This Information Applies To: Agilent 6100 and 6400 Series LC/MS Systems
Issue
Tuning a mass spectrometer involves optimizing the voltages applied to various optical elements in the ion path to achieve the required mass accuracy, resolution, and sensitivity that the user requires for their chosen application.
What are the differences between autotune, check tune, and manual tune and when is the best time to run them?
Background
Tuning involves the use of a calibrant that consists of a set of perfluorinated compounds with known ion masses across a broad mass range (100-3000 m/z). Calibrant is infused into the MS during the tuning operation to optimize voltages for optimal transmission, expected m/z assignment, spectral peakwidth (m/z resolution), and detection using these known ion masses as a reference point.
For more details on calibrant/tuning mix, check out the following articles:
Resolution
Autotune
Checktune
Manual tune
Where to perform tuning in different versions of software?
Select the software version that you are using from the list below for software-specific instructions.
MassHunter Workstation Data Acquisition Version 10 and older versions
MassHunter Workstation Data Acquisition version 12
OpenLab ChemStation Edition Online Session
MassHunter Workstation Data Acquisition Version 10 and older versions
Figure 1. Autotune and Checktune in MassHunter software
1. Tune context, 2. Autotune tab, 3. Autotune, 4. Checktune
Figure 2. Manual Tune in MassHunter software
MassHunter Workstation Data Acquisition version 12
Autotune
See Figure 3 for the location of each step in the Method Editor screen.
Checktune
Figure 3. Autotune/Checktune window in MassHunter Data Acquisition version 12
The numbers on figure 3 refer to steps listed in the Autotune section.
Advanced override settings (replaces old Manual Tune)
In MassHunter version 12, Manual Tune is replaced by Advanced override settings enabling some additional Tune parameters to be customized depending on the LC/TQ model being controlled.
Figure 4. Advanced Override setpoints window in MassHunter Data Acquisition version 12
1. TQ tab, 2. Advanced Override settings, 3. Parameter and input override value.
OpenLab CDS Acquisition
Figure 5. Autotune and Checktune in OpenLab CDS software
1. Method layout, 2. Autotune tab, 3. Autotune, 4. Checktune
Figure 6. Manual Tune in OpenLab CDS software
OpenLab ChemStation Edition Online Session
Figure 7. Autotune, Checktune and Manual Tune in OpenLab ChemStation software
1. MSD Tune tab, 2.Tune drop-down menu, 3.Check Tune, 4. Autotune, 5. Manual Tune
When to Run Checktune vs Autotune
Checktune | Autotune |
Uses current tune file but does not change it | Generates new tune file |
Does not update EMV | Updates EMV |
Takes 3–6 minutes per polarity | Takes 25–30 minutes per polarity |
To verify MS is OK before starting the worklist | At the initial installation of the MS Anytime MS has been vented for cleaning or repaired When checktune recommends it |
Recommended daily or weekly depending on instrument use or the lab’s SOP | Recommended monthly or quarterly depending on instrument use or the lab’s SOP |
Learn how to effectively operate your Agilent 6100 and 6400 Series LC/MS Systems:
LCMS-6400-1120sV2 - Tune and Calibrate Agilent 6400 Series Triple Quadrupole LC/MS with MassHunter Acquisition (Ver. 10.0) LCMS-6100-1101e - Tuning Agilent Single Quadrupole LC/MSD Systems in OpenLab CDS LCMS-6100-1102e - Tuning Agilent Single Quadrupole LC/MSD iQ System in OpenLab CDS eLearning courses available from Agilent education |