Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Articles VGA Setup
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Atomic Absorption Portal Home
  • Atomic Absorption Portal Articles
  • Atomic Absorption Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Systems Atomic Spectroscopy Accessories ICP OES Systems
  • Component :  Atomic Spectroscopy Sample Introduction Accessories
  • Product Name :  240FS AA 240Z AA 280FS AA 280Z AA 5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES VGA 77 Vapor Generation Accessory
  • Task :  Maintenance Operation
Related Articles
  • Shutting Down the Agilent Vapor Generation Accessory (VGA)

    This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP Issue: The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use. The Vapor Generation Accessory 77 (VGA 77) is a vapor generation system for Atomic Absorption (AA) and the VGA 77P is for Inductively Coupled Plasma Optical Emission Spectrometer (ICP-OES) applications. Unless otherwise specified, both the VGA 77 and VGA 77P will be referred...
  • VGA Maintenance and Cleaning

    Multiple Results
    This Information Applies To: Agilent VGA 77 Accessory Issue: The Agilent Vapor Generation Accessories use high concentrations of HCl and require thorough cleaning to maintain optimum working condition. Warning : The Vapor Generation...
    • Found In:
    • Articles
    • ,
    • Articles
  • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue The ICP-OES torch plasma ionizes the introduced gas or aerosol stream. In normal operating conditions, incomplete ionization occurs and material is deposited on the torch surface. Build up of deposits can create blockages, damage the torch or interact with the sample stream causing poor precision, accuracy, and memory effects. ICP-OES overview The ICP-OES sample introduction components are detailed in...
  • How to Clean the Ion Capillary in Agilent 6000 Series LC/MS Instruments

    This Information Applies To: Agilent 6000 series LC/MS instruments Issue: Cleaning the ion transfer capillary as part of regular maintenance or as part of troubleshooting when a loss of sensitivity in the instrument occurs. Steps to follow Cleaning the ion capillary is a part of regular maintenance of your LC/MS instrument and can also help to regain instrument sensitivity. There are two recommended methods for cleaning the ion capillary. Syringe adapter method Ultrasonic cleaning...
  • Flame AA Burner Maintenance

    This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280) Issue: This article describes when and how to clean an AA burner. Background For optimum and safe performance, burners (see Figure 1 ) should be cleaned after each day of use. Particularly if the solutions being analyzed have a high solids content or a rich nitrous oxide-acetylene flame has been used. If not cleaned regularly, hard deposits that are difficult to remove may build up...

VGA Setup

Answer

This Information Applies To: Agilent VGA 77 (Vapour Generation Accessory)


Issue: This article discusses the setup steps for the VGA 77. 


VGA 77

 Warning  Warning: Noxious Fumes and Corrosive Liquids.
Vapor generation methods of analysis involve the generation of toxic hydrides and require the use of materials that are toxic, highly corrosive, or otherwise hazardous. Careless, improper, or unskilled use of such materials can cause serious personal injury. Always ensure that laboratory safety practices governing the use, handling and disposal of such materials are strictly observed. These safety practices should include the wearing of appropriate safety clothing. In the VGA 77 system, highly concentrated acids are being pumped under pressure. If a leak occurs, acid could be sprayed from the system and cause serious personal injury. Always wear approved safety glasses warranted to protect the eyes. Never pump concentrated sulfuric acid through the VGA 77 system.

Steps to Follow:

  1. General Setup
  2. Peristaltic Pump Adjustment
  3. AA Setup
  4. Conditioning the System    


1. General Setup

  1. Ensure that the inert gas supply (argon or nitrogen) to the VGA 77 is turned on, with the cylinder regulator set to the recommended pressure of 350 kPa (50 psi).
  2. Ensure that the peristaltic pump is adjusted correctly as detailed below. 
  3. Set the instrument parameters for the element to be determined. Use the integration mode with Agilent AA systems.
  4. Fill the reagent container(s) with the required reagent(s). See VGA 77 User’s Guide (Section 5.)


2. Peristaltic Pump Adjustment

       1. Check that the gas supply is set to the recommended pressure (350 kPa (50 psi)) and turn on the inert gas supply.

 Tip 
Tip:  Ensure that the inert gas supply is turned on before switching on the VGA 77 power. If solutions are pumped through the system before the inert gas is turned on, the signal will be unreliable.

