logged in or a Guest and whether the user...
This Information Applies To: Agilent OpenLab CDS
Issue
Sequence Creation Template allows users to create a sequence template that identifies the order of injections or actions in the sequence.
This article outlines the step-by-step instruction...
This Information Applies To: Agilent 6100 and 6400 Series LC/MS Systems
Issue
Tuning a mass spectrometer involves optimizing the voltages applied to various optical elements in the ion path to achieve the required mass accuracy, resolution, and...
This Information Applies To: Agilent OpenLab ChemStation
Issue
In OpenLab ChemStation, creating a result set by combining multiple existing result sets can be useful in the following scenarios:
Combining single samples, sequences, or a combination...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The AriaMx analysis software shows an error when attempting to open a data file.
Background
The AriaMx analysis software can show several errors when something is wrong...