Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Articles VGA Maintenance and Cleaning
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Atomic Absorption Portal Home
  • Atomic Absorption Portal Articles
  • Atomic Absorption Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Systems Atomic Spectroscopy Accessories ICP OES Systems
  • Component :  Atomic Spectroscopy Sample Introduction Accessories
  • Product Name :  240FS AA 240Z AA 280FS AA 280Z AA 5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES VGA 77 Vapor Generation Accessory
  • Task :  Maintenance Operation
Related Articles
  • Shutting Down the Agilent Vapor Generation Accessory (VGA)

    This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP Issue: The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use. The Vapor Generation Accessory 77 (VGA 77) is a vapor generation system for Atomic Absorption (AA) and the VGA 77P is for Inductively Coupled Plasma Optical Emission Spectrometer (ICP-OES) applications. Unless otherwise specified, both the VGA 77 and VGA 77P will be referred...
  • VGA Setup

    Multiple Results
    This Information Applies To: Agilent VGA 77 (Vapour Generation Accessory) Issue: This article discusses the setup steps for the VGA 77. VGA 77 Warning : Noxious Fumes and Corrosive Liquids. Vapor generation methods of analysis involve...
    • Found In:
    • Articles
    • ,
    • Articles
  • Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization

    This Information Applies To: Agilent Graphite Furnace Atomic Absorption and Zeeman Instruments (240/280(Z) AA) Issue: How to optimize the Instrument to allow for better results, troubleshooting, after maintenance or lamp replacement. Steps to follow: Warning : The furnace runs at temperatures up to 3000 degree Celsius . Allow furnace to cool completely before touching. Optimize one or more lamps: Lamp optimization is required after cleaning or installing a new lamp...
  • Flame Optimization for Agilent Atomic Absorption Systems

    This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA) Issue: How to optimize the Agilent Flame AA for better instrument performance and response stability Steps to follow: Use Figure 1 as reference for the procedure on the Flame AA optimization. Figure 1. Agilent Flame AA instrument diagram Tips : Steps 1 through 4 are performed with the flame off in the Optimization Page of SpectrAA by selecting Optimize Lamps ...
  • Troubleshooting and Maintenance of AA Systems Part 2B- Flame Optimization

    This Information Applies To: Agilent Flame AA Systems: 55B, 240FS, 280FS Issue: After cleaning and re-installing the flame atomic absorption system components, it is important to optimize the burner position, nebulizer uptake rate, flame conditions, glass impact bead , and mixing paddles . Proper optimization ensures that the atomic absorption system will yield accurate and precise results throughout its lifetime. Steps to follow: Optimize the burner position vertically, horizontally...

VGA Maintenance and Cleaning

Answer

This Information Applies To: Agilent VGA 77 Accessory


Issue: The Agilent Vapor Generation Accessories use high concentrations of HCl and require thorough cleaning to maintain optimum working condition.


 Warning 
Warning: The Vapor Generation Accessories uses high concentrations of HCl. Continual exposure to HCl fumes may damage skin and lungs. Wear appropriate PPE when handling and cleaning the VGA components. Always maintain an efficient exhaust system during VGA determinations

The Vapor Generation Accessory 77 (VGA 77) is a vapor generation system for Atomic Absorption (AA) and the VGA 77P is for Inductively Coupled Plasma Optical Emission Spectrometer (ICP-OES) applications. Unless otherwise specified, in this article both the VGA 77 and VGA 77P will be referred to as VGA 77.


Figure 1. Vapor Generation Accessory 77 (VGA 77)

The working life of the absorption cell and the gas/liquid separator will depend on the type of analytical programs carried out. Clean immediately if contamination is suspected during conditioning, or whenever the system has been left unused for some time.

At some stage, the glass will start to turn white indicating the onset of devitrification. Initially, devitrification will not degrade the analytical performance, but eventually the devitrified area will extend over most of the cell at which point it may be necessary to discard the cell. If devitrification is excessive, the cell will eventually crack spontaneously or fail through thermal shock. 

Whether the cell becomes devitrified or not, and even with meticulous cleaning, the inside surface may deteriorate to the extent that the analytical signal will be suppressed. Regularly monitor the performance of the cell using absorbance measurements obtained from standard solutions. Discard the cell when the results of this monitoring become unacceptable. Sometimes, the center section of the cell may gradually soften in the flame and sag. If uncorrected, it will prevent the alignment of the cell with the optical path. When sagging is first noticed, turn the cell over so that it will return to its normal shape the next time it is heated.

