This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900)
Issue
How to run the ICP-OES Performance Test after installation, preventative maintenance, repairs, and when troubleshooting.
Steps to Follow
The ICP-OES Performance Test runs a suite of tests including Resolution, Sensitivity, and Precision to determine the analytical performance of your instrument and sample introduction components. The test uses the wavelength calibration solution and a blank to create a one-point calibration, then reruns the wavelength calibration solution as a sample.
|
Tips: If 'Advanced Valve System Test' and 'Use Autosampler' are selected on the Tests tab, the Performance Test is not available
|
Prepare the test solutions:
|
Warning: Wear appropriate Personal Protective Equipment when using acids
|
Prepare the instrument
Ensure that the instrument meets the following guidelines before performing the Performance Test:
Running the Performance Test
If the instrument performance tests pass, the system is in good working order. If any of the tests fail, contact Agilent Technical Support for additional troubleshooting.
|