       2. Place the three capillary tubes in a container of distilled water and switch on the VGA 77. The green ‘Power’ indicator light is illuminated when the power is on. 

 Tip 
Tip: The peristaltic pump will run continuously once the power is switched on.

      3. Slacken both pressure adjusting screws. Swing both tube beds against the pump tubing and close the pressure bar.

      4. Slowly tighten the front (sample) pressure adjusting screw until water is being pumped through the tube. Tighten the pressure adjusting screw a further half turn. At this setting, the pumping rate will be close to optimum. Any further tightening of the screw will tend to reduce the pumping rate. 

 Caution  Caution: Do not overtighten the pressure adjusting screws. Excessive tightening will shorten the life of the pump tubes and could cause permanent damage to the pump.
 Tip 
Tip:  At this stage, inert gas may be seen bubbling out of both the other tubes. This is a normal consequence of pressurizing the system and this situation is automatically corrected when you adjust the pumping pressure in these tubes.
 Tip 
Tip:  To see if the liquid is being pumped smoothly, you can deliberately create air bubbles by dipping the capillary in and out of the container

      5. Repeat Step 4 with the rear pressure adjusting screw.

      6. Measure the uptake rate through the sample pump tubes. The uptake rate should be within the range 6–8 mL/min.

      7. Measure the uptake rate for each of the other pump tubes. The uptake rate for each should be within the range 0.8–1.2 mL/min.

 3. AA Setup

     1. Remove the cell in its holder from the burner and place it at the front of the reagent module.

     2. Lower the burner so that it is clear of the optical path. It is recommended to use reduced slit height (0.5) in all methods.

     3. Fit the required hollow cathode lamp and set up the Spectrometer as described in the instrument operation manual.

     4. Select Optimize > Optimize Lamp from SpectrAA Software.

     5. Position the cell in its holder on top of the burner.

     6. Align the cell in the optical path as follows:

         a. Hold a piece of white card between the right end of the absorption cell and the sample compartment window.

         b. Using the burner positioning controls, adjust the position of the cell until the light from the hollow cathode lamp passes through the cell onto the card. Remove the card.

         c. Use the burner positioning controls to adjust the position of the cell for maximum transmission.

 Lighting the Flame

      1. Carry out any safety checks described in the instrument operation manual. Follow all instructions and observe all warnings.

      2. Set the acetylene flow higher than normal to obtain a fuel-rich air-acetylene flame. This is required for igniting the flame.

 Warning  Warning: Explosion Hazard. With the VGA 77 cell installed, the burner will be lower than usual, and a standard mixture or a fuel-lean mixture may not ignite readily. To avoid an explosive accumulation of fuel and oxidant in the sample compartment, ALWAYS ensure that the gas mixture is fuel-rich before operating the Ignite button.

      3. Ignite the flame and adjust the fuel flow to give a lean flame. If necessary, lower the cell back to its normal position.   

 Caution  Caution: If the absorption cell is left in a fuel-rich flame, carbon particles may be deposited on the outside wall of the cell. Under these conditions, the cell can be permanently damaged by local overheating.

 4. Conditioning the System

 The complete system requires “conditioning” to obtain a stable signal before calibrating the instrument and analyzing samples.

This procedure should be carried out:

  •  When the system is first installed
  •  After the system has been completely flushed and cleaned
  •  When a new absorption cell (or flow-through mercury cell) is fitted
  •  Whenever new pump tubes are fitted
  •  When the reaction coil is renewed

       1. Allow the pump to operate for three or four minutes to stabilize the flow rates.

      2. Make sure that the instrument is in ‘Absorbance’ mode with a suitably short integration period.

      3. Aspirate the highest standard, monitor the signal, and measure the time taken for the signal to rise from zero to its stable value. The measurement for this interval will be the pre-read delay that should be entered into the system software. 

 Tip  Tip:   When presenting solutions to the system, always wait until this period has elapsed before taking a reading.

      4. Aspirate the analytical blank.

      5. Continue to aspirate the standard and analytical blank in turn until a consistent response is obtained for the standard solution and then proceed to ‘Operation’ in VGA 77 User’s Guide (page 51).

 Tip 
Learn how to effectively maintain and operate your VGA 77: VGA 77 User’s Guide, VGA 77 High Productivity Accessory, VGA Supplies, Elemental Determinations using VGA

Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help