Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES
  • Task :  Troubleshooting
Related Articles
  • ICP-OES Interlocks and Internal Monitoring

    This Information Applies To: Agilent 5000 Series ICP-OES Instruments Issue The spectrometer incorporates interlocks and covers that are used to monitor critical operations to ensure Operator and instrument protection. If the instrument is used in any manner not specified by Agilent, this protection may be impaired. 5800 ICP-OES Warning : No interlock or cover must ever be bypassed, damaged or removed. It is good practice to develop safe working habits that do not depend upon...
  • Cleaning the Agilent ICP-OES Cone

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue The ICP-OES cone is the interface between the plasma and the optics via the orifice at the tip of the cone. Argon passes through the orifice of the cone to produce an oxygen free light path to the optics. Damage to the cone allows oxygen to enter the light path and reduces sensitivity at low wavelengths. Damage occurs during routine operation by build-up of material and corrosion on both surfaces of the cone...
  • Cleaning the Agilent ICP-OES Spray Chamber

    This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES systems Issue The ICP-OES spray chamber filters the sample droplets created by the nebulizer. Larger droplets fall to the chamber drain, while the smaller droplets are introduced to the gas flow going into the ICP torch plasma. A dirty spray chamber causes large droplets to form, which leads to incomplete transfer of sample to the torch, causing poor precision and memory effects. ICP-OES overview The ICP-OES sample introduction...
  • Warning or Error Messages Related to Temperature in ICP-MS Systems

    This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900) Issue Error messages related to temperature can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light and stay on, it is necessary to correct the issue. The troubleshooting procedure in this article is applicable for the following warnings/errors: 1432 Error: Inside housing temperature is too high. 1430 Error: Air intake temperature is too...
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...

Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES

Answer

This Information Applies To: 5100, 5110, 5800, and 5900 ICP-OES instruments


Issue

The ICP-OES operator can identify the source of plasma ignition failure. The main cause is air entering the plasma. 


Steps to follow

  1. Excessive air entering the plasma. Check the following:

    1. The peristaltic pump tubing is clamped and has sufficient clamping pressure. Usually, air is being pulled into the nebulizer because the pump tubing is not properly clamped during ignition. At ignition, the plasma is sensitive to too much air being self-aspirated into the plasma.

    2. Check that the autosampler probe is down inside the rinse vessel and that the rinse solution is flowing properly to the nebulizer

  2. The torch may be damaged, dirty, or not assembled correctly. Try a new torch. If the torch end is highly contaminated it can become conductive, making plasma ignition less consistent.

  3. The torch contains liquid sample. Clean and dry the torch.

  4. The nebulizer may have some damage and leak.

  5. There may be a leak at nebulizer gas fitting. Check to ensure the nebulizer gas tubing from the instrument to the nebulizer fitting is tight and secure.

  6. The spray chamber is not draining properly. If the spray chamber is not draining properly, due to insufficient pressure on the peristaltic pump pressure arms, then water droplets will go up into the torch and prevent ignition. 

  7. Check the drain tubing fitting and ensure that the rinse solution is flowing out the spray chamber properly. Always check the torch and ensure it is clean and completely dry.

  8. A large droplet may have entered the plasma. Check that the spray chamber is clean and nebulizer is producing a fine mist. If using a single pass spray chamber, consider changing to a double-pass spray chamber.

  9. Insufficient plasma flow and/or excessive auxiliary flow.

  10. Work coil damaged and bent down.

  11. Problem with ignitor. Check for a visible spark and listed for the ignition sound.

  12. Check that the argon supply has not recently changed.

  13. Check that there is sufficient argon supply and pressure is in the range 500 to 600 kPa (73 to 88 psi). It is better to set the pressure at the regulator to 600 kPa (88 psi) to identify possible line pressure drops. Ideally a second regulator has been installed within three meters or 10 feet of the instrument. If running 30 m of 6 mm or 1/4-" ID tubing to the instrument at 500 kPa, the line pressure drop is adequate to affect plasma ignition. If possible, watch the regulator while trying to ignite the plasma, and watch to ensure that the pressure doesn't fall below 80 psi. Raise the pressure if necessary.

  14. This test identifies a leak in the sample introduction system.

     Warning 
    Warning: The plasma operates at 10,000 K. Prevent damage to the torch by turning the plasma on for a short period of time. Do not exceed 10 seconds.  

    Unclamp and remove the spray chamber from the bottom of the torch. Seal the bottom of the torch with Parafilm and attempt ignition. If the plasma lights, the sample introduction system will have an air leak. Check that the nebulizer and spray chamber connections are sealed, in particular the helix seal.

Contact you local Agilent Service Center if you re unable to identify and correct the problem.   

 Tip 
Learn how to effectively troubleshoot your Agilent ICP-OES:
ICPOES-MULTI-2210e - Agilent 5000 Series ICP-OES Troubleshooting System Issues  e-learning course available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help