Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent MassHunter Qualitative Analysis Software
Issue
Error message prompted “Access to the path C:\Users\AgilentMad\AppData\Local\Temp\FILESPEC.TXT is denied" during search using NIST MS Program in MassHunter Qualitative...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
When sending a run configuration from the AriaMx PC software, no run configuration is transmitted. The expected blue confirmation icon does not appear at the bottom of the touch...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
Can the AriaMx software be used on Windows 11?
Resolution
The AriaMx a nalysis software 2.0 has not been validated for Windows 11, so Agilent...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to control more than one AriaMx with one PC.
Resolution
If you connect the PC and instruments using an ethernet cable, you will not be able to establish multiple connections...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
In the Export Data menu, after exporting data in a LIMS format, the output file's name is incomplete.
Background
An example of everything after...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
The AriaMx Data saving behavior when priming experiments from the PC is dependent upon whether the user is logged in or a Guest and whether the user...
This Information Applies To: Agilent OpenLab CDS
Issue
Sequence Creation Template allows users to create a sequence template that identifies the order of injections or actions in the sequence.
This article outlines the step-by-step instruction...
This Information Applies To: Agilent 6100 and 6400 Series LC/MS Systems
Issue
Tuning a mass spectrometer involves optimizing the voltages applied to various optical elements in the ion path to achieve the required mass accuracy, resolution, and...
This Information Applies To: Agilent OpenLab ChemStation
Issue
In OpenLab ChemStation, creating a result set by combining multiple existing result sets can be useful in the following scenarios:
Combining single samples, sequences, or a combination...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The AriaMx analysis software shows an error when attempting to open a data file.
Background
The AriaMx analysis software can show several errors when something is wrong...