This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
In the Export Data menu, after exporting data in a LIMS format, the output file's name is incomplete.
Background
An example of everything after the full stop in the file name being cropped can be seen in Figure 1.
Figure 1. Exported data where the output's name is truncated
Resolution
The issue comes from the . in the .amxd file name, for example: Test 1.2 Quantitative PCR DNA Binding Dye with Melt [Sample] (Figure 2).
Figure 2. Root cause of the cropped file name. The full stop causing the truncation is highlighted in yellow.
Once the full stop is removed, the LIMS file name is no longer cropped upon export (Figure 3).
Figure 3. Data exported without truncation of the file name.
Learn how to effectively operate your Agilent AriaMx:
AriaMx Real-Time PCR System Setup and User Guide on Agilent.com Aria Real-Time PCR Software User Manual on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.