Good morning,
We have an Agilent 7820A GC-FID equipped with a G4567A autosampler.
In the middle of a sequence, the system gave the following errors:
ALS Failure State: Preparing Samples
ALS Failure Code 945 : Autoinjector back tower plunger…
The Digest of upcoming webinars has been updated as of December 13.
Notable additions include:
Atomic (4)
GC & GCMS (9)
Vacuum Solutions Series (1)
Check Out:
New Year resolution for my ICP-MS: I will keep my ICP-MS analysis free from…
We have experienced this but on a 5975. The cooling fan for the turbo pump failed to restart after maintenance. The error message however did not refer to the fan at all. The easiest way to check if the fan is working is to hold a Kimwipe or other tissue…
Good morning It turns out that this is the first time something so particular has happened to me with my GC-MSD (6850-5975). I work with MassHunter 10.0.368 and from one moment to another the signal for two of my work standards (organophosphates and phenols…
Good morning, everybody!
I am currently doing untargeted analysis using the Unknown analysis software. I could «identify» some compounds of interest, which is great. Now, my next step is to do a semi-quantification of my compounds using my internal…
Hello everybody,
I operate an Agilent 7890B-GC/5977A-MSD and a 7820A-GC-FID, both with autosampler.
Those systems are yearly maintained by Agilent.
So far, GC-MS is for qualitative analysis (compounds identification through library search for…
I looked at your plot again. That little bounce is barely 0.01pA -- so very tiny! This is down in the lowest possible signals that can be measured with an FID and honestly should not matter as your smallest peaks of interest should be at least 50x to…
Hello jeremy_lai ,
In your expressions you must check for both the existence of the qualifier and the outlier yourself. In the initial if and in the final one you need to add a check to see if there is a qualifier by checking to see if it is None. …
We run a lot of our GCs with Polyarcs (methanizers) and FID responses can be theoretically calculated. While upgrading the software and importing chemstation methods, the calibration table did not get imported. If I could manually edit the calibration…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...