Hello ivynotterpek ,
I'm not aware of any way to do this with just one compound/cal curve. One solution would be to make duplicate compounds and assign one compound the low level standards and the other the high level standards. You can utilize different…
Thank you, unfortunately we tried to set a collection window much larger than the peak duration, but the peak got split again. And in the version of the software we are using no advanced settings allowing to collect different peaks into the same vial…
Hi CMZainea,
Thank you for your feedback. The leak test for the 6 ports valve was passing, so no leaks detected at the valve. Any other leaks were detected from the visible piping and connection from the back panel of the 7697A going to the 6 ways valve…
The two-way splitter without makeup G3181B can be used as a splitter or a combiner to connect two inlets to one detector. Discuss this part with your local sales rep and they can help you get a quote for this part.
https://www.agilent.com/cs/library…
The Digest of upcoming webinars has been updated as of February 20.
Notable additions include:
BioPharma (1)
GC & GCMS (2)
Pharma (1)
Software & Informatics (21)
Check Out:
Tips and Tricks Workshop on ICP-OES: Smarter Method Development…
This Information Applies To: Agilent 5000 Series ICP-OES Instruments
Issue
The spectrometer incorporates interlocks and covers that are used to monitor critical operations to ensure Operator and instrument protection. If the instrument is used in...
This Information Applies To: Agilent ICP-OES Systems (700 Series, 5000 series)
Issue
If any yellow or orange coloration on any part of the torch is observed (not due to the sample being presented or recently run) or if the torch appears to be visibly...
This Information Applies To: Agilent 700 Series ICP-OES Systems
Issue
The 700 Series ICP-OES requires regular maintenance, and routine maintenance tasks are able to be performed by the instrument operator.
Any maintenance procedures not...
This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900)
Issue
Correct ventilation is required to ensure proper function of the ICP-OES.
Resolution
The ICP-OES must be located under a flue that is vented by an exhaust...
This Information Applies To: Agilent demountable torches
Issue
The Agilent 5000 series ICP-OES instruments can be fitted with a demountable, semi-demountable, or bonded torch. The demountable and semi-demountable torches can be disassembled for...
This Information Applies To: 5100, 5110, 5800, and 5900 ICP-OES instruments
Issue
The ICP-OES operator can identify the source of plasma ignition failure. The main cause is air entering the plasma.
Steps to follow
Excessive air entering...
This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900)
Issue
How to run the ICP-OES Performance Test after installation, preventative maintenance, repairs, and when troubleshooting.
Steps to Follow
The ICP-OES Performance...
This Information Applies To: Agilent 5000 Series ICP-OES
Issue
Blocked nebulizers are the source of many instrument performance issues including loss of sensitivity and poor precision, especially when running organics, high solids, or high salt...
VGA Setup
Multiple Results
This Information Applies To: Agilent VGA 77 (Vapour Generation Accessory)
Issue: This article discusses the setup steps for the VGA 77.
VGA 77
Warning : Noxious Fumes and Corrosive Liquids. Vapor generation methods of analysis involve...
This Information Applies To: Agilent VGA 77 Accessory
Issue: The Agilent Vapor Generation Accessories use high concentrations of HCl and require thorough cleaning to maintain optimum working condition.
Warning : The Vapor Generation...