This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any AriaMx optical module:
Steps to Follow
Caution: Finger prints or scratches on the glass window of the optical module (the surface opposite of the label) may impact its functionality. Avoid touching the glass window of the optical modules. Do not put the optical module down with the glass window in contact with the lab bench or other surface.
|
The optical module configuration can be changed either with the AriaMx switched on or off.
AriaMx is Switched On
Figure 4. Optical module shipping box – top piece of foam removed
AriaMx is Switched Off.
Caution: Only move the optical module housing carrier by hand if the AriaMx is switched off and the power cable is disconnected
|
Figure 7. Optical module housing carrier, centered in the instrument door opening
Learn how to effectively operate your Agilent AriaMx real-time PCR system:
AriaMx Real-Time PCR System Setup and User Guide on Agilent.com Aria Real-Time PCR Software User Manual on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.