This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed regularly as part of your routine maintenance schedule.
Items Required
Steps to Follow
|
Warning: Surfaces exposed to samples may be contaminated, and the inlet is set at high temperatures. Wear appropriate PPE when handling contaminated and/or hot surfaces.
|
|
Caution: Wear clean, lint free gloves when handling the inlet liner, O-ring, and cleaned inlet surfaces to prevent contamination with dirt, fibers, and skin oils.
|
Figure 3: Inlet liner installation direction:
1. Top of Inlet 2. Bottom of inlet 3. The exception with flow direction is printed on liner.
Figure 4: Install the inlet liner
Tips: The liner O-ring will slide up the liner as the liner is pushed down into the inlet ensuring a correct seal.
|
|
Caution: Insufficient purging time can lead to residual oxygen (air) remaining in the column. Damage to the stationary phase of the column may occur if the column is heated with oxygen present.
|
Learn how to effectively maintain your Agilent GC system:
Get Started with your New Agilent GC system eLearning path available from Agilent education |