Cleaning Procedures:

  • General
  • Absorption Cell
  • Gas/Liquid Separator
  • Fluoro-elastomer Tubing
  • Pump tubing

General:

Always clean up any spilt liquids immediately. The front of the Spectrometer, the burner/torch compartment and, the VGA 77 should be cleaned daily with a soft cloth. If there are spills, use a cloth dampened with distilled water and appropriate PPE.

Absorption Cell cleaning: 

 Tip 
This section applies to the VGA 77 on Atomic Absorption instruments only

Always clean the absorption cell (Figure 2) thoroughly immediately after use, when contamination is suspected while conditioning the system, or whenever the system has been left unused for some time. The cell is fragile and must be handled with appropriate care.

 


Figure 2. The absorption cell

 Warning 
Warning: The cell is heated when in use. Always turn off the flame and allow the cell to cool before attempting to remove it from the sample compartment.
  1. Disconnect the black fluoro-elastomer tubing from the cell inlet.

  2. Release the spring clips and remove the absorption cell from the cell support.

  3. Soak the absorption cell in dilute nitric acid for at least 30 minutes.

  4. Rinse thoroughly with distilled water and allow it to air dry in a dust-free location.

  5. If the cell has been exposed to contamination with potassium iodide, soak the cell in sodium hydroxide solution (0.5% w/v) for at least 30 minutes. Wash thoroughly with dilute hydrochloric acid and then wash thoroughly with distilled water. Allow to air dry in a dust-free location. 

     Caution  Caution: Do not leave the cell in the sodium hydroxide for too long as this solution may accelerate cell devitrification. Avoid handling the central part of the quartz cell with bare hands after it has been cleaned or while it is in use.

Gas/Liquid Separator:

  1. Disconnect the tubing from the vapor outlet of the separator (Figure 3), and the sample inlet.


    Figure 3. The Gas/Liquid Separator

  2. Carefully remove the rubber mounting strap holding the separator in place.

  3. Disconnect the drain tubing from the drain in the bottom of the module. You may now lift the separator out of the module.

  4. Disconnect the tubing from the inert gas inlet of the separator.

  5. If the separator has been exposed to solutions containing potassium iodide, soak the separator in sodium hydroxide solution (0.5% w/v) for at least 30 minutes. Wash thoroughly with dilute hydrochloric acid and then wash thoroughly with distilled water. Allow to air dry in a dust-free location. 

     Tip 
    Tips: If the separator has not been exposed to potassium iodide solutions, it may be treated in the same way as the absorption cell. To avoid KI contamination, it is recommended that you dedicate a separate module to the analysis of those elements requiring prereduction with KI (i.e., As or Sb).
         
  6. Refit the separator into its mounting bracket, secure it with the retaining strap and reconnect all tubing. If the system is not to be used for some time, the separator should be filled with distilled water.

Fluoro-elastomer Tubing

The fluoro-elastomer tubing connecting the gas/liquid separator to the absorption cell is susceptible to contamination with potassium iodide. Agilent recommends using a dedicated separate module for the analysis of those elements requiring prereduction with KI (i.e., As or Sb). If there is no KI contamination, the tubing can be cleaned by flushing it well with distilled water.  If there is KI contamination, remove traces of KI as follows:

  1. Disconnect and remove the tubing.

  2. Soak the tubes in sodium hydroxide solution (0.5% w/v). Wash thoroughly with dilute hydrochloric acid and then wash thoroughly with distilled water.

  3. Allow the tubes to air dry in a dust-free location.

  4. Reconnect the tubing.

Pump Tubes

Check tubing daily and look for the following signs of wear:

  • Loss of elasticity and stretching. Roll the tubing between your fingers. Stretched tubing will feel oval or flat in shape and will not reshape quickly after being pinched. Tubing in good condition will feel round and reshape quickly.
  • Cracks in the tubing, which can cause liquid on the outside or bubbles in the tubing. 
  • Visible sample deposit blocking or partially blocking the tubing

When the system is not being used, the pressure bar should be released, the pump tubes removed from around the rollers and released from the retaining bracket to minimize wear. Spray the outside of the tubes and the surface of the pump rollers with silicone lubricant to reduce mechanical wear. 

An alternative method of removing iodine from the tubing and gas/liquid separator is to pump a freshly prepared 1% w/v sodium thiosulfate solution through the system for 5–10 minutes. Remove the thiosulfate by pumping distilled water through the system for 5–10 minutes.

 Tip 
Learn how to effectively maintain your Agilent Vapor Gas generator:
Agilent Vapor Generation Accessory VGA 77 and VGA 77P Users Guide on Agilent.com

Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help