Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Calculating the Headspace Sampler GC Cycle Time
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  6 Sep 2023
  • Modified :  10 Nov 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Sample Preparation and Introduction
  • Component :  Headspace Samplers
  • Product Name :  7694 Headspace Sampler 7697A Headspace Sampler 8697 Headspace Sampler G1888A Headspace Sampler
  • Task :  Operation
Related Articles
  • How to Fix "Data system not ready" on 7697A Headspace Sampler

    This Information Applies To: Agilent 7697A headspace sampler Issue During sequence running, there is error message: "Sequence aborted: Data System Not Ready" displayed on the 7697A display screen. Resolution Check GC cycle time. If the GC cycle time is set too short in the GC Method the sequence will abort before the scheduled run time is reached. See the article How to calculate GC Cycle Time for use with Headspace for further detail. Check and ensure that the remote start cable...
  • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler

    This Information Applies To: Agilent 7697A headspace sampler. Issue: The headspace sampler may present a flow shutdown or poor chromatography with decreased area counts or shifted retention times, caused by leaks or restrictions in the sampler. The hardware has a built-in restriction and pressure decay test that can be utilized to help with troubleshooting. Things to know about the test: Runs two tests successively, the restriction and pressure decay test and the crossport leak test...
  • Replacing and Adjusting the Agilent ICP-MS System Peristaltic Pump Tubing

    This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems Issue Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened or does not require stretching to fit over the pump rollers, it should be replaced with new pump tubing. Background Agilent pump tubing, PVC tubing and connector barb sizes vary with application. Some sized tubing will not require a connector...
  • How Does the AriaMx Warm up Feature Work?

    This Information Applies To: Agilent AriaMx, Software version 1.8 and later Issue The Aria Warm up Feature was introduced with Aria SW/FW 1.8 to reduce baseline drift. The warm-up feature is used to speed up the heating process that would normally take up to 3 hours after the instrument is turned on. Background Sometimes, the baseline fluorescence signals gradually decrease across the baseline cycle range (see Figure 1). Warming up the optical modules may help signals remain steady throughout...
  • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule

    This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900) Issue: The ICP-OES and AVS 6/7 components require regular routine maintenance. Agilent-authorized field service engineers must perform any maintenance procedures not listed in this document. Warning : Shock Hazard. Do not remove instrument covers that expose electrical circuits, devices, and components operating at dangerous voltages. Contact with these circuits, devices, and components can result in...

Calculating the Headspace Sampler GC Cycle Time

Answer

This Information Applies To: Agilent headspace samplers (HS)


Issue

GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause reduced sample throughput as vials wait longer than needed before processing.


Background

GC cycle time is the total time required for the GC (or GC/MS) system to return to a ready state after a sample run. HS uses the GC cycle time to calculate throughput and timing. An accurate GC cycle time is crucial to optimize throughput and for correctly processing samples. 

  
Tips: It is better to enter a longer time than needed than to enter too short a time and possibly reduce sample quality.

Steps to follow

  • Manual calculation
  • For Agilent GCHS systems controlled by Agilent software (Note: Applicable to OpenLab CDS 2.x, OpenLab ChemStation C.01.X, and MassHunter Acquisition only)

Manual calculation

1. Check GC Run time.

2. Check GC Oven equilibration time. The default value is 3 minutes.

3. Check GC Cooling time. The cooling time is the time to go down to the original temperature after your analysis is over. It is recommended to calculate this parameter manually.

4. If needed, check GC prep run time and inlet cooling time.  However, these parameters are not commonly used and should be added only when necessary.

5. Next, add all timings consolidated from Step 1 to 4. 

6. Finally, with the total time calculated, do an addition of 3~5 minutes. The reason for adding this time is that GC is highly influenced by the surrounding environment, so there may be variables. Therefore, this addition of time is required for that variable.

   Caution: This calculation can be flexible because it does not always come out with the same value. The time may increase or decrease depending on the laboratory environment or changes in ambient temperature.

 

Follow these steps for Agilent GC/HS systems controlled by the following software products:

  • OpenLab CDS 2.x 
  • OpenLab ChemStation C.01.X
  • MassHunter Acquisition

To track GC cycle time via OpenLab CDS 2.x

  1. To edit the headspace sampler configuration, click Control > Configuration. See Figure 1.
  2. The Configuration-Agilent 7697A window will be displayed, click the Preference Tab. 
  3. Make sure to tick the Track GC cycle time check box under Method Development. 
  4. Click OK to set the HS configuration. 
  5. Run a sequence of at least two samples with and GC cycle time setting of at least 5 minutes greater that the GC oven run time plus the time needed for the oven to return the initial setpoint.
  6. When the sequence is completed, enable the Activity Log in Acquisition. Scroll down the activity list and find the Calculated GC cycle time. See Figure 2.


Openlab CDS Headspace sampler configuration page. 
Figure 1 OpenLab CDS headspace sampler configuration page
1.Control menu, 2. Configuration tab, 3. Preferences tab, 4. Option to tick Track GC cycle time, 5. OK option 


Activity log under Home of Openlab CDS. 
Figure 2. Activity log under Home of OpenLab CDS
1.Home tab, 2. Activity log option, 3. Calculated GC cycle time shown in Activity log

To track GC cycle time via OpenLab ChemStation 

  1. In the instrument control view, click Instrument > Agilent 7697A Configuration. See Figure 3.
  2. The Agilent 7697A Configuration window is displayed. See Figure 4. 
  3. Click the Preferences tab, and tick the Track GC cycle time check box under the Method Development section. 
  4. Click OK. 
  5. Run a sequence of at least 2 samples with and GC cycle time setting of at least 5 minutes greater that the GC oven run time plus the time needed for the oven postrun and return to the initial setpoint.
  6. When the sequence is complete, you can find the Calculated GC Cycle Time by clicking in the sequence name.log file in the sequence folder or by looking in Run.log file in each date file folder. See Figure 5 and 6. 


Instrument menu under Method and Run Control. 
Figure 3. Instrument menu under Method and Run Control 
1. Instrument menu, 2. Agilent 7697A Configuration


Agilent 7697A Configuration page. 
Figure 4. Agilent 7697A Configuration page 
1. Preference tab, 2. Tick box for Track GC cycle time


Example of sequence log file. 
Figure 5. Example of sequence log file
1. Sequence name.log file, 2. Calculated GC cycle time


Example of Run.log file.
Figure 6. Example of Run.log file 
1. Run.log file, 2.Calculated GC cycle time 

To track GC cycle time via Masshunter Acquisition

  1. Click Instrument > HS Edit Configuration menu. See Figure 7.
  2. Click on Preferences tab and check the Track GC Cycle Time box under Method Development. See Figure 4.
  3. Run a sequence of at least two samples with and GC cycle time setting of at least 5 minutes greater that the GC oven run time plus the postrun time and time needed for the oven to return the initial setpoint.
  4. The Calculated GC cycle time will be recorded in the mslogbook_history folder in {date and time} mslogbk.htm file. Example is shown in Figure 8. 


Instrument Menu.
Figure 7. Instrument Menu
1. Instrument tab, 2. Option for HS Edit Configuration Second label


Overview of mslogbook_history folder. 
Figure 8. Overview of mslogbook_history folder
1.Directory of mslogbook_history folder, 2. MSlogbk.htm files for sequence run on specific date and time

 

 
Learn how to effectively maintain and operate your Agilent Headspace
SI-7697-2100sV3 - Agilent 7697A Headspace Sampler Operation and Maintenance
SI-8697-2100sV2 - Agilent 8697 and 8697XL Headspace Sampler Operation and Maintenance e-learning courses available from Agilent education